The user is experiencing higher than expected low-frequency noise and thermal drift on a custom PCB using the ADR1001. The final answer suggests referring to the ADR1001 datasheet for stability and noise data, noting that the evaluation board's layout closely matches the datasheet's recommended design. Accompanying answers highlight the importance of the GND path to avoid IR drops and suggest checking all thermocouple connections for additional thermal errors.
AI Generated Content
Category: Hardware
Product Number: Voltage Reference
Hi all, could anyone share the test results from the ADR1001 official evaluation board? I’m specifically looking for:
  • Stability data (Thermal drift plots)
  • Noise data (Spectrum/FFT and Time-domain plots)
Any insights or shared data would be greatly appreciated. Thanks!
  • Hi 00yujie,

    I'd refer to the ADR1001 datasheet for specs and plots for stability and noise data. The layout used in the evaluation board was designed to demonstrate best practices, and is very similar to layouts of boards used to larger scale datasheet data collection.

  • Thank you for your advice. Currently, the issue is that the noise (below 100Hz) measured on the PCB I built following the datasheet's recommended design is significantly higher than the values specified in the datasheet. The stability is also noticeably worse. Therefore, I would like to check the noise and stability performance of the official demo board to determine if my suboptimal results are a reasonable discrepancy due to the transition to a custom PCB, or if there are flaws in my circuit design. I would greatly appreciate it if any experts could share the official demo board's test results.
  • My test results are as follows:
    1. Stability: Measured at 0.6 ppm/°C, whereas the ADR1001 datasheet specifies 0.2 ppm/°C.
    2. Noise Performance (compared to Figure 3 in the datasheet; my system's noise floor is low enough for precision): 
    • Above 100 Hz: Results are consistent with the datasheet.
    • Below 100 Hz: Noise is significantly higher than the datasheet specifications, with the discrepancy increasing at lower frequencies.
    • Specific Noise Levels:
      • 100 Hz: 20 nV
      • 50 Hz: 30 nV
      • 10 Hz: 100 nV
      • 1 Hz: 500 nV
      • 0.5 Hz: 700 nV
      • 0.1 Hz: 1000 nV
  • Testing environment/setup:
    The chip is battery-powered, and the PCB features thermal isolation slots around the device. Additionally, both the top and bottom surfaces of the PCB are covered with anti-draft foam.
    Given these precautions, I would appreciate any expert guidance on potential PCB design flaws that could be contributing to the elevated low-frequency noise. Thank you for your help in improving the noise performance!

  • It's our experience that the GND path is critical. Any IR drop is going to add TC error (Cu is >4000ppm/C TCR). You've already addressed the air current issue with foam, but please check all thermocouples carefully. This can include the connections to the meter you are using. Any thermocouple will have additional TC and can look like low-frequency noise.