What is "the offset of the internal op amp"

Category: Datasheet/Specs
Product Number: LTC2944

Hi all.

The LTC2944 datasheet says the following:

For input signals with an absolute value smaller than the
offset of the internal op amp the LTC2944 stops integrating
and does not integrate its own offset

Is this equivalent to Vose in the electrical characteristics (5uV typ/10uV max)? I don't see any other offset specification relating to the coulomb counter.

Regards,

Mike

Thread Notes

  • Any help here?

    I tried to test it by drawing a small constant current and observing how fast ACR changes over the course of ~10 minutes. My Rsense value is 1mohm.Prescaler was 256, so each LSB of the ACR is 1.0625mAhr.

    Actual current (mA) ACR rate of change (LSB/hr) Measured current (mA)
    20.0 0 0.0
    40.0 9.1 9.7
    60.0 37.6 39.9
    80.0 61.3 65.2

    So there is apparently a level of current that the LTC2944 cannot detect. At 20mA, the ACR had not changed even after a full hour. Based on the datasheet Vose I would expect it to respond to currents above 10mA, but apparently that's not the case.

    And even when the current is high enough for the ACR to change, it has far greater error than I would expect compared to the datasheet Vose value. But I see that the Vose spec is for the condition Vsense > 500uV, and the plot for "Total Charge Error vs Differential Sense Voltage" also only extends down to Vsense=500uV. Could it be that Vose is actually much greater at lower sense voltages?