ADG1606 Multiplexer used for impedance measurements at low frequencies

Category: Datasheet/Specs
Product Number: ADG1606

Dear

I was looking for a multiplexer to speed up my impedance measurements. I have metal traces with a coating as samples to test. I want to test the insulation integrity over time. On my sample, I have one low-impedance channel and multiple high impedance channels, they are measured to be > 100 MOhm at 100 Hz and linearly increase to GOhm at 1 Hz. When measuring with the multiplexer, it is in this region that the measurements are influenced by the multiplexer for only the high imepdance channels.  In terms of grounding and removing noise, I use a farday cage with the floating ground of the potentiostat connected to it. I also use the same earth ground for the ground of the potentiostat, arduino and multiplexer. 

I looked at the Idc values of my potentiostat when measuring and they look like this: 
 
Mux high impedance channel
  • -8,859613E-005 op 1 MHz
  • -1,672427E-008 op 1 kHz
  • -1,680497E-008 op 1 Hz
High impedance channel without multiplexer
  •  -7,194425E-007 op 1MHz
  • -2,132992E-011 op 1 kHz
  • 5,289324E-014 op 1Hz
The cannel-to-channel capacitance or crosstalk capacitance:
  • Cd off: 248 pF
  • Cd, Cs On: 271 pF
The Channel On Leakage, ID, IS (On): +/- 0.3 nA 
 
Could this cause the influence or is there something else i need to take into account? Are there better suitable types of multiplexers do these kind of measurements at low frequencies, like ADG1208 for example?

Kind regards