I found this figure in CN0565's circuit note for its resistance measuing capability. However I'm having trouble reproducing this result.

Figure below is my experiment setup, I'm just connecting electrode 0 & 1 to different resistors (10~1MΩ), configuring 500mV and 10kHz, running 'python cn0565_example_single.py 0 1 1 0' and reading the raw magnitute. However I found
- The saturation point is around 10kΩ. (see figure)
- The sensitivity is not that right, I'm reading the same number for 5kΩ and 7kΩ.
- If I turn to 'impedance mode' (other than magnitute mdoe), seems only 1kΩ is correctly measured (maybe because of calibration). Others show around 20% error.


I also found an paper using a similar setup to evaluate this board, obtaining similar result of mine: CN0565 – Multichannel Bipolar Impedancemeter Evaluation | Springer Nature Link.
In fact, I'm trying to use this board to explore human body EIT research (to penetrate through skin and sense subtle impdance change), so I really need to test its sensitivity and penetration. May you please tell:
- Is there anything wrong with my evaluation setup? Since I can always pass the test board (cn0565_prod_test.py), could it be the issue of where AGND is connected to?
- Is this board potentially capable of non-invasive body impedance sensing?
Thanks a lot for you board and your time!
Sherlockk