ADPD4100/ADPD4101 Medical grade

Is ADPD4100/ADPD4101 belonging to Medical grade IC? 

  • 0
    •  Analog Employees 
    on Jul 15, 2021 6:45 PM

    Hi John,

    Although we don't market ADPD4100/4101 as a Medical grade part, we have performed rigorous tests of related medical applications of the ADPD4100/4101 part: 

    The ADPD4100/4101 ECG is based on current measurement and is different from traditional ECG solution which is typically an instrumentation amplifier. The ADPD4100/4101 ECG has been tested against IEC60601-2-27 and IEC60601-2-47 and it meets requirements by the standards.

    Along with ADI proprietary algorithms, systems built with ADPD4100/4101 for heart rate monitoring and SpO2 based on PPG gave very promising results in clinical trials.

    I hope the information above helps.

    Regards,

    Glen B.