Question about NULL_CNFG for IMU device

Thread Summary

The user asked about the function of NULL_CNFG for IMU devices and its availability on the ADIS16480. The engineer clarified that the ADIS16480 uses the Tare command in the GLOB_CMD register for auto-nulling, which orients the IMU's axes to East/North/Up (ENU) and sets pitch and roll to 0 degrees. The NULL_CFG in Table 3 refers to the stall time required for the Tare command. Setting the Body Frame bit in the EKF_CONFIG register to 1 allows viewing accelerometer readings relative to the IMU's orientation.
AI Generated Content

I have some questions about NULL_CNFG(Continous bias estimation) for IMU devices.

#1, what's the function of NULL_CNFG? If you have some material to explain this function, please let me know.

#2, About ADIS16480, Table3 says mention “NULL_CNFG” stall times, but I can’t find “NULL_CNFG" on the register of USE REGISTERS map.
Is it possible to use NULL_CFG on ADIS16480?
(It looks like ADIS16489 has the NULL_CFG register on the user reg map)

https://www.analog.com/media/en/technical-documentation/data-sheets/ADIS16480.pdf

Best,

Tarzan2