I tried to use the dc388c evaluation board to stabilize my laser system. the board is not working properly as my temperature gets continuously oscillating for a long long time. can anybody suggest what to do in this case or am I making some mistake. I'm also uploading the variation of my thermistor resistance with time. I'm using %V supply and also my current is working at two extreme ends always while cooling(320mA) and heating(280mA). with time it's also not decreasing
    Thanks for contacting Analog Devices!
    Thanks for sharing the waveform. This is actually a very classic instability issue in TEC (thermoelectric controller) loops, especially with the DC388C.
    What your waveform is showing?
    From your image:
    • Thermistor value oscillates between roughly 8.6kΩ and 8.8kΩ
    • Oscillation is slow and periodic
    • It continues for a very long time (15+ minutes shown)
    • Current is hitting:
      • ~320 mA (cooling max)
      • ~280 mA (heating max)
    -> This clearly indicates:
    The control loop is saturating and oscillating (limit cycling)
    Root cause (very important insight)
    This is NOT a hardware failure.
    This is almost certainly:
    Loop compensation / tuning issue
    Your system is behaving like:
    • High loop gain
    • Very slow thermal system
    • Aggressive controller
    Result:
    • Controller overshoots temperature
    • Corrects too aggressively
    • Overshoots again
    • Repeats → oscillation
    Why this happens in your setup
    The DC388C is designed as a generic platform, but:
    -> Your laser + TEC + thermistor system is custom
    -> Thermal dynamics are very slow and nonlinear
    Mismatch occurs between:
    • Controller compensation
    • Actual thermal system response
    Key symptom confirmation
    These two observations confirm instability:
    1. Current always at extremes
    • Either max heating or max cooling
    • Never settles in mid-range
    This means:
    Controller has no fine control → behaving like bang-bang
    1. Oscillation does not decay
    • Amplitude not reducing with time
    This means:
    Loop is underdamped or unstable
    What you should do (actionable steps)
    Step 1: Reduce loop gain (most important)
    You must slow down the control loop
    Try:
    • Increase compensation capacitor (I-term cap)
    • Reduce gain resistor
    -> Effect:
    • Slower response
    • Less overshoot
    • Oscillation reduces
    Step 2: Increase integration time constant
    For thermal systems:
    • System is inherently slow
    • Controller should be even slower
    -> If controller is faster than thermal system → oscillation