Post Go back to editing

EVAL-AD5940ELCZ Fast Chronoamperometry

Category: Software
Product Number: AD5940

Hi,

We have tested EVAL-AD5940ELCZ with on-board dummy and by connecting external 10uf+6.8kohms for various pulse amplitudes and below are the results we are gettiing

EXTERNAL ELECTRODES   RC JP9,10,11=C (10uf + 6.8kohms)											
AmpODR = 0.01  PULSE LENGTH = 1000mS

PA 500mV			Ipeak=73.52uA
LPTIARTIA10K
-69.204643
-64.748947
-59.661701
-55.889885
-51.308575
-44.539547
-38.450905
-33.586967
-29.023104
-25.460636
-21.978426
-19.358042
-16.67485
-14.752308
-12.651814
-11.249157
-9.598769
-8.586904
-7.271477
-6.559682
-5.498968
-5.024437
-4.152139

PA 900mV    Ipeak=132.35uA
LPTIARTIA6K
-108.445564
-101.244171
-93.135323
-87.010078
-79.482925
-69.163643
-59.367886
-52.195572
-44.872021
-39.729832
-34.116463
-30.405247
-26.030891
-23.349276
-19.882366
-17.986038
-15.182263
-13.885082



ON BOARD DUMMY SENSOR      JP6=A, JP9,10,11=B    RC   (10uf + 6.8kohms)											
AmpODR = 0.01  PULSE LENGTH = 1000mS											
PA 500mV			Ipeak=73.52
LPTIARTIA10K
-60.111782
-56.015472
-51.388805
-48.136879
-44.152225
-38.000782
-32.801884
-28.353165
-24.532505
-21.245684
-18.412462
-15.95956
-13.84162
-12.006307
-10.418725
-9.047472
-7.850679
-6.821369
-5.928135
-5.153535
-4.480121
-3.897426
-3.398472
-2.962323

PA 900mV
LPTIARTIA6K
-108.684792
-101.465897
-93.322105
-87.574921
-80.536346
-69.612022
-60.362995
-52.422791
-45.582001
-39.677742
-34.564606
-30.137869
-26.292831
-22.948059
-20.045378
-17.514982

1.Why are we getting all negative values?

Ported ad5940lib to nrf mcu and able to read chip id and adi id.

Now, we want to measure transient response of rc 1nf+4.7kohm.

We are trying to determine whether the current generated by rc 1nf+4.7kohm can be measured using LPTIA or HSTIA is required due to bandwidth and settling time contraints.

To evaluate this, we attempted measuring using HSTIA by setting switch config given as here  sw_cfg and ADC signal path as "Analog -> ADC -> FIFO (raw, source = "sinc3")". But during acquisition we only receive 10–11 samples in console.

2.Is the LPTIA sufficient for measuring current through a 1 nF + 4.7 kΩ RC network, or is the HSTIA required?

3.Are we missing any configurations to obtain continuous high-speed samples (beyond 10–11) when using the HSTIA with the raw ADC → FIFO path?

SM Config
.DswitchSel = SWD_OPEN,
.PswitchSel = SWP_OPEN,
.NswitchSel = SWN_OPEN,
.TswitchSel = SWT_SE0LOAD | SWT_TRTIA,

FIFO CONFIG
fifo_cfg.FIFOEn = bTRUE;
fifo_cfg.FIFOMode = FIFOMODE_FIFO;
fifo_cfg.FIFOSize = FIFOSIZE_4KB;                      
fifo_cfg.FIFOSrc = FIFOSRC_SINC3;
fifo_cfg.FIFOThresh = 10;  

ADC MUX CONFG
dsp_cfg.ADCBaseCfg.ADCMuxN = ADCMUXN_HSTIA_N;
dsp_cfg.ADCBaseCfg.ADCMuxP = ADCMUXP_HSTIA_P;   

ADC FILTER CONFIG
dsp_cfg.ADCFilterCfg.ADCAvgNum = ADCAVGNUM_16; 
dsp_cfg.ADCFilterCfg.ADCRate = ADCRATE_800KHZ;
dsp_cfg.ADCFilterCfg.ADCSinc2Osr = AppCHRONOAMPCfg.ADCSinc2Osr;
dsp_cfg.ADCFilterCfg.ADCSinc3Osr = AppCHRONOAMPCfg.ADCSinc3Osr;
dsp_cfg.ADCFilterCfg.BpSinc3 = bTRUE;
dsp_cfg.ADCFilterCfg.BpNotch = bTRUE;
dsp_cfg.ADCFilterCfg.Sinc2NotchEnable = bTRUE;         

Thank you

Edit Notes

Links
[edited by: Kala at 5:46 AM (GMT -5) on 20 Nov 2025]

Thread Notes

Parents
  • Hi,  .

    Can your team please help look into this query?

    Regards,
    Jo

  • Hi,
    Did not get a response to the query...??

    Modified Chronoamperometric example to configure HSDAC+HSTIA for transient measurement. Able to get 800KSPS ODR (ADCRATE_1P6MHZ ,ADCSINC3OSR_2).

    1.When tested with 1nf+4.7Kohm RC for 1ms pulse length, got 800 samples. But if changing pulse length to 2ms, not printing any data on console.

    ADCRATE_1P6MHZ ,ADCSINC3OSR_2 1mspulse -> DataCount: 803  no values when 2ms pulse

    ADCRATE_1P6MHZ ,ADCSINC3OSR_4 1mspulse -> DataCount: 402, 2mspulse ->DataCount: 802 no values when 3ms pulse

    same with ADCSINC3OSR_5 when pulselength is morethan 3ms, no values. In ISR fifocnt is zero. Why it is?

  • Hi,

    -ve value is due to direction of measured current.

    Use HPTIA for capcitive and inductive measurements. Use LPTIA for pure resistive measurements.

    You have set fifo_cfg.FIFOThresh = 10; 

    It displays every 10 measured samples.

  • Hi,

    When using HSDAC+HSTIA to get transient measurement, we are able to get correct pulse amplitude and pulse length, but bias voltage is not.

    Below is the modified code

    /* Generate init sequence */
    static AD5940Err AppCHRONOAMPSeqCfgGen(void)
    {
      AD5940Err error = AD5940ERR_OK;
      uint32_t const *pSeqCmd;
      uint32_t SeqLen;
    
      AFERefCfg_Type aferef_cfg;
      LPLoopCfg_Type lp_loop;
      DSPCfg_Type dsp_cfg;
      HSLoopCfg_Type hs_loop;
    
      /* Start sequence generator here */
      AD5940_SEQGenCtrl(bTRUE);
    
      // AD5940_AFECtrlS(AFECTRL_ALL, bFALSE);  /* Init all to disable state */
    
      aferef_cfg.HpBandgapEn = bTRUE;
      aferef_cfg.Hp1V1BuffEn = bTRUE;
      aferef_cfg.Hp1V8BuffEn = bTRUE;
      aferef_cfg.Disc1V1Cap = bFALSE;
      aferef_cfg.Disc1V8Cap = bFALSE;
      aferef_cfg.Hp1V8ThemBuff = bFALSE;
      aferef_cfg.Hp1V8Ilimit = bFALSE;
      aferef_cfg.Lp1V1BuffEn = bTRUE;
      aferef_cfg.Lp1V8BuffEn = bTRUE;
      /* LP reference control - turn off them to save power*/
      aferef_cfg.LpBandgapEn = bTRUE;
      aferef_cfg.LpRefBufEn = bTRUE;
      aferef_cfg.LpRefBoostEn = bFALSE;
      AD5940_REFCfgS(&aferef_cfg);
    
      lp_loop.LpDacCfg.LpdacSel = LPDAC0;
      lp_loop.LpDacCfg.LpDacSrc = LPDACSRC_MMR;
      lp_loop.LpDacCfg.LpDacSW = LPDACSW_VBIAS2LPPA | /*LPDACSW_VBIAS2PIN|*/ LPDACSW_VZERO2LPTIA | LPDACSW_VZERO2PIN | LPDACSW_VZERO2HSTIA;
      lp_loop.LpDacCfg.LpDacVzeroMux = LPDACVZERO_6BIT;
      lp_loop.LpDacCfg.LpDacVbiasMux = LPDACVBIAS_12BIT;
      lp_loop.LpDacCfg.LpDacRef = LPDACREF_2P5;
      lp_loop.LpDacCfg.DataRst = bFALSE;
      lp_loop.LpDacCfg.PowerEn = bTRUE;
      lp_loop.LpDacCfg.DacData6Bit = (uint32_t)((AppCHRONOAMPCfg.Vzero - 200) / DAC6BITVOLT_1LSB);
      lp_loop.LpDacCfg.DacData12Bit = (int32_t)((AppCHRONOAMPCfg.SensorBias) / DAC12BITVOLT_1LSB) + lp_loop.LpDacCfg.DacData6Bit * 64;
      if (lp_loop.LpDacCfg.DacData12Bit > lp_loop.LpDacCfg.DacData6Bit * 64)
        lp_loop.LpDacCfg.DacData12Bit--;
    
      lp_loop.LpAmpCfg.LpAmpSel = LPAMP0;
      lp_loop.LpAmpCfg.LpAmpPwrMod = LPAMPPWR_NORM;
      lp_loop.LpAmpCfg.LpPaPwrEn = bFALSE;
      lp_loop.LpAmpCfg.LpTiaPwrEn = bFALSE;
      lp_loop.LpAmpCfg.LpTiaRf = AppCHRONOAMPCfg.LpTiaRf;
      lp_loop.LpAmpCfg.LpTiaRload = AppCHRONOAMPCfg.LpTiaRl;
      if (AppCHRONOAMPCfg.ExtRtia == bTRUE)
      {
        lp_loop.LpAmpCfg.LpTiaRtia = LPTIARTIA_OPEN;
        lp_loop.LpAmpCfg.LpTiaSW = LPTIASW(9) | LPTIASW(2) | LPTIASW(4) | LPTIASW(5) | /*LPTIASW(12)|*/ LPTIASW(13);
      }
      else
      {
        lp_loop.LpAmpCfg.LpTiaRtia = AppCHRONOAMPCfg.LptiaRtiaSel;
        lp_loop.LpAmpCfg.LpTiaSW = LPTIASW(7) | LPTIASW(5) | LPTIASW(2) | LPTIASW(4) | /*LPTIASW(12)|*/ LPTIASW(13);
      }
      AD5940_LPLoopCfgS(&lp_loop);
    
      dsp_cfg.ADCBaseCfg.ADCMuxN = ADCMUXN_HSTIA_N;
      dsp_cfg.ADCBaseCfg.ADCMuxP = ADCMUXP_HSTIA_P;
    
      dsp_cfg.ADCBaseCfg.ADCPga = AppCHRONOAMPCfg.ADCPgaGain;
    
      memset(&dsp_cfg.ADCDigCompCfg, 0, sizeof(dsp_cfg.ADCDigCompCfg));
      // memset(&dsp_cfg.DftCfg, 0, sizeof(dsp_cfg.DftCfg));
    
      dsp_cfg.ADCFilterCfg.ADCAvgNum = ADCAVGNUM_2;           /* Don't care because it's disabled */
      dsp_cfg.ADCFilterCfg.ADCRate = AppCHRONOAMPCfg.ADCRate; // ADCRATE_1P6MHZ; /* Tell filter block clock rate of ADC*/
      dsp_cfg.ADCFilterCfg.ADCSinc2Osr = AppCHRONOAMPCfg.ADCSinc2Osr;
      dsp_cfg.ADCFilterCfg.ADCSinc3Osr = AppCHRONOAMPCfg.ADCSinc3Osr;
      dsp_cfg.ADCFilterCfg.BpSinc3 = AppCHRONOAMPCfg.BpSinc3;
      dsp_cfg.ADCFilterCfg.BpNotch = bTRUE;
      dsp_cfg.ADCFilterCfg.Sinc2NotchEnable = AppCHRONOAMPCfg.Sinc2NotchEnable;
    
      memset(&dsp_cfg.StatCfg, 0, sizeof(dsp_cfg.StatCfg)); /* Don't care about Statistic */
      AD5940_DSPCfgS(&dsp_cfg);
    
      hs_loop.HsDacCfg.ExcitBufGain = AppCHRONOAMPCfg.ExcitBufGain;
      hs_loop.HsDacCfg.HsDacGain = AppCHRONOAMPCfg.HsDacGain;
      hs_loop.HsDacCfg.HsDacUpdateRate = AppCHRONOAMPCfg.HsDacUpdateRate;
    
      /* Configure HSTIA to measure Current */
      hs_loop.HsTiaCfg.DiodeClose = bFALSE;
      hs_loop.HsTiaCfg.HstiaBias = HSTIABIAS_1P1;
      hs_loop.HsTiaCfg.HstiaCtia = AppCHRONOAMPCfg.CtiaSel; /* 31pF + 2pF */
      hs_loop.HsTiaCfg.HstiaDeRload = HSTIADERLOAD_OPEN;
      hs_loop.HsTiaCfg.HstiaDeRtia = HSTIADERTIA_OPEN;
      hs_loop.HsTiaCfg.HstiaRtiaSel = AppCHRONOAMPCfg.HstiaRtiaSel;
    
    
      hs_loop.SWMatCfg.Dswitch = SWD_CE0;
      hs_loop.SWMatCfg.Pswitch = SWP_RE0; 
      hs_loop.SWMatCfg.Nswitch = SWN_SE0; 
      hs_loop.SWMatCfg.Tswitch = SWT_TRTIA | SWT_SE0LOAD;
    
      AD5940_HSLoopCfgS(&hs_loop);
    
      /* Enable all of them. They are automatically turned off during hibernate mode to save power */
      AD5940_AFECtrlS(AFECTRL_HSTIAPWR | AFECTRL_WG | AFECTRL_DACREFPWR | AFECTRL_HSDACPWR | AFECTRL_INAMPPWR | AFECTRL_EXTBUFPWR, bTRUE);
      AD5940_SEQGpioCtrlS(0);
      /* Sequence end. */
      AD5940_SEQGenInsert(SEQ_STOP()); /* Add one extra command to disable sequencer for initialization sequence because we only want it to run one time. */
      /* Stop here */
      error = AD5940_SEQGenFetchSeq(&pSeqCmd, &SeqLen);
      AD5940_SEQGenCtrl(bFALSE); /* Stop sequencer generator */
      if (error == AD5940ERR_OK)
      {
        AppCHRONOAMPCfg.InitSeqInfo.SeqId = SEQID_1;
        AppCHRONOAMPCfg.InitSeqInfo.SeqRamAddr = AppCHRONOAMPCfg.SeqStartAddr;
        AppCHRONOAMPCfg.InitSeqInfo.pSeqCmd = pSeqCmd;
        AppCHRONOAMPCfg.InitSeqInfo.SeqLen = SeqLen;
        /* Write command to SRAM */
        AD5940_SEQCmdWrite(AppCHRONOAMPCfg.InitSeqInfo.SeqRamAddr, pSeqCmd, SeqLen);
      }
      else
        return error; /* Error */
      return AD5940ERR_OK;
    }
    
    static AD5940Err AppCHRONOAMPTransientMeasureGen(void)
    {
      AD5940Err error = AD5940ERR_OK;
      uint32_t const *pSeqCmd;
      uint32_t SeqLen;
      uint32_t VbiasCode, VzeroCode;
      uint32_t WaitClks;
      ClksCalInfo_Type clks_cal;
    
      SWMatrixCfg_Type sw_cfg;
    
      uint32_t hs_mid = 0x800, hs_hi = 0x800;
      float vfs_mv;
    
      clks_cal.DataType = AppCHRONOAMPCfg.DataFifoSrc;
      clks_cal.DataCount = AppCHRONOAMPCalcDataNum(AppCHRONOAMPCfg.pulseLength);
      clks_cal.ADCSinc2Osr = AppCHRONOAMPCfg.ADCSinc2Osr;
      clks_cal.ADCSinc3Osr = AppCHRONOAMPCfg.ADCSinc3Osr;
      clks_cal.ADCAvgNum = 0;
      clks_cal.RatioSys2AdcClk = AppCHRONOAMPCfg.SysClkFreq / AppCHRONOAMPCfg.AdcClkFreq;
      AD5940_ClksCalculate(&clks_cal, &WaitClks);
      LOG_INF("WaitClks : %d\n", WaitClks);
      LOG_INF("PulseAmplitude: %f\n", AppCHRONOAMPCfg.pulseAmplitude);
    
      vfs_mv = 404.0f * 2 * 1;
    
      {
        int32_t delta = (int32_t)((AppCHRONOAMPCfg.pulseAmplitude / vfs_mv) * 2048.0f + 0.5f);
        if (delta > 2047)
          delta = 2047;
        if (delta < -2047)
          delta = -2047;
        hs_hi = (uint32_t)(0x800 + delta);
        printk("\rhs_hi : %d\r\n", hs_hi);
      }
    
      AD5940_SEQGenCtrl(bTRUE);
      AD5940_SEQGpioCtrlS(AGPIO_Pin1);
    
      /* Ensure switch matrix is correctly connected for the HS path */
      {
        SWMatrixCfg_Type sw_cfg;
        sw_cfg.Dswitch = SWD_CE0;
        sw_cfg.Pswitch = SWP_CE0;
        sw_cfg.Nswitch = SWN_SE0;
        sw_cfg.Tswitch = SWT_TRTIA | SWT_SE0LOAD;
        AD5940_SWMatrixCfgS(&sw_cfg);
      }
    
      AD5940_AFECtrlS(AFECTRL_HSTIAPWR | AFECTRL_WG | AFECTRL_DACREFPWR | AFECTRL_HSDACPWR | AFECTRL_INAMPPWR | AFECTRL_EXTBUFPWR, bTRUE);
    
      /* Power up ADC and start conversions just before applying the step */
      AD5940_AFECtrlS(AFECTRL_ADCPWR, bTRUE);
      AD5940_SEQGenInsert(SEQ_WAIT(16 * 250)); /* small settle */
      AD5940_WriteReg(REG_AFE_HSDACDAT, hs_mid);
      AD5940_SEQGenInsert(SEQ_WAIT(16 * 100));
      AD5940_AFECtrlS(AFECTRL_ADCCNV, bTRUE);
      AD5940_SEQGenInsert(SEQ_WAIT(16 * 10));
      AD5940_WriteReg(REG_AFE_HSDACDAT, hs_hi);
      AD5940_SEQGenInsert(SEQ_WAIT(WaitClks));
      AD5940_AFECtrlS(AFECTRL_ADCPWR | AFECTRL_ADCCNV, bFALSE);
      /* Return HSDAC to midscale and stop ADC */
      AD5940_WriteReg(REG_AFE_HSDACDAT, hs_mid);
    
      AD5940_AFECtrlS(AFECTRL_HSTIAPWR | AFECTRL_WG | AFECTRL_DACREFPWR | AFECTRL_HSDACPWR | AFECTRL_INAMPPWR | AFECTRL_EXTBUFPWR, bFALSE);
    
      AD5940_SEQGpioCtrlS(0);
      /* Sequence end. */
      AD5940_SEQGenInsert(SEQ_STOP()); /* Add one extra command to disable sequencer for initialization sequence because we only want it to run one time. */
      AD5940_EnterSleepS();            /* Goto hibernate */
      /* Sequence end. */
      error = AD5940_SEQGenFetchSeq(&pSeqCmd, &SeqLen);
      AD5940_SEQGenCtrl(bFALSE); /* Stop sequencer generator */
    
      if (error == AD5940ERR_OK)
      {
        AppCHRONOAMPCfg.TransientSeqInfo.SeqId = SEQID_2;
        AppCHRONOAMPCfg.TransientSeqInfo.SeqRamAddr = AppCHRONOAMPCfg.MeasureSeqInfo.SeqRamAddr + AppCHRONOAMPCfg.MeasureSeqInfo.SeqLen;
        AppCHRONOAMPCfg.TransientSeqInfo.pSeqCmd = pSeqCmd;
        AppCHRONOAMPCfg.TransientSeqInfo.SeqLen = SeqLen;
        /* Write command to SRAM */
        AD5940_SEQCmdWrite(AppCHRONOAMPCfg.TransientSeqInfo.SeqRamAddr, pSeqCmd, SeqLen);
      }
      else
        return error; /* Error */
      return AD5940ERR_OK;
    }

    With

    pAMPCfg->SensorBias = 0;
    pAMPCfg->Vzero = 1100;
    pAMPCfg->pulseAmplitude = 500;
    pAMPCfg->pulseLength = 1;

    Pulse is captured in oscilloscope(CH1 -Yellow -CE0 , CH2-Blue-SE0) is like below

    after pulse

    And also when testing multiple times there is offset in current measured

    Any calibration needed to remove the offset?

    Hi,
    Did not get a response to the query...??

    Modified Chronoamperometric example to configure HSDAC+HSTIA for transient measurement. Able to get 800KSPS ODR (ADCRATE_1P6MHZ ,ADCSINC3OSR_2).

    1.When tested with 1nf+4.7Kohm RC for 1ms pulse length, got 800 samples. But if changing pulse length to 2ms, not printing any data on console.

    ADCRATE_1P6MHZ ,ADCSINC3OSR_2 1mspulse -> DataCount: 803  no values when 2ms pulse

    ADCRATE_1P6MHZ ,ADCSINC3OSR_4 1mspulse -> DataCount: 402, 2mspulse ->DataCount: 802 no values when 3ms pulse

    same with ADCSINC3OSR_5 when pulselength is morethan 3ms, no values. In ISR fifocnt is zero. Why it is?

    Can you also look into this?

    Thank you

  • VBias and Vzero are expected to be at 1.1V for you configuration. Is that what you are getting?

    In 2 ms or higher pulse length case, you may reduce the sample rate (ODR-Output data rate or sample delay) to give enough time to complete each measurement. 

  • Hi,

    1.As you can see in the captured waveform, we are not getting Vbias and Vzero 1.1V. 

    In 2 ms or higher pulse length case, you may reduce the sample rate (ODR-Output data rate or sample delay) to give enough time to complete each measurement. 

    2.We want to measure FCA with ADCRATE_1P6MHZ bypassing SINC filters. But, as I said earlier, not able to get any data if pulse length is 1ms also. Can you help us with the configurations to measure with different pulse length with 1.6M adc rate with sinc filters bypassed?

    Thank you

  • You may sweep pAMPCfg->AmpODR from 1 to 20 and check.

  • Hi,

    We want to measure only transient response when pulse is applied, not normal aerometric measurements.

    while (1)
      {
        /* Check if interrupt flag which will be set when interrupt occurred. */
         if(AD5940_GetMCUIntFlag())
        {
          printk("FILE : %s LINE : %d\n",__FILE__,__LINE__);
           AD5940_ClrMCUIntFlag(); /* Clear this flag */
          temp[IntCount] = APPBUFF_SIZE;
          AppCHRONOAMPISR(AppBuff[IntCount], &temp[IntCount]); /* Deal with it and provide a buffer to store data we got */
          if (pAMPCfg->bMeasureTransient == bFALSE)
          {
            // AppCHRONOAMPCtrl(CHRONOAMPCTRL_STOPNOW, 0);
            //AMPShowResult((float *)AppBuff[0], temp[0]);
          }
          if (pAMPCfg->EndSeq) /* End sequence only set at end of transient */
          {
            for (int i = 0; i < IntCount; i++)
            {
              AMPShowResult((float *)AppBuff[i], temp[i]); /* Show the results to UART */
            }
            pAMPCfg->EndSeq = bFALSE;
            pAMPCfg->bMeasureTransient = bFALSE;
            IntCount = 0;
          }
        }
      }

    And transient sequence is triggered by mmr not by wakeup timer. Will pAMPCfg->AmpODR make any difference in our case?

    Thank you

Reply
  • Hi,

    We want to measure only transient response when pulse is applied, not normal aerometric measurements.

    while (1)
      {
        /* Check if interrupt flag which will be set when interrupt occurred. */
         if(AD5940_GetMCUIntFlag())
        {
          printk("FILE : %s LINE : %d\n",__FILE__,__LINE__);
           AD5940_ClrMCUIntFlag(); /* Clear this flag */
          temp[IntCount] = APPBUFF_SIZE;
          AppCHRONOAMPISR(AppBuff[IntCount], &temp[IntCount]); /* Deal with it and provide a buffer to store data we got */
          if (pAMPCfg->bMeasureTransient == bFALSE)
          {
            // AppCHRONOAMPCtrl(CHRONOAMPCTRL_STOPNOW, 0);
            //AMPShowResult((float *)AppBuff[0], temp[0]);
          }
          if (pAMPCfg->EndSeq) /* End sequence only set at end of transient */
          {
            for (int i = 0; i < IntCount; i++)
            {
              AMPShowResult((float *)AppBuff[i], temp[i]); /* Show the results to UART */
            }
            pAMPCfg->EndSeq = bFALSE;
            pAMPCfg->bMeasureTransient = bFALSE;
            IntCount = 0;
          }
        }
      }

    And transient sequence is triggered by mmr not by wakeup timer. Will pAMPCfg->AmpODR make any difference in our case?

    Thank you

Children
  • No. For transient PULSE test, AmpODR doesn't matter.

    PULSE test is done only once.

    With ADCSINC2OSR_44 and ADCSINC3OSR_4, 

    For 2ms, datacount is AppCHRONOAMPCalcDataNum(2) = 9 samples => (pAMPCfg->FifoThresh must be 9)

    For 1ms, datacount calculated will be 4 samples => (pAMPCfg->FifoThresh must be 4)

    With ADCSINC2OSR_44 and ADCSINC3OSR_2, 

    For 2ms, datacount is AppCHRONOAMPCalcDataNum(2) = 18 samples => (pAMPCfg->FifoThresh must be 18)

    For 1ms, datacount calculated will be 8 samples => (pAMPCfg->FifoThresh must be 8)

  • Hi,

    Now, with RC 4.7K+1nf we are able to achieve  ODR of 1.6M, but the problem is we are not able to test with this ODR for pulse length >500us.

    Similarly with 800K ODR not able to test for pulse length >1ms

    We want to test FCA for different pulse lengths, with 1.6M ODR and no SINC filters.

    For pulse test,  fifo threshold in CHRONOAMPCTRL_PULSETEST will be considered right? That we are keeping 1000.

    Application Context: We are targeting an oversampled capture rate because our sensor reactions are extremely fast (time constant of ~7-10µs). We need to capture as much detail as possible during a very brief window (<200µs). We are aware that higher speeds will increase noise, but we need to evaluate the SNR trade-offs to find the optimal ODR.

    Thank you

  • For transient test, you need to keep FifoThresh  as I mentioned above. If you keep 1000 as threshold, interrupt will not occur till 1000 samples are filled, which will not happen because ADC is turned off after clks_cal.DataCount number of samples. Hence no output will be seen.

    If no Sinc filters are needed,

    you need to update in AppCHRONOAMPTransientMeasureGen() as below:

    clks_cal.DataCount =1000;

    instead of

    clks_cal.DataCount = AppCHRONOAMPCalcDataNum(AppCHRONOAMPCfg.pulseLength);

  • Hi,

    Thank you for the input, will test with the modifications suggested.

    VBias and Vzero are expected to be at 1.1V for you configuration. Is that what you are getting?

    We are not able to get vbias and vzero 1.1v as expected.

    modified code

    1. Is configuration correct in the given code?

    2. Also about current offset, what could be the possible reasons and any calibrations need to be done?

    Thank you

  • VBias and Vzero return to DC bias values after transient response is measured.

    For this measurement must be first completed, which needs correcting Datacount and ensuring that output is seen. 

    BUFSENCON[V1P1LPADCCHGDIS, V1P8HSADCCHGDIS] register bits may be closed to discharge and remove unexpected offset, and then opened again.

  • Hi,

    1.

    If no Sinc filters are needed,

    you need to update in AppCHRONOAMPTransientMeasureGen() as below:

    clks_cal.DataCount =1000;

    If datacount is fixed to 1000, then pulse duration will not be set to pAMPCfg->pulseLength given right? We want to run very long experiments by bypass all sinc filters and capture raw ADC @ 1.6 MSPS. With the limitation of 1k FIFO samples we are not able to capture the Pulse data of even 1ms pulse length.

    Can you let me know if it is possible to achieve this in AD5940,if it is possible, can you provide the FIFO and other configuration required to achieve this? What is the ADC data rate we can use to achieve this?
    2.

    VBias and Vzero return to DC bias values after transient response is measured.

    For this measurement must be first completed, which needs correcting Datacount and ensuring that output is seen. 

    As I mentioned before, we want to capture only transient response. After output is read, we are stopping the measurement. If we are able to see output on console after endseq means transient response is measured so Vbias and vzero should return to DC bias right? But in our case, it's not.

     
    3. DAC slope in FCA : the goal is to limit the fastest slope (dV/dt) of the voltage applied, preventing it from changing faster than a user-defined rate. How to set this?
    Thank you
  • 1) Datacount = (PulseLength_in_ms)*800/ADCSinc2Osr/ADCSinc3Osr;

    Max data count = 1024 as data FIFO size is 4kB .

    If ADCSinc2Osr = 1333, ADCSinc3Osr = 5,

    Datacount is 1024 for pulse length of 8.5 secs.

    Hence you may measure for pulse length upto 8.5 seconds.

    If Sinc filters need to be bypassed,

    you may set datacount = x.

    Pulselength would be "waitclks" (derived from datacount =x) in ms. 

     2) You can see the output even when pulse is not fully completed. 

    This happens if fifothreshold is set smaller than datacount. 

    VBias and Vzero are restored in code only after Pulse length duration is completed.

    3) LPDAC can source the data from waveform generator:

    lp_loop.LpDacCfg.LpDacSrc = LPDACSRC_WG;

    where DAC update rate is set by:

    HsDacCfg.HsDacUpdateRate = 7;       /* DAC update rate equals to WG_CLK/HsDacUpdateRate*/

    Reference: ad5940-examples/examples/AD5940_WG at master · analogdevicesinc/ad5940-examples · GitHub

  • Hi,

    As i mentioned earlier,

    modified code

    1.we are using HSDAC to apply pulse, and want to update DAC at user defined slope rate which is v/s. How to set it using HsDacUpdateRate?

    2. In the modified code given,

    AD5940_WriteReg(REG_AFE_HSDACDAT, hs_hi);
    AD5940_SEQGenInsert(SEQ_WAIT(WaitClks));
    AD5940_AFECtrlS(AFECTRL_ADCPWR | AFECTRL_ADCCNV, bFALSE);
    /* Return HSDAC to midscale and stop ADC */
    AD5940_WriteReg(REG_AFE_HSDACDAT, hs_mid);
    
    AD5940_AFECtrlS(AFECTRL_HSTIAPWR | AFECTRL_WG | AFECTRL_DACREFPWR | AFECTRL_HSDACPWR | AFECTRL_INAMPPWR | AFECTRL_EXTBUFPWR, bFALSE);

    setting all AFECTRLs to False, once conversion is done. Is it the reason for not getting correct bias?

    Thank you

  • DAC update rate is set by 8 bit value:

    You may fetch input byte from user and set HsDacUpdateRate.

  • Hi,

    1. I’m having difficulty understanding this part. Could you please explain it more clearly or with an example how it works to set v/s slope given by user?

    2.To measure OCP,  AD5941 Potentiometry Basic Example and OCP Basic Example, it is given VRE0, VSE0 wrt VZERO0. Measuring OCP (Open Circuit Potential) of a 3 electrode system, here it is given wrt ADCMUXN_VSET1P1. When exactly to use vzero0 and vset1p1??

    float App_OCP(void)
    {
      uint32_t afeResult;
      float V_RE0, V_SE0;
    
      SWMatrixCfg_Type sw_cfg;
    
      sw_cfg.Dswitch=SWD_OPEN;
      sw_cfg.Pswitch=SWP_RE0;
      sw_cfg.Nswitch=SWN_SE0;
      sw_cfg.Tswitch = SWT_TRTIA | SWT_SE0LOAD;
    
      AD5940_SWMatrixCfgS(&sw_cfg);
    
        /*Measure VSE*/
      //AD5940_ADCMuxCfgS(ADCMUXP_VSE0, ADCMUXN_VSET1P1);
      AD5940_ADCMuxCfgS(ADCMUXP_VSE0, ADCMUXN_VZERO0);
      AD5940_AFECtrlS(AFECTRL_ADCPWR /*| AFECTRL_SINC2NOTCH*/, bTRUE);
      AD5940_Delay10us(16 * 25);
      AD5940_AFECtrlS(AFECTRL_ADCCNV, bTRUE); /* Start ADC convert*/
      while (!(AD5940_INTCTestFlag(AFEINTC_1, AFEINTSRC_ADCRDY)))
        ;
      AD5940_INTCClrFlag(AFEINTSRC_ADCRDY);
      afeResult = AD5940_ReadAfeResult(AFERESULT_SINC3);
      AD5940_AFECtrlS(AFECTRL_ADCPWR | AFECTRL_ADCCNV /*| AFECTRL_SINC2NOTCH*/, bFALSE); /* Stop ADC */
      V_SE0 = AD5940_ADCCode2Volt(afeResult & 0xffff, ADCPGA_1, 1.82);
    
      /*Measure VRE*/
      //AD5940_ADCMuxCfgS(ADCMUXP_VRE0, ADCMUXN_VSET1P1);
      AD5940_ADCMuxCfgS(ADCMUXP_VRE0, ADCMUXN_VZERO0);
      AD5940_AFECtrlS(AFECTRL_ADCPWR /*| AFECTRL_SINC2NOTCH*/, bTRUE);
      AD5940_Delay10us(16 * 25);
      AD5940_AFECtrlS(AFECTRL_ADCCNV, bTRUE); /* Start ADC convert*/
      while (!(AD5940_INTCTestFlag(AFEINTC_1, AFEINTSRC_ADCRDY)))
        ;
      AD5940_INTCClrFlag(AFEINTSRC_ADCRDY);
      afeResult = AD5940_ReadAfeResult(AFERESULT_SINC3);
      AD5940_AFECtrlS(AFECTRL_ADCPWR | AFECTRL_ADCCNV /*| AFECTRL_SINC2NOTCH*/, bFALSE); /* Stop ADC */
      V_RE0 = AD5940_ADCCode2Volt(afeResult & 0xffff, ADCPGA_1, 1.82);
    
      return (V_SE0 - V_RE0);
    }
    
    wrt VZERO0, results are as below
    OCP=-0.1539 V 
    OCP=-0.1544 V 
    OCP=-0.1532 V 
    OCP=-0.1449 V 
    OCP=-0.1540 V 
    OCP=-0.1539 V 
    OCP=-0.1529 V 
    OCP=-0.1552 V 
    OCP=-0.1535 V 
    OCP=-0.1551 V 
    OCP=-0.1532 V 
    OCP=-0.1550 V 
    OCP=-0.1543 V 
    OCP=-0.1553 V 
    OCP=-0.1545 V 
    OCP=-0.1557 V 
    OCP=-0.1547 V 
    OCP=-0.1552 V 
    OCP=-0.1543 V 
    OCP=-0.1556 V 
    
    wrt VSET1P1
    OCP=-0.4020 V 
    OCP=-0.4303 V 
    OCP=-0.4166 V 
    OCP=-0.4314 V 
    OCP=-0.4346 V 
    OCP=-0.4293 V 
    OCP=-0.4475 V 
    OCP=-0.4209 V 
    OCP=-0.5000 V 
    OCP=-0.3882 V 
    OCP=-0.4947 V 
    OCP=-0.4169 V 
    OCP=-0.4478 V 
    OCP=-0.5628 V 
    OCP=-0.4950 V 
    OCP=-0.4123 V 
    OCP=-0.4752 V 
    OCP=-0.3938 V 
    OCP=-0.4337 V 
    OCP=-0.4250 V 
    
    

    3.Want to use external rtia.

    On EVAL-AD5940ELCZ, as you have mentioned here EXT DISCRETES, desired external resistance should be placed in RTIA1/RTIA2/RTIA3 right? According to data sheet, it should be connected between DE0 and AINx, but in schematic it is between AIN0 and AIN1/2/3. How to connect HS external rtia and what should be the switch matrix config?

    Thankyou

Before You Switch