Question About Compatibility of EVAL-CN0579-ARDZ with PCB393B04 IEPE Sensor

Thread Summary

The user is troubleshooting the EVAL-CN0579-ARDZ evaluation board with a PCB393B04 accelerometer, which requires 18-30 VDC excitation and 2-10 mA constant current. The support engineer confirmed the board can provide these and suggested checking the voltage at P6-P9 test points, ensuring the current sources and switches (U9, 11, 13, 15) are enabled. The `iio_info` command should be run to verify system and driver functionality.
AI Generated Content
Category: Hardware
Product Number: AD7768-4

Dear Sir or Madam,

We are currently trying to use the EVAL-CN0579-ARDZ evaluation board together with a PCB393B04 accelerometer in one of our research projects. However, we have not been able to obtain a signal so far.

According to the documentation, the EVAL-CN0579-ARDZ board is compatible with IEPE sensors. We would like to ask whether you have any experience using this board with the PCB393B04 sensor, and whether these two devices are fully compatible. Additionally, we would appreciate any guidance regarding important considerations or configuration steps when using this setup.

The PCB393B04 sensor requires a constant excitation voltage in the range of 18 to 30 VDC. ICP/IEPE sensors contain a small internal pre-amplifier circuit, and the excitation voltage powers this circuit. If this voltage is not provided, the sensor will not produce any signal.

In addition, the sensor requires a constant current supply between 2 and 10 mA. This constant current powers the internal electronics of the IEPE accelerometer, enabling it to generate the acceleration signal and transmit it through the same cable.

I couldn't determine whether the issue is related to the P17–P20 solder jumpers or if it is caused by something else. Additionally, could you explain the function of the P6–P9 solder jumpers?

We would greatly appreciate your advice on whether the EVAL-CN0579-ARDZ can provide the required excitation voltage and constant current for this sensor, and whether there are any recommended settings or precautions when using this combination.

Thank you very much for your time and support.

Kind regards,