ADXL345 Self test
Question asked by meir.mardechayev@phnx.co.il on Apr 19, 2015
Latest reply on May 4, 2015 by venkat
One of my customer uses Digital Accelerometer ADXL345 of Analog Devices.
We would like to implement the self-test for ADXL345 in our SW according to its data sheet. We use the full resolution of ADXL345 (13 bits, -16 to +16g).
We understood, that according to the “Self-test” chapter of datasheet, the Table 15 of X, Y and Z limits should be used for results estimation.
The “Using self-test” explains how to use it:
A) We need to measure acceleration results with axis offsets 0 and self-test mode OFF - A st_off
B) We need to measure acceleration results with axis offsets 0 and self-test mode ON   - A st_on
The difference between B and A results is our test value, that we compare with axis limits from Table 15
A st_change = A st_on – A st_off
Test passed if: AXIS LIMIT MINIMUM < A st_change < AXIS LIMIT MAXIMUM
(all limits and results are in ADXL units).
The 1-st question:
We performed tests and
- obtained results after sending to ADXL345 the setup with axis offsets 0 and self-test mode ON   - A st_on
- obtained results after sending to ADXL345 the setup with axis offsets 0 and self-test mode OFF - A st_off
We noticed, that the difference A st_change is affected only on the VERTICAL axis of ADXL345(has the G) and only on the VERTICAL axis we have the results meet the test logic.
The question: Is it correct ? Is test relevant only for currently VERTICAL axis ?
The 2-nd question:
How we should consider the positive or negative SIGN of  A st_on and A st_off when we calculate the difference ?
If we calculate the difference considering the results sign, the A st_change could be out of its limits.
Example of real tests made by us:
Y axis of ADXL345 is VERTICAL axis and directed UP.
A st_on = - 2
A st_off = 256(approximately 1G in ADXL345 units)
A st_change = -2 - 256 = -258
According to Table 15:
Y AXIS LIMIT MINIMUM  =   -540
Y AXIS LIMIT MAXIMUM =   -50
-500 < -258 < -50  - TEST PASSED
Y axis of ADXL345 is VERTICAL axis and directed DOWN.
A st_on = 5
A st_off = -256(approximately -1G in ADXL345 units)
A st_change = 5 - (- 256) = +261
TEST FAILED
The question: Is it correct to take the sign of results under consideration ? May be we should use the ABLOLUTE values  in calculations ?
Meir