Getting wrong values while measuring capacitance as Device under Test (DUT) with Resistor as calibration impedance

Thread Summary

The user is experiencing issues with the AD5933 impedance analyzer when measuring capacitors at low frequencies, leading to incorrect phase angles and capacitance values. The final answer suggests that the ADC is not capturing a full signal cycle due to the low MCLK frequency, and recommends increasing MCLK to at least 16*1024*fmin to achieve better accuracy. The user should also consider using a band-pass filter or a summing amplifier to manage DC offsets and noise, especially for measurements below 1 kHz.
AI Generated Content
Category: Hardware
Product Number: AD5933

Hi All,

I am literally struggling to understand the behaviour of AD5933 when I am trying to measure capacitor as Device under Test or Zunknown.

I have measured 47pF of capacitor with 

Fmin-30Khz, Vac=200mV RFB,Zcal=220Kohm,external oscillator where I got the phase around -84degrees and Cap value as 53pF. Is it acceptable? Similarly I have measured 10pF of Zunknown and found Cap of 13pF and phase of -84degrees.

I have also tried to measure the capacitance of 1microFarad with same Rfb and Rcal but getting wrong results and phase.

I also tried to measure 47pF of cap at low frequencies from 300Hz but again getting wrong results.

   

Any kind of help would be appreciable.

Thanks in advance.