Information about the evaluation board
Product Number: EVAL-CN0350-PMDZ
I purchased a EVAL-CN0350-PMDZ evaluation board, i want a few clarifications about the board which are defined as follows:
1. I want a complete schematic diagram of the EVAL-CN0350-PMDZ board, other than which is given in circuit note CN-0350.
2. I am using 1510A precision signal generator for generating charge signals which is used for giving input to the board. Now, i want to know the expected input (charge) and output (voltage) relation for the board so that i can verify it by giving suitable input charge signals.
3. I also want to know the minimum value of AC charge signals which is supported by the board for conversion to voltage.
Thread Notes
  • Hi  ,
    Moving your query to reference designs. Someone here will be able to help you with this.
    Francis
  • Hi  ,
    Please see the schematic and answers below. Let me know if this works for you.
    1. I have attached a copy of the Schematic for EVAL-CN0350-PMDZ
      02-047068-01-a-1.pdf
    2. The CN‑0350 signal chain consists of a charge amplifier stage (U1A with feedback capacitor C2) followed by a non‑inverting gain stage (U1D, set by R7 and R8), with both stages referenced to a mid‑scale offset voltage V_HREF of 1.25 V. The charge injected at the summing node is integrated onto C2 and then scaled by the second stage, giving the overall transfer function:
      When you drive the board through the CAL input (TP1–TP2) using your 1510A precision signal generator, the injected charge is defined by:
      where C_CAL is the calibration capacitor in series with the 1510A output. Substituting this in, the end‑to‑end relationship becomes:
      Using the default board values (C2 = 1 nF, C_CAL = 1 nF, R8 = DNP giving unity gain), a ±1 V peak signal from the 1510A injects ±1000 pC of charge and produces an ADC output swing between 0.25 V and 2.25 V centered around 1.25 V, which matches the calibration result shown in Figure 4 of the circuit note. Please also ensure that the 1510A output is floating (isolated) with respect to the board ground, since the CAL input sits at the 1.25 V HREF common‑mode level.
    3. There are two limits to consider — one in amplitude and one in frequency. On the amplitude side, the smallest resolvable charge is set by the 12‑bit AD7091R ADC's 1 LSB step, given by:
      Referred back to the input through the charge amplifier:
      at the default unity gain with C2 = 1 nF. Accounting for AD8608 noise and offset contributions, a reliably distinguishable AC charge is on the order of a few pC peak. On the frequency side, the low‑frequency corner of the charge amplifier is set by R3 in parallel with C2:
      with the default R3 = 100 MΩ and C2 = 1 nF — so AC signals below roughly 1.6 Hz will be attenuated by the DC feedback path and should be avoided. For best results during your verification, I would recommend keeping the test frequency in the flat region between approximately 10 Hz and 100 kHz (the circuit note itself uses 1 kHz for calibration), which keeps you well inside both the low‑ and high‑frequency corners of the design.
    Michelle
  • Hi  !
    Do you still need any assistance?
    Regards, Elena