ADIS16480 test conditions of Static and Dynamic Accuracy value

Thread Summary

The user inquired about the static and dynamic accuracy values of the ADIS16480's yaw, pitch, and roll angles. The dynamic accuracy is tested using a sinusoidal angular rotation of 30 degrees peak-to-peak at 0.6 Hz, with 4 cycles over a 10-second window. The sensor samples data at 1200 SPS, and the dynamic accuracy is determined by the standard deviation of the angular position error.
AI Generated Content
Category: Datasheet/Specs
Product Number: ADIS16480

Hi sir

According to the datasheet, The static accuracy of the ADIS16480's yaw angle is 0.3 degrees, and the roll and pitch angles are 0.1 degrees. The dynamic accuracy of the yaw angle is 0.5 degrees, and the roll and pitch angles are 0.3 degrees.

I would like to know how these values are obtained and whether they are based on any relevant international standards, or if there are any related tests(include both static and dynamic testing)?

I have read the following article 'ADIS16480 test conditions of Dynamic Accuracy value', but I don't understand the meaning of '4 cycles @ 0.6Hz.' Does it refer to a sine wave with 4 cycles and a frequency of 0.6Hz? If so, the total time would be less than 10 seconds, which seems to contradict the 'measured over 10 sec window' statement.

Hi Tzuyang,

Dynamic accuracy refers to the sensor’s ability to measure angles like yaw, pitch, and roll while in motion. The ADIS16480’s dynamic accuracy is characterized under specific test conditions that simulate real-world motion scenarios. Here’s a detailed overview of the conditions used during these tests (referenced here (1) ADIS16480 test conditions of Dynamic Accuracy value - Q&A - MEMS Inertial Sensors - EngineerZone (analog.com)):

  • Sinusoidal Angular Rotation: The sensor is subjected to a smooth, oscillatory motion like a sine wave. This type of motion closely resembles the kinds of movements the sensor might experience in practical applications.
  • Frequency and Cycles: The test involves rotating the sensor through four complete cycles at a frequency of 0.6 Hz, which means each cycle (back and forth) takes approximately 1.67 seconds. The relatively slow frequency allows for a detailed sensor response assessment.
  • Amplitude: The motion covers a peak-to-peak angular range of 30 degrees. This means the sensor rotates 15 degrees from the center position in one direction, then 15 degrees in the opposite direction. Testing over this range ensures the sensor’s accuracy is evaluated across significant angular movements.
  • Duration of Measurement: Data is collected over 10 seconds, long enough to encompass all four motion cycles. This duration provides sufficient data to assess the sensor’s performance accurately.
  • Alignment and Motion Generation: The Device Under Test (DUT) is aligned directly with the axis of rotation to ensure the motion is accurately applied to the measured axis. The motion is generated using a motor, and the sensor is tested around both horizontal and vertical axes, which correspond to pitch, roll, and yaw measurements.
  • Sampling Rate and Synchronization: The sensor samples data at 1200 samples per second (SPS) with no decimation (Decimation = 0). The data collection is synchronized with an external trigger, ensuring that all the measurements are perfectly aligned in time across the different axes.
  • Error Analysis: After collecting approximately 10 seconds of data, the standard deviation of the error in angular position is calculated. This statistical measure reflects the sensor’s dynamic accuracy, showing how much the measured angular position deviates from the actual value during motion.

I hope this explanation helps. If you have any other questions, please let me know. 

thank you