ADG5413 high impedance

Thread Summary

The user inquires if the ADG5413 can be supplied with 5V to keep switches open (high impedance) without switching, due to safety requirements in a medical device. The final answer confirms that 5V is outside the ADG5413's operational range, and the switch positions cannot be guaranteed. The recommended solution is to use the ADG5413F, which remains off and high impedance when no power is supplied, despite its higher cost.
AI Generated Content

Hi,

Is it allowed to supply the ADG5413 with 5V when all switches only have to remain open (high impedance)?

During this device state the inputs off the ADG5413 will be 2.5V +- 0.3V

When the ADG5413 is switching, it's supply will be >=9V (Minimal listed single supply voltage).

The 5V supply is required during the state where the ADG5413 only has to remain high impedance due to single fault safety of a medical device.

A alternative is using the ADG5413F because quoting from the datasheet:

When no power supplies are present, the switch remains in the off condition, and the switch inputs are high impedance.

The ADG5413F is considerable more expensive though. Since I need 12 of these muxes per device, this would add quite a lot to the cost price. Also the over voltage protection of the ADG5413F is not required.

Thanks,

Koen