AD8302 Eval Board – VPHS Output Behavior and SW2 Function

Category: Hardware
Product Number: AD8302

Hi Team,

I’m testing the AD8302 Evaluation Board to measure phase difference between two RF signals at 1500 MHz, −10 dBm each. I’m seeing unexpected results at the VPHS (phase output) SMA connector:

  • No input signals:
     SW2 set to “M” → VPHS fluctuates between ~1.13 V and 1.2 V
     SW2 set to “C” → VPHS steady at ~2.4 V

  • With both RF inputs connected (same frequency, unknown phase):
     VPHS reading remains unchanged from the no-signal case.

Could you clarify:

  1. What is the exact function of SW2?

  2. Is ~2.4 V at VPHS a valid phase output, or is the device saturated?

  3. What is the correct procedure to measure phase difference in this setup?

Thanks in advance for your support.

  • The AD8302 has two modes that are set by SW1 and SW2. If these two switches are open (VMAG not connected to MSET and VPHS not connected to PSET (see Figure 14 in datasheet), the device operates in a comparator mode, that is, the VMAG and VPHS signals are digital output that tell you whether INPA is bigger or smaller than INPB and whether the phase of INPA is greater or less than the phase of INPB.

    If you want to measure the phase between two signals, you do not want to be in this mode. So set SW1 and SW2 so that VMAG is connected to MSET and VPHS is connected to PSET. 

    If your two signals are not referenced to the same source, the phase output voltage will wander. If you have two signal generators, you should connect both to the same 10 MHz reference (or slave one off of the other). An alternative way to test the board is to use one signal generator. Feed the output of the signal generator into a power splitter and feed the two outputs of the power splitter into INPA and INPB. Then vary the relative cable lengths between the power splitter and the INPA and INPB inputs. 

     

     

    I’m testing the AD8302 Evaluation Board to measure phase difference between two RF signals at 1500 MHz, −10 dBm each. I’m seeing unexpected results at the VPHS (phase output) SMA connector:

    • No input signals:
       SW2 set to “M” → VPHS fluctuates between ~1.13 V and 1.2 V
       SW2 set to “C” → VPHS steady at ~2.4 V

    • With both RF inputs connected (same frequency, unknown phase):
       VPHS reading remains unchanged from the no-signal case.

    Could you clarify:

    1. What is the exact function of SW2?

    2. Is ~2.4 V at VPHS a valid phase output, or is the device saturated?

    3. What is the correct procedure to measure phase difference in this setup?

    Thanks in advance for your support.