The MAX86150 is an integrated electrocardiogram, pulse oximeter, heart rate monitor sensor module. It includes internal LEDs, photodetector, and low-noise...
Datasheet
MAX86150 on Analog.com
I am recording data with MAX86150, the sample rate is set to 200sps for ECG and PPG. but when I record the data for each IC the sample rate is a number near 200 and it is not accurate.
this is the result:
recording time = 24:29:06 (HH:MM:SS) = 88146s
the desired samples based on the recording time = 88146 x 200 = 17629200 samples
but total samples generated by MAX86150 = 17645525 samples
real sample rate = 17,645,525 / 88,146 = 200.185 sample per second
this test with another MAX86150 IC results 199.45sps.
Has anyone else faced this problem? and is there any solution to solve this problem?
stephenv - Moved from Other Products (EN) to Optical Sensing Library. Post date updated from Saturday, August 3, 2024 11:35 PM UTC to Tuesday, February 3, 2026 9:08 PM UTC to reflect the move.
Based on the datasheet it can vary between 196 to 204 samples per second!
![]()