ADIS16505 Scaled Sync Mode with GNSS PPS Signal Reports DIAG_STAT = 0x8000 (Clock Error)

Category: Hardware
Product Number: ADIS16505-2

I'm feeding the ADIS16505 SYNC pin with a 1Hz PPS signal from a GNSS receiver and the IMU is configured in scaled sync mode with UP_SCALE = 2000.

I'm receiving data continuously from the IMU at 400Hz in burst mode successfully, however the DIAG_STAT flags sporadically report 0x8000 (clock error). When this occurs it seems to be sticky for a period of several seconds, i.e. there will be a long period of time in which this error is reported continuously, followed by an extended period of time where the error disappears. In a period of 15 minutes I recorded 215000 clock errors @ 400 Hz output sample rate = approximately 60% of burst transfers reported this error flag.

The datasheet suggestion to "adjust the frequency of the clock signal on the SYNC pin to operate within the appropriate range" is useless in this situation as the timing of the PPS signal is out of my control. Is the meaning intended here to instead make an adjustment to the UP_SCALE register?

This seems to be a fairly common problem across a number of different ADI IMUs when configured for external sync mode with a 1 Hz external clock. Here are some similar posts I found, none of which had a satisfactory resolution that actually got to the root of the issue.

 ADIS16467 Clock Error in Scaled Sync mode 

 ADIS16465 Reading buffer data is abnormal in scaled sync mode 

 ADIS 16465 Clock Error 

 ADIS16475 vs GPS

Is the datasheet wrong in stating that a 1Hz external clock is a supported configuration?

I have confirmed the PPS signal has consistent timing (measured in an edge capture interrupt by a microcontroller running off an 8MHz 30ppm oscillator / 216MHz core clock).

PPS period - mean 999983 us, sd 1 us, min 999971 us, max 999994 us

I've also confirmed the PPS waveform meets the requirements of the ADIS - it is a 1ms low pulse idling high at 3.3V otherwise, and the waveform is clean.

The DRDY signal coming out of the IMU looks good as well (400Hz 32-bit burst format, 5:1 decimation)

DRDY period - mean 2500 us, sd 3 us, min 2263 us, max 2530 us

What can I do to resolve this issue?

* As an aside - does a burst read automatically clear the error flags? The behaviour observed here suggests that is the case but it isn't explicitly stated in the datasheet.

  • Hi, thank you for posting this question.

    In IMUs like the ADIS16505, the external clock signal, such as a 1 Hz PPS from a GNSS, is scaled internally to synchronize sensor data sampling. This ensures that data sampling remains consistent with the external clock, maintaining accurate timing across data sets. The IMU also monitors the integrity of this clock signal, setting error flags like the 0x8000 in the DIAG_STAT register if it detects deviations due to noise or signal drop, indicating synchronization issues. 

    Even if the PPS signal appears stable, high-frequency noise or short-duration anomalies might not be visible without specialized equipment. Consider using a high-bandwidth oscilloscope to check the signal more thoroughly for precision and repeatability (Share these measurements). 

    Another suggestion is to test the system with a different clock source to see if the issue persists.

    Second Querry: Clear error flags in the DIAG_STAT register require specific action, such as writing to a register or power-cycling the device, as burst reading doesn't achieve this.

  •    I will try to get an oscilloscope capture of the PPS signal as close to the IMU as possible to see if there are any anomolies.

    re: clearing the DIAG_STAT during a burst read - Are you sure it is the case that the burst read function doesn't internally trigger the DIAG_STAT flags to be cleared? I ask as I have evidence to the contrary. My code currently does not explicitly perform a read of the DIAG_STAT register between burst reads to clear the flags - I skip this step because I can see that the flags are clearing themselves after some time. Is there perhaps a timeout or some other logic that would explain the flags being reset to 0 if they are not being reset by the burst read?

    This distinction matters because my current interpretation is that the clock errors are occurring frequently. If the behaviour is as you stated then the occurrence may be much rarer and I will have to do a deeper investigation to produce conclusive evidence.

  •   

    This is the 1 PPS signal input to the IMU. Both of these were captured with infinite persistence over several minutes. During this measurement I observed hundreds of thousands of clock errors in the DIAG_STAT register.

    In this capture I've set a GPIO to output high whenever the DIAG_STAT 0x8000 error is received from a burst transfer (blue trace). You can see the PPS signal (yellow trace, very faint from the large horizontal scale) is present during all the error periods.

    Zooming in on one of these pulses immediately prior to the error flag being set (with limited resolution from the large capture) it appears to have no spurious artifacts


    I tried injecting an external 1 Hz signal into the SYNC pin (3.3V, positive pulse polarity, 10% duty cycle) however I get the same errors.

  • Hi BHillam, your symptoms match almost exactly to what I am experiencing.

    I am frequently getting clock sync errors when I read the diag stat part of the data burst read.

    My setup:

    ADIS16507

    GNSS module 1Hz PPS -> Sync pin of IMU

    Up scale factor 2000

    Decimation set to give 200 Hz ODR (although I have tried many differentz )

    No filtering set.

    IMU in scaled Sync mode

    32Bit Burst reads

    Noteworthy about my setup is

    1) I communicate with the IMU via a FT4222 (USB to SPI converter)

    2) I am polling the IMU (I do not have data ready connected) (yes I know not ideal!)

        I have tested varying polling rate, and I see the clock sync error in all configs. However i get fewer errors if polling below the output data rate.

    (eg if I configure decimation to give 200 Hz ODR but poll at 100 Hz.)

    I have been able to inspect the 1Hz PPS signal on the oscilloscope and its squeaky clean and stable, as one would expect from a GNSS receiver.

    I have also monitored the DR signal on the oscilloscope. This is reasonably stable, however suffers outages and stutters occasionally

    (eg the min/max statistic shows min 192.3 Hz max 217.4 Hz mean 206.4 Hz)

    (I presume this correlates to when I get a flood of "clock synchronisation" errors, however I haven't correlated the two things so directly as you have with your test where you trigger a GPIO when you see the clock sync error in the diag stat register.

    In a separate side experiment, I switched to direct sync mode, and fed a 2000 Hz signal into the sync pin (my GNSS allows a reconfiguration of the PPS to higher freqs)

    In this mode when I monitor the data ready pin with the oscilloscope it's rock steady to whatever I have configured the ODR to be. I also get not clock sync errors of course, as this error is not applicable in direct sync mode.

    I have also done a little investigation into the differences in the Diag stat... what you get when you read it in a burst and what you get if you do a direct read of the diag stat register only, and have found  that the error bit for "clock sync error" does seem to be more "sticky" in the burst read. Ie it hangs around stuck/latched for longer. However the error bit does eventually clear from burst reads, even if Im not additionally polling the diag stat register directly.

    This sounds like it matches up a bit with what "FormerMember" said ie...
    "Second Querry: Clear error flags in the DIAG_STAT register require specific action, such as writing to a register or power-cycling the device, as burst reading doesn't achieve this."

    Questions to you

    1) Are you triggering your reads based on the data ready signal, or do you poll like I do?

    2) have you managed to make progress on this error?

  • I'm sorry to hear that you are experiencing the same problem.

    1. We are using the data ready signal to trigger an interrupt - no polling.

    2. We did not find any solution to this problem. It's disappointing that Analog never treated this issue as important enough to warrant deeper investigation.