CN0565‘s accuracy for resistance or impedance measurement

Thread Summary

The user is facing issues with the CN0565 board's resistance measurement capability, including saturation at 10kΩ and poor sensitivity for resistors between 5kΩ and 7kΩ. The support engineer suggests selecting an appropriate RTIA value based on the DUT impedance to improve accuracy and avoid ADC saturation. The user also mentions a connection issue with the ADICUP3029, which may be a common problem, and is now exploring the AD5933 for their EIT research.
AI Generated Content
Category: Datasheet/Specs
Product Number: CN0565

I found this figure in CN0565's circuit note for its resistance measuing capability. However I'm having trouble reproducing this result.

 


Figure below is my experiment setup, I'm just connecting electrode 0 & 1 to different resistors (10~1MΩ), configuring 500mV and 10kHz, running 'python cn0565_example_single.py 0 1 1 0' and reading the raw magnitute. However I found

  1. The saturation point is around 10kΩ. (see figure)
  2. The sensitivity is not that right, I'm reading the same number for 5kΩ and 7kΩ.
  3. If I turn to 'impedance mode' (other than magnitute mdoe), seems only 1kΩ is correctly measured (maybe because of calibration). Others show around 20% error.




I also found an paper using a similar setup to evaluate this board, obtaining similar result of mine: CN0565 – Multichannel Bipolar Impedancemeter Evaluation | Springer Nature Link. 

In fact, I'm trying to use this board to explore human body EIT research (to penetrate through skin and sense subtle impdance change), so I really need to test its sensitivity and penetration. May you please tell:

  1. Is there anything wrong with my evaluation setup? Since I can always pass the test board (cn0565_prod_test.py), could it be the issue of where AGND is connected to?
  2. Is this board potentially capable of non-invasive body impedance sensing?

Thanks a lot for you board and your time!