ADG417 as a thermocouple switch?

Thread Summary

The user seeks advice on designing a simple frontend for a sapflow sensor with a datalogger that has a single analog input and requires 10uV resolution. The engineer recommends the second circuit, which uses two amplifiers (AD620) and a switch (ADG417) to minimize errors from switch leakage, over the first circuit. The user acknowledges the advice and plans to prototype and test the recommended solution.
AI Generated Content

Hello,

First sorry for my English. I have a problem. In short time I have to make a simple frontend for sapflow sensor. The measuring value will be recorded by datalogger. I have two thermocouplers but only one analog input in datalogger. Unfortunatelly I have limited access to Analog Devices parts. For example there is ADG417 and AD620 at fingertips. Here are my conditions:

-single supply: 5VDC or 12VDC

-logger input voltage range: 0..5V/16bit resolution

-resolution: 10uV (from sensor voltage)

-sampling frequency: one per hour

There are two circuits on pictures below. Please tell me are these circuits correct? I know that sapflow1.png is a worse solution but there is only one AD620. Maybe the ADG417 noise aren't to big to receive 10uV resolution. Which one I should to choose? Maybe some another one?