I did peak detection tests using a drop impact tester.
Setting
ODR: 6400Hz
HPF_CORNER :11
1. Configure the FIFO_CTL register for peak detect mode(b0011100X to Register 0x3A).
2. Set the desired activity threshold and time settings (Register0x23 to Register 0x29).
3. Set the desired inactivity threshold and time settings (Register0x2A to Register 0x31).
4. Set the activity mode to linked or loop mode (Register 0x3E).
Result
1.duration of shock pulse:20ms
Acceleration 372peakdet
value Value
17.64G 0G @activity threshold:1G
17.55G 0G @activity threshold:3G
17.76G 16.1G @activity threshold:5G
17.61G 13.4G @activity threshold:10G
17.67G 0G @activity threshold:15G
2.duration of shock pulse:10ms
Acceleration 372peakdet
value Value
57.60G 0G @activity threshold:1G
57.51G 0G @activity threshold:3G
57.76G 57.8G @activity threshold:5G
57.65G 58.9G @activity threshold:10G
57.75G 58.1G @activity threshold:15G
Refer:https://www.lansmont.com/whats-waveform-part-2
Q1.The longer the duration of the shock pulse and the lower the shock acceleration, the more unstable the peak detection becomes.
Can you give any advice on this?
Q2.Peak detection operation does not occur when the threshold installation is 1G or 3G. Can you give any advice on this,also?
Best regards,
Yukio