MAX17320 plethora of issues. Charge and discharge termination lead to OVP and UVP, FETs do not get deactivated, poor documentation
The user is experiencing issues with the MAX17320 where the CHG and DIS FETs do not turn off as expected, leading to overcharging and under-discharging of a 3S battery pack. The user provided an INI file and detailed observations, including cell imbalance and incorrect SOC reporting. The final answer suggests opening a support ticket with the Central Apps Team for a deeper review, and mentions that the MAX17320 requires a charge source to wake up the first time batteries are connected.
Product Number: MAX17320
Dear community
We are working on a custom evaluation board for the MAX17320, entirely based on the MAX17320G1EVKIT board. Our system uses a 3S pack of 2600 mAh batteries and we have used the exact same component values from the EVKit.
After several weeks of studying and trying to decipher the poorly-written datasheet from the chip and additional resources regarding the ModelGauge m5 algorithm, we managed to get a good number of settings for our particular case working with the help of the Wizard. In spite of the lack of a friendly C/C++ library, we managed to workout some logic to communicate with the device and we are now able to read every register from the system and decode the most relevant bits and values.
Amongst the several issues we encountered during the past weeks, we are now facing an extremely pressing issue: the FETs gate drivers CHG and DIS never go off. We have set the drive voltage to 10V and we ALWAYS read 9.8V between the Gate and Source from both Charge and Discharge FETs. This represents potential risks and serious issues for our application. On the one hand, when the batteries are fully charged (according to the SOC at 100% and cell voltages at 4.2V (+/-0.05V)) the battery pack keeps taking amps. The current reading even goes below the charge termination current of 80 mA that we have successfully written to shadow RAM. No matter how long we leave the charger connected, the voltages per cell keep increasing to the point of reaching nearly 4.3V and above (we have even registered a cell going near 4.5V!)
Please clarify if, under the described conditions, the CHG FET should go OFF. We have even computed a small snippet to determine if the End-of-Charge Detection conditions are present (datasheet page 58) and all three conditions are met, yet the CHG FET stays on!
Similarly, when fully depleting our battery pack, the DIS FET is NEVER turned OFF despite the cells reach the desired discharge voltage of 3.0V. If a cell dips below our desired UVP threshold of 2.8V, the alter kicks-in and the battery pack simply dies.
The issues described above are a small sample of many other problems we are facing with this chip. The description of several registers is lacking clarity, there are several access bits that do not have any explanation whatsoever, the math used to encode floating-point values to hexadecimal is not clearly explained thus creating much more frustration and confusion.
It is extremely disappointing (and rather discouraging) to face this situation with a chip from a reputable company like ADI/Maxim. The documentation is extremely poor, lacks consistency, makes a whole bunch of assumptions, and leaves the reader (designer) helpless.
On top of this, the forum is filled with questions with no answers. WE NEED ANSWERS, please.
In our case, our project is the prototype of a product that could potentially hit the market in the hundreds of thousands of units per month. However, in light of the high unreliability of this chip, we cannot fully decide whether to continue the development using the MAX17320 or moving to a Texas Instruments one to avoid losing a valuable customer. ADI/Maxim SHOULD care about their customers
EDIT 1:
As an example of how bad the documentation is, how is it possible that one user had to figure out by themself that the I2C address had to be changed for writing to shadow RAM?
MAX17320 Receiving error on nonvolatile memory write
This is totally unacceptable
EDIT 2:
More examples of poor documentation. In the official datasheet for the MAX17320 there is no mention whatsoever of any register for checking if Balacing is active or Off. I just found out that there is an "obscure" HConfig2 register (0x0F5)
 How to check the status of cell balancing in MAX17320?
Why is this relevant piece of information not included in the datasheet?
Edit Notes
Included more examples of missing information in the datasheet
[edited by: dmelendrez at 5:48 PM (GMT -5) on 11 Nov 2025]
  • Perhaps  may be willing to assist with this?
    I would deeply appreciate some insights
  • Hello,
    The behavior described above is definitely not the expected behavior.  The OVP and UVP are configurable values and the CHGFET should open when the cell voltage of any cell exceeds the OVP threshold and the DISFET should open when the cell voltage of any cell drops below the UVP threshold.  If you can provide the Complete INI file that should be generated by the EVKIT GUI, we are happy to review the settings to see if anything is not configured properly.
    Regarding the charging behavior, the MAX17320 will not limit the charge and stop it once the termination current is reached.  The external charger should be used to limit the charging and stop it at the appropriate time.  The MAX17320 will open the CHG FET if the OVP threshold is exceeded.  When the current tapers to the configured IChgTerm value, the Fuel gauge will report 100%, but it doesn't stop the charging.
    Thanks
  • Dear  thank you for mentioning
    Dear  thanks for your answer, and apologies for the late reply.
    For context, we are using a custom 3S battery pack (2500 mAh nominal). Our MAX17320 is a custom board and is controlled by an ESP32-S2 via I2C. We use the EVkit Wizard as an INI file generator, which we manually load using our custom firmware.
    Here is the latest INI file we have been working with for your consideration. Below, I will describe the most recent issues we have been facing for the past weeks or so.
    Device = MAX173xx
    Title = EVKIT Configurator Profile Generated on 2025/11/06 16:38:37
    0x180 = 0x0000		//nXTable0 Register
    0x181 = 0x0000		//nXTable1 Register
    0x182 = 0x0000		//nXTable2 Register
    0x183 = 0x0000		//nXTable3 Register
    0x184 = 0x0000		//nXTable4 Register
    0x185 = 0x0000		//nXTable5 Register
    0x186 = 0x0000		//nXTable6 Register
    0x187 = 0x0000		//nXTable7 Register
    0x188 = 0x0000		//nXTable8 Register
    0x189 = 0x0000		//nXTable9 Register
    0x18A = 0x0000		//nXTable10 Register
    0x18B = 0x0000		//nXTable11 Register
    0x18C = 0x0000		//nVAlrtTh Register
    0x18D = 0x0000		//nTAlrtTh Register
    0x18E = 0x0000		//nIAlrtTh Register
    0x18F = 0x0000		//nSAlrtTh Register
    0x190 = 0x0000		//nOCVTable0 Register
    0x191 = 0x0000		//nOCVTable1 Register
    0x192 = 0x0000		//nOCVTable2 Register
    0x193 = 0x0000		//nOCVTable3 Register
    0x194 = 0x0000		//nOCVTable4 Register
    0x195 = 0x0000		//nOCVTable5 Register
    0x196 = 0x0000		//nOCVTable6 Register
    0x197 = 0x0000		//nOCVTable7 Register
    0x198 = 0x0000		//nOCVTable8 Register
    0x199 = 0x0000		//nOCVTable9 Register
    0x19A = 0x0000		//nOCVTable10 Register
    0x19B = 0x0000		//nOCVTable11 Register
    0x19C = 0x0100		//nIChgTerm Register
    0x19D = 0x0000		//nFilterCfg Register
    0x19E = 0x0000		//nVEmpty Register
    0x19F = 0x0000		//nLearnCfg Register
    0x1A0 = 0x1050		//nQRTable00 Register
    0x1A1 = 0x8002		//nQRTable10 Register
    0x1A2 = 0x078C		//nQRTable20 Register
    0x1A3 = 0x0880		//nQRTable30 Register
    0x1A4 = 0x0000		//nCycles Register
    0x1A5 = 0x0A6C		//nFullCapNom Register
    0x1A6 = 0x08CC		//nRComp0 Register
    0x1A7 = 0x223E		//nTempCo Register
    0x1A8 = 0x0000		//nBattStatus Register
    0x1A9 = 0x08FC		//nFullCapRep Register
    0x1AA = 0x0000		//ndQTot Register
    0x1AB = 0x0000		//nMaxMinCurr Register
    0x1AC = 0x0000		//nMaxMinVolt Register
    0x1AD = 0x0000		//nMaxMinTemp Register
    0x1AE = 0x0000		//nFaultLog Register
    0x1AF = 0x0000		//nTimerH Register
    0x1B0 = 0xB695		//nCONFIG Register
    0x1B1 = 0x0204		//nRippleCfg Register
    0x1B2 = 0x0000		//nMiscCFG Register
    0x1B3 = 0x08FC		//nDesignCap Register
    0x1B4 = 0x0008		//nSBSCFG Register
    0x1B5 = 0x4205		//nPACKCFG Register
    0x1B6 = 0x083B		//nRelaxCFG Register
    0x1B7 = 0x2241		//nConvgCFG Register
    0x1B8 = 0x0AC0		//nNVCFG0 Register
    0x1B9 = 0x0102		//nNVCFG1 Register
    0x1BA = 0xBE2D		//nNVCFG2 Register
    0x1BB = 0x0909		//nHibCFG Register
    0x1BC = 0x0000		//nROMID0 Register
    0x1BD = 0x0000		//nROMID1 Register
    0x1BE = 0x0000		//nROMID2 Register
    0x1BF = 0x0000		//nROMID3 Register
    0x1C0 = 0x0000		//nChgCtrl1 Register
    0x1C1 = 0x0000		//nPReserved1 Register
    0x1C2 = 0x2061		//nChgCfg0 Register
    0x1C3 = 0x00E1		//nChgCtrl0 Register
    0x1C4 = 0x0000		//nRGain Register
    0x1C5 = 0x0000		//nPackResistance Register
    0x1C6 = 0x0000		//nFullSOCThr Register
    0x1C7 = 0x0000		//nTTFCFG Register
    0x1C8 = 0x4000		//nCGAIN Register
    0x1C9 = 0x0001		//nCGTempCo Register
    0x1CA = 0x7135		//nThermCfg Register
    0x1CB = 0x0000		//nChgCfg1 Register
    0x1CC = 0x0000		//nManfctrName Register
    0x1CD = 0x0000		//nManfctrName1 Register
    0x1CE = 0x0000		//nManfctrName2 Register
    0x1CF = 0x01F4		//nRSense Register
    0x1D0 = 0x784F		//nUVPrtTh Register
    0x1D1 = 0x3700		//nTPrtTh1 Register
    0x1D2 = 0x5528		//nTPrtTh3 Register
    0x1D3 = 0x21E7		//nIPrtTh1 Register
    0x1D4 = 0x2C60		//nBALTh Register
    0x1D5 = 0x2D05		//nTPrtTh2 Register
    0x1D6 = 0x2A58		//nProtMiscTh Register
    0x1D7 = 0x0D00		//nProtCfg Register
    0x1D8 = 0x3243		//nJEITAC Register
    0x1D9 = 0xF659		//nJEITAV Register
    0x1DA = 0x34D3		//nOVPrtTh Register
    0x1DB = 0x61F5		//nStepChg Register
    0x1DC = 0xAFB5		//nDelayCfg Register
    0x1DD = 0x4F99		//nODSCTh Register
    0x1DE = 0x4355		//nODSCCfg Register
    0x1DF = 0xAFAB		//nProtCfg2 Register
    0x1E0 = 0x0000		//nDPLimit Register
    0x1E1 = 0x0000		//nScOcvLim Register
    0x1E2 = 0x0000		//nAgeFcCfg Register
    0x1E3 = 0xA5B9		//nDesignVoltage Register
    0x1E4 = 0x0000		//nVGain Register
    0x1E5 = 0x0000		//nRFastVShdn Register
    0x1E6 = 0x0000		//nManfctrDate Register
    0x1E7 = 0x0000		//nFirstUsed Register
    0x1E8 = 0x0000		//nSerialNumber0 Register
    0x1E9 = 0x0000		//nSerialNumber1 Register
    0x1EA = 0x0000		//nSerialNumber2 Register
    0x1EB = 0x0000		//nDeviceName0 Register
    0x1EC = 0x0000		//nDeviceName1 Register
    0x1ED = 0x0000		//nDeviceName2 Register
    0x1EE = 0x0000		//nDeviceName3 Register
    0x1EF = 0x0000		//nDeviceName4 Register
    1. Regarding the discharging configuration (Step 8/21 in the Wizard); we set an empty voltage of 3V and a UVP of 2.8V. However, when discharging the pack, both the alert and protection kick in at 3V, and the MAX17320 disables the DIS FET, which "kills" the pack output.
     Discharging configuration
    2. With regards to cell balancing, we just recently faced a major issue with one of our battery packs, where one of the cells deviated from the others as far as ~200mV. This resulted in the MAX17320 reporting a wrong Time to Empty (TTE) value, and the SOC was stuck at 60% despite the cells' voltages being around 4,2, 4.18 and 4.0V. At the time of typing this, one of the cells (using a different set of Li-Ion batteries) is reporting a spread of ~600mV.
    We have monitored the thermals of our custom board, and we have noticed that the balancing resistors get pretty hot (despite using the same value and rating from the EVKit schematic). At some point, one of the resistors (the one attached to BT2 - the "offending" lowest cell) showed some sort of "thermal flickering" or "pulsating" behaviour, which can be evidently related to variations in the current passing through this resistor. I will assume that this is a result of the internal FETs trying to balance the cells; however, there seems to be something wrong with them.
    It is important to mention that, at first, we thought that our battery holders could be the issue, since we were using spring-loaded type 18650 battery holders. We assumed that these could potentially introduce some stray resistance between cell nodes. However, we just changed our approach, and we are spot-welding our cells.
    Thermal profile while chargingBalancing resistors area
    3. This very morning, after flashing the shadow RAM with the INI settings, we proceeded to commit these to the NVM (we now have only 5 remaining NVM commits). We attached a fresh set of batteries, and after pressing the "Wakeup" button, the battery would NOT turn on the "pack voltage" at the output terminals. The only way to make it come to life was by keeping the button pressed WHILE plugging in the charger input. This is NOT a very desirable behaviour. Imagine having to tell this to our customers?
    4. Similarly, at the time of writing this, my colleague just noticed something VERY wrong with the SOC. Despite the battery pack measuring a voltage of ~12.4V, the SoC reports only 17.2%. Why?! We are also showing all the voltages per cell here. Please note that these images don't show the spot-welded version of the battery pack. We had to temporarily go back to the cell holders to continue our tests
    Wrong SoC  Cell 2Cell 3
    These are the most pressing issues at the moment. We will DEFINITELY come back with more issues since this is still a work in progress.
    Thanks for your help in advance
    EDIT:
    The last picture of the battery cells' voltages is wrong. For some reason, I uploaded one of the pictures showing the wrong voltage. I accidentally took my shot while my colleague was moving the test leads. The last cell voltage was close to the one reported in the serial terminal
  • The MAX17320 is very good at keep cells balanced.  However, it is HIGHLY recommended that the cells start off well balanced, and the MAX17320 can try to keep them balance.  But if they are starting off with a very large delta, it may be more than the MAX17320 can help with.  What typically will happen is one cell will reach full first and limit how full the pack is charged.  It may also trip the OVP fault if that single cell is too much higher than the rest.  Or a single cell may reach UVP first, and limit how much can be discharged out of the pack before tripping UVP.
    The MAX17320 measures each cell individually for UVP and OVP protection.  The fuel gauge also uses the higher or lower cells when estimating the SOC.  So if there is a big imbalance, this can impact the fuel gauge behavior.
    For best results, I would recommend balancing the cells individually before assembling into a pack.
    The MAX17320 does require a charge source connection to wake it up the first time the batteries are connected to a pack.
  • Dear
    Thanks for your answer.
    Here are some very much needed clarifications to your answers.
    "The MAX17320 is very good at keep cells balanced."
    Unfortunately, we have to strongly disagree with this statement. We are still actively experiencing huge imbalances between cells. We need to clarify that we are effectively using pre-balanced cells (using an external charger) where all cells measure 4.15V +/- 0.05V (More of this later)
    "However, it is HIGHLY recommended that the cells start off well balanced, and the MAX17320 can try to keep them balance."
    We have been totally aware of this since the beginning of this project. We have now tested 15 well-matched cells (binning the closest ones in terms of voltage and freshly installed after charging finishes), yet we are still facing these issues and triggering all sorts of alerts and protections.
    "But if they are starting off with a very large delta, it may be more than the MAX17320 can help with.  What typically will happen is one cell will reach full first and limit how full the pack is charged.  It may also trip the OVP fault if that single cell is too much higher than the rest.  Or a single cell may reach UVP first, and limit how much can be discharged out of the pack before tripping UVP. "
    We fully agree with this, and it is (partially) the behaviour that we have observed on several occasions. The main drawback here is that we have to programmatically remove the fault, which complicates the control algorithm on our system. We are dealing with a consumer-grade device, and the end-user should not have to deal with any issues arising from the fuel gauge itself.
    "The MAX17320 measures each cell individually for UVP and OVP protection.  The fuel gauge also uses the higher or lower cells when estimating the SOC.  So if there is a big imbalance, this can impact the fuel gauge behavior."
    While this statement makes a lot of sense "on paper", it is far from reality. We have provided evidence in our last response where the SOC was reported at 17.2% while the worst-case cell was merely at 4.095V. There is no way that this makes any mathematical sense.
    "For best results, I would recommend balancing the cells individually before assembling into a pack."
    While this is totally understandable, we ARE indeed assembling packs with well-balanced batteries (as mentioned before). However, once we cycle the battery pack with any load, the cells start spreading apart.
    For instance, here is a clear example of what I am trying to explain (part of it, at least). We observed this imbalance WHILE charging a freshly pre-charged pack (all cells to 4.16V +/-0.05V). Cell 2 suddenly started deviating from the rest of them. We first discharged them for a while, reaching 9V per pack (approx. 3.15V/ per after bouncing back with a worse deviation of 300mV - unnaceptable but we carried on)
    Cell 1 - ImbalanceCell 2 - Imbalance while chargingCell 3 - Imbalance
    Regardless of the observed imbalance, we continued charging this same pack. One of our cells reached 4.26V (!!) and, evidently, OVP got triggered.
    OVP - Huge imbalance
    This cannot be any clearer.
    What is the point of setting a charging voltage to 4.15V (at room temperature - we are at 26C here in the shop), an OVP at 4.28V and an OVP release at 4.26V in step 5 if the MAX17320 will not try to keep the cells at such levels? It seems to us as if the MAX17320 was simply ignoring the settings it has in the NVM.
    Step 5
    Please note that our PCB is basically a "carbon copy" of the Evaluation board provided for this chip, and thus, we are using the same values and ratings from the provided BOM.
    "The MAX17320 does require a charge source connection to wake it up the first time the batteries are connected to a pack."
    Based on the hardware we have on the bench (two custom MAX17320 boards), this is not entirely true. One of the boards showed the behaviour that you describe here (as we did before, where we had to connect the charger WHILE pressing the wakeup button). On the other hand, we did NOT need to do this with the second board. Attaching a pack and pressing the wakeup button was enough to wake it up. Both systems are running the same NVM configuration from the same INI file we shared before.
    We are exploring all possibilities to get this system up and running as intended. We have invested time, effort and quite a lot of money in prototyping these systems. We want to extinguish all chances of changing our design and delve into a different chip. If we do not see any other option, we will have to migrate our design to a TI chip or similar. Our system will invariably hit the market in the thousands of units per month, and we originally thought that the MAX17320 was our best bet.
    We are now questioning our choices.
  • I think this may be getting much more in depth that we can continue on the EngineerZone page.  The best course of action here is to redirect you to the Select Support Type  · Support Portal to open a support ticket and get connected to someone on our Central Apps Team to continue this discussion.
  • We thought that we were already dealing with Technical Support. Oh, well...
    So, open a ticket and start over again. Very nice.
    We need to get to the bottom of this.
    I stand by my initial claims. Maxim/ADI are leaving us customers alone in the dark.
    In any case, thanks for your help,
  • Hi
    Thank you for the detailed follow-up and for clearly outlining your concerns. We appreciate the time and effort you have invested in investigating and documenting these issues.
    EngineerZone is a technical support community, and your questions and feedback are appropriate for this forum. In some cases, however, the depth and system-specific nature of a problem requires engagement through our Central Apps Team (CAC). This ensures the issue can be reviewed in greater depth with the appropriate applications specialists.
    Please open a CAC support ticket, and once it has been created, post the ticket number here. I will reference this EngineerZone thread, so the support team has full context, and you are not required to restate the information you have already provided.
    Thank you for your patience as we work to ensure the appropriate technical resources are engaged.