CN0565‘s accuracy for resistance or impedance measurement

Thread Summary

The user is facing issues with the CN0565 board's resistance measurement capability, including saturation at 10kΩ and poor sensitivity for resistors between 5kΩ and 7kΩ. The support engineer suggests selecting an appropriate RTIA value based on the DUT impedance to improve accuracy and avoid ADC saturation. The user also mentions a connection issue with the ADICUP3029, which may be a common problem, and is now exploring the AD5933 for their EIT research.
AI Generated Content
Category: Datasheet/Specs
Product Number: CN0565

I found this figure in CN0565's circuit note for its resistance measuing capability. However I'm having trouble reproducing this result.

 


Figure below is my experiment setup, I'm just connecting electrode 0 & 1 to different resistors (10~1MΩ), configuring 500mV and 10kHz, running 'python cn0565_example_single.py 0 1 1 0' and reading the raw magnitute. However I found

  1. The saturation point is around 10kΩ. (see figure)
  2. The sensitivity is not that right, I'm reading the same number for 5kΩ and 7kΩ.
  3. If I turn to 'impedance mode' (other than magnitute mdoe), seems only 1kΩ is correctly measured (maybe because of calibration). Others show around 20% error.




I also found an paper using a similar setup to evaluate this board, obtaining similar result of mine: CN0565 – Multichannel Bipolar Impedancemeter Evaluation | Springer Nature Link. 

In fact, I'm trying to use this board to explore human body EIT research (to penetrate through skin and sense subtle impdance change), so I really need to test its sensitivity and penetration. May you please tell:

  1. Is there anything wrong with my evaluation setup? Since I can always pass the test board (cn0565_prod_test.py), could it be the issue of where AGND is connected to?
  2. Is this board potentially capable of non-invasive body impedance sensing?

Thanks a lot for you board and your time!

Sherlockk

  • Hi  ,

    Based on the CN0565 Circuit Note, the accuracy you can achieve in resistance or impedance measurements is strongly linked to the selected RTIA value and the system’s defined performance limits, particularly around its 20 kHz sampling/measurement frequency region.

    The CN0565 uses the AD5940 analog front end, where the RTIA determines the transimpedance gain and ultimately sets the dynamic range and sensitivity of the measurement; using an RTIA that is too small causes early saturation on higher‑value resistances, while too large an RTIA reduces resolution for low‑impedance measurements.

    In the system performance characterization, the note highlights that measurements taken around 20 kHz fall within the optimized bandwidth of the device, ensuring the ADC, excitation signal chain, and RTIA remain within linear operating regions.

    Would you be able to share the results from your resistor sweep so we can compare them directly? Also, what RTIA value are you currently using in your setup? If there were no firmware modifications on your side, I assume the configuration is still using the default 20 kΩ RTIA.

    We’ve also attached our recently captured reference results for your review. The test board setup is purely resistive.

    Best Regards,
    Michelle

  • Hi  ,

    Thanks for your quick reply. 

    I'm afraid my board is now facing connection issue, so I'm not able to share further results.However I'm still curious about this problem so to decide whether I need a new board or not.

    May you please share:

    1. Is it that I should know the approximated resistance in advanced, so to adjust a proper RTIA value in firmware?

    2. I see in your result that the error between 'REAL' and sum of resistors is also around 10% at some measurements. Can it be boosted by some calibration or it's just the best we can do?

    Best Regards,

    Sherlockk

  • Hi  ,

    Could you please let me know if you’re still experiencing any connection issues?

    To answer your questions:
    According to the CN0565 Circuit Note, you need to know the approximate DUT impedance ahead of time so you can choose the appropriate RTIA value in the firmware. The RTIA sets the gain for current‑to‑voltage conversion and selecting the wrong value can lead to problems—high RTIA values can cause ADC saturation for low‑impedance loads, while low RTIA values can reduce measurement resolution for high‑impedance loads.

    How to Select RTIA:

    • Low DUT impedance (≈100 Ω): 200 Ω or 600 Ω RTIA
    • Medium impedance (1 kΩ–10 kΩ): 1 kΩ to 40 kΩ RTIA
    • High impedance (50 kΩ–100 kΩ): 100 kΩ to 600 kΩ RTIA
    • Very high impedance (>100 kΩ): 1 MΩ or higher RTIA

    Rule of thumb: Choose an RTIA value that is roughly the same order of magnitude as the DUT impedance for the best accuracy.

    You can further reduce measurement error by selecting the optimal combination of RTIA value, excitation voltage, and frequency.

    Regards,
    Michelle

  • Hi  ,

    My CN0565 is still failing, referring to some previous Q&As I think my ADICUP3029 is not working, so I'm quiting this line of experiment. May it be some common issue for this set of kit? I see several posts of cn0565 saying ADICUP3029 failing.

    Thanks for your reply anyway. I'm currently working on AD5933 and I am now understanding your answer and calibration things. 

    Best regards,

    Zhikai Qin

  • Closing thread due to inactivity.

    If you still have some questions and require further assistance please open a new one.