Requests for Interpreting Output Results
I am testing emstat pico with STM32f4 controller. Based on the LSV measurements, I got these results. So what does this result mean in terms of diagnosing problem?
const char * Linear_Sweep = "e\n"
							 "var c\n"
							 "var p\n"
							 // Set hardware configuration to high speed mode.
							 "set_pgstat_mode 3\n"
							 // Set maximum bandwith of the signal being measured. Any signal of significant higher frequency than the set bandwidth will be filtered out.
							 // Set the maximum bandwidth to frequency of 200 Hz.
								 "set_max_bandwidth 200\n"
							 // Selects the lowest current range that can measure the 'max value' current without causing an overload. That is 500uA.
							 "set_range ba 100u\n"
							 // Set WE potential to -500mV
							 "set_e -500m\n"
							 // Turn the cell on. The instrument will start applying the configure potential current.
							 "cell_on\n"
							// Wait 1 second
							 "wait 100m\n"
							 // The LSV performs a potential sweep from -500 mV to 500 mV with steps of 50
							 //mV at a rate of 100 mV/s
							 "meas_loop_lsv p c -500m 500m 50m 500m\n"
							// Start measurement data packet.
							 "pck_start\n"
							 // Add variable p to the measurement data package
							 "pck_add p\n"
							 // Add variable c to the measurement data package
							 "pck_add c\n"
							 // Signal the end of a measurement package
							 "pck_end\n"
							 "endloop\n"
		 	 	 	 	 	 "on_finished:\n"
							 "cell_off\n"
							 "\n";
e
Failed to parse package: -5
Pda7F85F3Fu;ba955CB9Cf,14,27
Receiving measurement response:1	E set[V]: -4.999E-01	I[A]: 2.240E-08	status: Underload        	CR: 125 uA
2	E set[V]: -4.497E-01	I[A]: 2.240E-08	status: Underload        	CR: 200 uA (High speed)
3	E set[V]: -3.995E-01	I[A]: 1.120E-08	status: Underload        	CR: 200 uA (High speed)
4	E set[V]: -3.493E-01	I[A]: 1.120E-08	status: Underload        	CR: 200 uA (High speed)
5	E set[V]: -2.992E-01	I[A]: 1.120E-08	status: Underload        	CR: 200 uA (High speed)
6	E set[V]: -2.490E-01	I[A]: 1.120E-08	status: Underload        	CR: 200 uA (High speed)
7	E set[V]: -1.988E-01	I[A]: 1.120E-08	status: Underload        	CR: 200 uA (High speed)
8	E set[V]: -1.486E-01	I[A]: 1.120E-08	status: Underload        	CR: 200 uA (High speed)
9	E set[V]: -9.840E-02	I[A]: 1.120E-08	status: Underload        	CR: 200 uA (High speed)
10	E set[V]: -4.821E-02	I[A]: 1.120E-08	status: Underload        	CR: 200 uA (High speed)
11	E set[V]: 1.976E-03Pd4	I[A]: 1.120E-08	status: Underload        	CR: 200 uA (High speed)
12	E set[V]: 1.976E-03	I[A]: 1.120E-08	status: Underload        	CR: 200 uA (High speed)
13	I[A]: 1.120E-08	status: Underload        	CR: 200 uA (High speed)
14	E set[V]: 1.525E-01	I[A]: 1.120E-08	status: Underload        	CR: 200 uA (High speed)
15	E set[V]: 2.027E-01	I[A]: 1.120E-08	status: Underload        	CR: 200 uA (High speed)
16	E set[V]: 2.529E-01	I[A]: 1.120E-08	status: Underload        	CR: 200 uA (High speed)
17	E set[V]: 3.031E-01	I[A]: 1.120E-08	status: Underload        	CR: 200 uA (High speed)
18	E set[V]: 3.533E-01	I[A]: 1.120E-08	status: Underload        	CR: 200 uA (High speed)
19	E set[V]: 4.035E-01	I[A]: 1.120E-08	status: Underload        	CR: 200 uA (High speed)
20	E set[V]: 4.537E-01	I[A]: 1.120E-08	status: Underload        	CR: 200 uA (High speed)
21	E set[V]: 5.039E-01	I[A]: 1.120E-08	status: Underload        	CR: 200 uA (High speed)
22	E set[V]: 5.039E-01	I[A]: 1.120E-08	status: Underload        	CR: 200 uA (High speed)
23	E set[V]: 5.039E-01	I[A]: 1.120E-08	status: Underload        	CR: 200 uA (High speed)
24	E set[V]: 5.039E-01	I[A]: 1.120E-08	status: Underload        	CR: 200 uA (High speed)
25	E set[V]: 5.039E-01	I[A]: 1.120E-08	status: Underload        	CR: 200 uA (High speed)
26	E set[V]: 5.039E-01	I[A]: 1.120E-08	status: Underload        	CR: 200 uA (High speed)
27	E set[V]: 5.039E-01	I[A]: 1.120E-08	status: Underload        	CR: 200 uA (High speed)
28	E set[V]: 5.039E-01	I[A]: 1.120E-08	status: Underload        	CR: 200 uA (High speed)
29	E set[V]: 5.039E-01	I[A]: 1.120E-08	status: Underload        	CR: 200 uA (High speed)
30	E set[V]: 5.039E-01	I[A]: 1.120E-08	status: Underload        	CR: 200 uA (High speed)