AD9467 FMC Start Capturing data and see the output waveform in ADI IIO Oscilloscope

Category: Software
Product Number: AD9467_Eval_board
Software Version: no-OS (master branch)

Hi  

Thank you for your recommendation. At this point, we are happy with the project's success because it has been made possible by your excellent suggestions that have helped us advance to new levels on a daily basis. The way you help in this thread was precious  RE: Generates error in No-OS (2019_R2) using SDK 2019.1 

You suggested the iio-oscilloscope, which is described here: wiki.analog.com/.../tinyiiod demo
With the help of this link's instructions, we were able to create a project utilizing the TINYIIOD=y flag and attach it to a iio-oscilloscope. Here is the snapshot
 

Here, the connection was successfully established, and then the ADI IIO Oscilloscope window appeared.

Now, we'll go over three scenarios with you in which we use an input signal from Analog Discovery 2 and an ADI IIO Oscilloscope to view the ad9467's output voltage signal.

Scenario:1 You can examine the same sinusoidal signal on the oscilloscope during run time by providing sinusoidal input from Analog discovery 2. The image is shown below. Waveform signal (Running), IIO Oscilloscope (Running)
 

Scenario:2 You can examine the same sinusoidal signal on the oscilloscope during run time by providing triangle wave input from Analog discovery 2. Here's a picture, Waveform signal (Running), IIO Oscilloscope (Running)

In scenario 2, is it required to launch or debug the project in order to see the various waveforms in the oscilloscope? By the way, even after a relaunch and another round of code-debugging, the sinusoidal wave remains. Can you provide me some ideas on how we can view various waveform signals in the oscilloscope while it is in run-time?

Scenario:3 When we stop the Analog Discovery 2 input signal during operation, there is no signal at first and a sinusoidal wave arises at the ending. This is the picture. Waveform signal (STOP), IIO Oscilloscope (Running)

Here are the issues:

What are the methods, concerns, and things to keep in mind to obtain the same run time environment if the ADI IIO Oscilloscope plots the data continuously in run time as you instructed? Does the IIO Oscilloscope adjust appropriately if we change the input signal from Analog Discovery 2 or not? If so, what should we do?

Regards


Abbas Ali
FPGA Engineer

Parents Reply Children
  • Hi,

    From what I see in your previous thread.
    Your sampling clock is 20MHz. According to the datasheet the minimum sampling clock should be 50MHz.
    Your clock source is one of the the Analog Discovery 2 outputs, right? It's DAC is capable of 100MSPS, so, 50M is doable, but keep in mind that this setup is not exactly reliable. maybe not even as a proof of concept.
    My suggestion is to use use ad9467 onboard clock and filter out samples, if bandwidth is your concern. Or use a more reliable clock source.

    Andrei

  • Hi  and  ,

    Your suggestion is valid to use the ad9467 onboard clock frequency. Now, we will work on it to use the onboard clock frequency range between 50Mhz-250Mhz. After that, we will update you soon. 

    If we want to look at the DDR data, can anyone tell me where the DDR data gets stored and in which format 16 or 32-bit?

  • ,

    You already know the address, ADC_DDR_BASEADDR:

    https://github.com/analogdevicesinc/no-OS/blob/master/projects/ad9467/src/devices/adi_hal/parameters.h#L52

    Data is stored in 16-bit.

    And as I said in the other thread, you may use the xilinx_capture.tcl script in no-OS repo to retrieve the data.

  • Hi  and  ,

    We apologize for the delayed response; there are some problems on our end.
    Although you advised using the ad9467 onboard clock frequency, you missed providing a link describing how to modify the configuration of the ad9467 onboard clock; instead, we used this source link.

    https://wiki.analog.com/resources/fpga/xilinx/fmc/ad9467 



    However, we get quite unpredictable results in Vitis Serial Terminal.

    When we provide input from the ad9467_fmc.c using this function 

    • ad9517_frequency(ad9517_device, 3, 150000000);
    • ad9517_frequency(ad9517_device, 3, 250000000);

    With the given inputs of 150 Mhz or 250 Mhz, we are getting the clock frequency in the region of 170-180 Mhz in Vitis Serial Terminal, why do these unpredictable results occur?

    Please take a look at it to see why providing the clock input from the internal source code does not produce the desired results on the Vitis Serial Terminal or in case it's not the issue then provide another source that how to configure the ad9467 onboard clock.

    Thanks

    Abbas Ali
    FPGA Engineer 

  • Hi Abbas,

    However, we get quite unpredictable results in Vitis Serial Terminal.


    What do you mean? From one run to the other or from one setting to another?

    Have you populated all c304 - c305 and c306 - 307? If so, you should drop one of the pairs(c306 - 307). If you take a look at the datasheet connecting all previous mentioned capacitors will short the out 3 and 5 of the clock chip.

    We made the changes and tested things out too.
    It seems that there is a limitation of the clock chip(or software config).
    Desired freq vs obtained freq 
    250M -> 250M
    200M -> 250M
    150M -> 124M

    We will have a more in-depth look tomorrow.

    Andrei

  • Hi  ,

    I truly appreciate your response, and tomorrow we'll take an additional look. 

    We will have a more in-depth look tomorrow.

    Once you take a closer, more in-depth look tomorrow, kindly let me know as soon as you are able to.

    Yes, we have populated all four caps (C304, C305, C306, and C307). You are right about this configuration, we are shorting output 3 and Output 5 of the Clock Generation IC. We will unpopulated C306 and C307 and will run the test again and update you soon.

    It would be preferable if you could clear out our confusion on another question we have with an Analog Input Bandpass Filter.

    Given the desired signal at 1 Mhz (50% duty cycle PWM signal), we knew that 1 Mhz is relatively low in comparison to 50 Mhz, but when we applied that 1 Mhz analog sinusoidal input to the ad9467, the analog input bandpass filter was fine.

    Here is the picture:

    • Results that are required to be displayed on the ADI IIO Oscilloscope are obtained at the clock frequency of 170-180 Mhz displayed on the Vitis Serial Terminal.

    However, when we provided an analog signal other than sinusoidal input, the bandpass filter at the input side distorted our entire signal. As a result, the bandpass filter eliminated the majority of our signal's components. 

    Here is the snapshot:

    • Results that are required to be displayed on the ADI IIO Oscilloscope are obtained at the clock frequency of 170-180 Mhz displayed on the Vitis Serial Terminal.

    Therefore, please provide the input filter configuration that how the resistors and capacitors will be utilized to retain the voltage waveform (square, triangular, etc) at 1Mhz and 50% duty cycle.

    Thanks 
     and  

    Abbas Ali 
    FPGA Engineer

  • Hi  ,

    Did you try to resolve the clock chip issue..?

    Actually, we unpopulated pairs (C306-C307) and run the test again.

    After these modifications, we obtain the outcomes shown in Vitis Serial Terminal. We achieve outcomes comparable to your ADI Team, but with a 10x reduction.

    Desired freq vs obtained freq in Vitis Serial Terminal
    250M -> 25M
    200M -> 25M
    150M -> 12.5M

    Here are the issues:

    • Why this 10x reduction occurs in these obtained frequencies?
    • Is it still a problem to change the ad9467 onboard clock setup as described in this link:  https://wiki.analog.com/resources/fpga/xilinx/fmc/ad9467? We are getting stuck at this point, and the project is not moving forward, so if any changes need to be made to this link, kindly let me know as soon as possible.

    Please take a close look at it and inform us as soon as you can.

    Thanks 
    Abbas Ali
    FPGA Engineer 

  • Hi  ,

    What is your progress regarding clock chip limitation?

    We have looked more closely today and have come to the conclusion that the internal clock frequency can be adjusted up to 217MHz.

    The default setting for vco clk sel is '0', so we modify it to "1" because when we initialize it to "0," it switches to an external clock source in the otherwise condition of ad9517.c and the Vitis serial terminal displays a 10x reduction in frequency.

    Desired freq vs obtained freq in Vitis Serial Terminal (When vco_clk_sel=0)
    250M -> 25M
    200M -> 25M
    150M -> 12.5M

    We wish to use the ad9467's internal clock source, which ranges from 50 to 250 MHz, but we haven't provided it with any external clock sources, therefore it still enters this condition.

    When vco clk sel is initialised to '0', the code runs in the else condition (GREEN BOX) while debugging, but when we change vco clk sel to "1," it runs in the true condition (RED BOX) of the if statement.

    Desired freq vs obtained freq in Vitis Serial Terminal (When vco_clk_sel=1)
    250M -> 217M
    200M -> 163M
    150M -> 130M
    100M -> 81M

    Since we want a fixed clock frequency, we offer it from internal source code using the function ad9517 frequency(ad9517 device, 3, 250000000); as a consequence, kindly take a look at it to understand why this unexpected results are displayed in Vitis Serial Terminal.

    We were going to post three responses to you, but the ADI Team didn't offer any comments. To help us finish the project quickly, kindly expedite your response.

    Thanks
    Abbas Ali
    FPGA Engineer

  • Hi,

    I wrote a post but something went wrong, my reply did not get posted and I could only recover part of it.

    Abbas047 said:

    Desired freq vs obtained freq in Vitis Serial Terminal
    250M -> 25M
    200M -> 25M
    150M -> 12.5M

    This is strange. What frequency is written on the clock crystal(Y200). The board that I have has a 250MHz, but from what I see on the wiki it should be a 25MHz one, which does not make sense.
    What other changes have you made to the design, regarding the clock? 

    This is the clock path I'm using:

    Removed:
    C202
    C209 - C210
    J300 - jumper.

    Populated:
    C205 - C206
    C302 - C303
    C304 - C305



    Looking at the AD9517-4 clock chip and the schematic, there are only two options.

    1. Dividing the frequency passed through from the clock crystal.
    250/2=125
    250/3=83
    ...


    2. Using the Internal VCO (1.6GHz) of the AD9517-4 to obtain the desired(aprox) clock frequency.

    This will gave you more frequency options.
    To use this you have to select the VCO as a clock source.
    https://github.com/analogdevicesinc/no-OS/blob/master/projects/ad9467/src/app/ad9467_fmc.c#L110

     - 0, // vco_clk_sel
    + 1, // vco_clk_sel

    There is no option to use the external clock with the internal PLL to get a better result because of how the schematic is made.

    I see that you found the part of what I was looking.
    My results when using the VCO where closer to what I wanted.
    250 -> 249
    150 -> 148
    50 ->48

    Do you have an oscilloscope with at least 2G sampling to double check the clock result returned by software?

    Abbas047 said:

    Therefore, please provide the input filter configuration that how the resistors and capacitors will be utilized to retain the voltage waveform (square, triangular, etc) at 1Mhz and 50% duty cycle.

    You should  ask this on the High speeds ADC forum. Here we should keep the topic related to the software, besides, we do not have the resources. My guess is that you will not receive a straight answer there as well. In the links I posted above, you have the circuit model that you can simulate yourself. And solutions for scenarios where signals like square waves are used.

    Andrei