Hello.
I would like to know something about the offset register.
I use a translator, so I write the minutes separated by as short a sentence as possible.
The calibration method using the offset register does not work.
I have carried out the Calibration of the WATTOS Registers on page 8 of AN-1076.
First, the LCYCMODE register was set to 0x09 in order to apply only to phase A.
"LINE CYCLE MODE" was used to measure the AWATTHR.
Systematically the following conditions were used
fl = 50 Hz
---------------
VN = 200 V
IN = 0.7227A
WTHRN = 2645190203 (THR0 = 11167291, THR1 = 157)
LINECYCN = 1000
AWATTHRN = 346
---------------
Vmin = 40V
Imin = 0.1449A
WTHRmin = PMAX = 33516139 (THR0 = 16738923, THR1 = 1)
LINECYCmin = 1000
AWATTHRmin = 1065
---------------
↓
AWATTOS = 1153599
After measuring AWATTOS, the power was returned to normal accumulated power mode and the THR register was also restored to its original state. (power analyser was used to compare the accumulated power)
After setting AWATTOS, the error in accumulated active power was measured and the error was about 345% at VN=40V. (Before setting AWATTOS, the error in accumulated active power was around 0.2% when VN = 200 V and around 3% when VN = 40 V.)
I think the AWATTOS setting is wrong, but I could not understand what is wrong.
Could you please tell me what is wrong?
As an aside, before setting xIRMSOS and xVRMSOS, the error in accumulated apparent power was about 3% at VN=40V, but by setting xIRMSOS and xVRMSOS, it improved to 0.01% even at VN=40V.
The accumulated active power error remains large because the AWATTOS setting does not work.
Please help us.