Post Go back to editing

Simultaneous measurement of pH and ORP

Hello, I use CN0326 to measure pH (mV), the result is very good, but I want to measure and add ORP parameters on the same circuit at the same time. i changed Opamp AD8606 (1 channel) to AD8607 (2 channel). Then connect two pH and ORP sensors simultaneously. however i noticed the result when i connected 2 different pH + ORP sensors at the same time than the pH single connection (pH voltage deviation +-5- 10mV). Please explain this phenomenon to me
Thanks a lot

Parents
  • Hi thuanbkhn,

    Can you provide a schematic showing exactly how you have connected the additional channel? A high resolution photograph would also be very helpful.

    Note that because of the high impedances involved, all circuitry must be kept very clean. Did you modify the circuit by hand? Is it possible that there is some flux residue or other contamination on the circuit?

    -Mark

Reply
  • Hi thuanbkhn,

    Can you provide a schematic showing exactly how you have connected the additional channel? A high resolution photograph would also be very helpful.

    Note that because of the high impedances involved, all circuitry must be kept very clean. Did you modify the circuit by hand? Is it possible that there is some flux residue or other contamination on the circuit?

    -Mark