4Wire Measurment - Influence of Leakage Currents
The user asks how to model leakage currents in a 4-wire RTD measurement setup using CMOS analog multiplexers (e.g., ADG7xx/ADG79x-class devices). The final answer suggests that the datasheet provides worst-case leakage values, and actual leakage can be determined through physical measurement. The user also inquires about the interpretation of leakage current values for 16 channels, specifically whether the leakage of inactive channels is included in the on-state leakage current of the active channel.
Product Number: ADG706
Hi all,
I have a conceptual question about modeling leakage currents in CMOS analog multiplexers in a 4‑wire RTD measurement setup, and I’d appreciate some expert feedback.
Setup (simplified):
  • DMM: Keysight 34465A in 4‑wire RTD mode (Pt1000)
  • 4‑wire connections (force+/force−, sense+/sense−) to multiple Pt1000 sensors
  • Channel selection via CMOS analog multiplexers (e.g. ADG7xx/ADG79x‑class devices)
  • Ideally, all four leads per RTD are switched through a MUX path so that only one sensor at a time is connected to the DMM.
I have read quite a bit about internal leakage mechanisms (reverse‑biased PN junctions, parasitic diodes, “channel on leakage”, “off leakage”, etc.). For example, TI’s application brief models a MUX channel as an ideal switch plus a leakage current source at the channel pin that flows through the input impedance of the following stage and creates a DC offset error TI_MUX_ErrorInformation.pdf. The ADG798/ADG5298 paper similarly uses “channel on leakage” as a key per‑channel parameter 67745-high-prec...nments.pdf.
My problem is not whether leakage exists, but how to model it:
  • In theory, one could go down to transistor level (P‑MOS/N‑MOS, body diodes) and derive leakage for each PN junction.
  • In practice this is not tractable, and the datasheets already collapse all of that into per‑channel specs like “drain/source/channel leakage”.
So I’d like to model my setup at system level by:
  • assuming an effective leakage current Ileak,effI_\text{leak,eff}Ileak,eff at the sense node,
  • composed of on‑leakage of the active MUX channel plus off‑leakages of all inactive channels that are “visible” at the bus,
  • and then using ΔU=Ileak,eff⋅Rin\Delta U = I_\text{leak,eff} \cdot R_\text{in}ΔU=Ileak,effRinΔR=ΔU/Imeas\Delta R = \Delta U / I_\text{meas}ΔR=ΔU/ImeasΔT=ΔR/(dR/dT)\Delta T = \Delta R / (dR/dT)ΔT=ΔR/(dR/dT) to estimate the RTD temperature error.
My concrete questions:
  1. Mapping datasheet → model:
    When a datasheet specifies “drain leakage”, “channel on leakage”, or I_S(off)/I_D(off):
    Is it reasonable to interpret these values as an effective leakage current source at that physical pin (e.g. the COM/sense node), as TI’s model suggests? In other words:
    – active channel → on‑leakage at the COM pin,
    – inactive channels → off‑leakage at their pins that are connected (directly or via COM) to the sense bus?
  2. Direction/sign of leakage currents:
    Datasheets usually give magnitudes (±x nA) but not exact directions.
    From your experience, is it acceptable at system level to model symmetrically distributed leakage currents (some pulling current out of the sense node towards the rails, some pushing current into it), such that a realistic net leakage in the pA range results?
    Or are there common assumptions (e.g. “dominantly towards VDD”) that you would apply in such a model?
  3. Abstraction level:
    For an uncertainty analysis, is it technically sound to ignore the detailed P‑/N‑MOS junction level and rely on the per‑channel leakage model at the sense node (as TI/ADI effectively do), or am I missing an important effect by not going down to transistor detail?
  4. Practical observation:
    Experimentally, I see only about 0.1–0.2 °C difference between measurements “with MUX” and “without MUX” for Pt1000. Back‑calculating this corresponds to an effective leakage current through the DMM input in the few‑pA range.
    This is much smaller than the nA worst‑case figures from the datasheets, but seems plausible if many individual leakages partially cancel.
    Would you agree with this interpretation, or are there more typical explanations for such small net currents?
I want to justify this modeling approach cleanly in a write‑up and I’m looking for best practices on how MUX leakage is usually modeled in 4‑wire measurement chains:
– purely as per‑channel leakage at the sense node,
– or with additional assumptions about current distribution / direction?
Thanks a lot for any insights!
  • Hi,  .
    I'm currently into this and will get back to you.
    Christian
  • Hi Christian,
    While searching for more information about leakage currents in multiplexers, I found this document:
    ez.analog.com/.../4760.Switches-and-Multiplexers_2D00_Advanced.pdf
    The chapter on leakage currents was helpful, and I have some assumptions I want to share with you.
    The chapter delving deep into the causes of leakage currents was helpful, and I have some assumptions I would like to share with you.
    First, the leakage current is caused by internal reversed bias diode junctions. OK, there is an NMOS and a PMOS (if I consider the MUX internal circuit is like the one described in the paper), so there are some diodes from source to VDD, drain to VDD, or source to VSS, drain to VSS. Am I right?
    Secondly, if the first assumption is true, I would consider the path of the leakage current to go from one MUX in the positive wire through the DUT (with a resistance of about 1 kΩ) to the MUX in the negative wire (where the sense wires are mostly important). Then there is the question of whether the current takes the path through the multimeter or whether it is possible for the current to flow through VCC/VDD/Source back to the first MUX. (The source is a LDO, if my research is true the resistance is much lower than the multimeters input resistance, so i assume the leakage to take this way)
    Thirdly, in further measurements 4 Wire + Mux and only 4 Wire on the same DUT there was an even less difference (0.03°C to 0.08°C)...
    I hope you can understand my line of thinking.
    Thanks!
  • Hi,  .
    Your assumptions are on the right track. The only key refinement is that, in a 4-wire system, the dominant error mechanism is not leakage through the DUT but leakage into the high-impedance sense node. Most leakage currents are shunted through low-impedance paths (e.g., supply rails), leaving only a small residual current that creates the observed pA-level error. This also explains why your measured temperature deviation is significantly smaller than worst-case datasheet estimates.
    Christian
  • Hi Christian, thank you very much for your feedback.
    Am I understanding this correctly: While you can roughly calculate the effect using the datasheet, it only shows the worst-case scenario? So the actual situation can only be accurately determined through my own physical measurement?
    I also have a question about the leakage currents (drain and source) when the device is turned on: How exactly should the values for the 16 channels be interpreted? Is the leakage current of the inactive channels already included in this value?
    Gabriel