Need clarification on CN0359 Configuration

Thread Summary

The user is experiencing incorrect and intermittent conductivity readings with the LFS1305 sensor on the CN0359 Rev. C board, expecting μS/cm but getting mS/cm. The final answer suggests removing JP1, JP2, and JP3, and soldering JP4 and JP5, but the user reports no improvement. The issue may be related to the 4-electrode configuration and capacitive coupling changes in Rev C. The user also notes that the excitation voltage setting is for Vp, not Vpp, and should be set to 0.35V for a 0.7Vpp excitation.
AI Generated Content
Category: Hardware
Product Number: CN0359

Hi, I am using the CN0359 Rev. C Conductivity Measurement System. I am using it with this conductivity sensor from IST: LFS1305

The sensor requires the following: 

  • Excitation voltage: <0.7Vpp
  • Excitation frequency: 100Hz to 10kHz
  • Cell constant = 0.86/cm

Based on the medium I am using, my readings should be in the range of μS/cm, but instead I am in the mS/cm range. I do occasionally get the right reading, but it is intermittent. 

My questions are: 

  1. When setting the excitation voltage, is this the Vp or the Vpp? For example, if i set it to 0.7V, does that mean the Vpp is 0.7V or 0.35V?
  2. What could be the possible cause of my intermittent values? 
    1. For context, when I do get good readings the excitation voltage is set to 0.6V, frequency is 820Hz, and setup time is around 8% and hold time around 5%.  

Thanks!

Parents
  •  , I am still getting the same results. Did these changes work for you?

  •  Hello , I received your sensors yesterday (LFS1505, LFS1305, and LFS1107) and tested them. At first, I also got a “sensor error” but after dipping the sensor probes into water, I started getting stable readings. The values were very close to those from my conductivity meter.

    The reason you're seeing the “sensor error” message is that the reading is outside the acceptable range, which is expected when the probe is not in a liquid solution. Try dipping the probe into water to get a valid reading.

    -Rey

  • Hi  , 

    I am no longer getting a "sensor error" message, but the readings are not close to the expected values. I used another conductivity meter, the water I am using is within the uS/cm range, while both boards are showing readings within the mS/cm range. Additionally, when I was getting the "sensor error" messages in the past, it was while the probe was immersed inside calibration solution within the range the sensor is rated for. 

    Were you able to get good readings of other fluids? 

    Can you share the configuration of the board you used? 

Reply
  • Hi  , 

    I am no longer getting a "sensor error" message, but the readings are not close to the expected values. I used another conductivity meter, the water I am using is within the uS/cm range, while both boards are showing readings within the mS/cm range. Additionally, when I was getting the "sensor error" messages in the past, it was while the probe was immersed inside calibration solution within the range the sensor is rated for. 

    Were you able to get good readings of other fluids? 

    Can you share the configuration of the board you used? 