Which Thermocouple type to use with 5B47 for -100C to 300C
The user is building an application to measure air temperatures between -100°C to 300°C using the 5B47 Isolated, Linearized Thermocouple Input. Despite the 5B47's highest accuracy with type J thermocouples, type J is not suitable for temperatures below 0°C due to condensation and rust concerns. Type T thermocouples are recommended for this temperature range, offering better low-temperature performance.
I am building an application to measure air temperatures between -100 deg C to 300 deg C.  I'm considering using the  5B47 Isolated, Linearized Thermocouple Input from AD.  The spec sheet for the 5B47 says that the highest accuracy for this temperature range is available from a type J thermocouple.  (+/- 0.5 deg C accuracy).  However, all the thermocouple manufacturers I have looked at say type J is not rated for temperatures below 0 deg C.  They all recommend type T thermocouples.  What am I missing?  Should I use type J or type T?
  • Hi seanengtxct,
    I'm not very familiar with the 5B47, but generally speaking, type J thermocouples are not used to measure temperatures below ambient due to concerns about condensation forming on the wires, which can cause the iron to rust. Type T thermocouples are good for cryogenic and laboratory applications where low temperatures are needed and the higher temperature range of type J is not needed. It sounds to me like type T would fit your measurement temp range very well, and personally I would choose type T despite the slightly lower accuracy of the module.
    Scott
    Thank you,
    EZ Admin