Hello,
We develop a bioimpedance device using the AD5933 and we would like to replace it with AD5934.
How the lower sample rate of 250Ksps of AD5934, compare to AD5933 with 1Msps effect on the device resolution,accuracy and measurment time?
The data sheet of the AD5934 is not clear about the effect of the difference in the sample rate on the device performance.
We are using the AD5933 for sweep bioimpedance measurment across 50 frequncies from 1Khz to 100Khz and 4 electrode measurment type.
Looking forward to your answer,
Best regard,
gal
charlyelkhoury - Moved from Boston Engineering to Precision ADCs. Post date updated from Tuesday, July 9, 2024 6:52 AM UTC to Tuesday, July 9, 2024 1:39 PM UTC to reflect the move.
dendalisay - Moved from Precision ADCs to Direct Digital Synthesis (DDS). Post date updated from Tuesday, July 9, 2024 1:39 PM UTC to Wednesday, July 10, 2024 6:16 AM UTC to reflect the move.
There is some impact on impedance accuracy when the sampling rate is reduced but it is difficult to quantify since it depends on multiple factors, including the nature of the impedance being measured.
I am attaching a simulation model that includes the ADC and the DSP so that you can evaluate the degradation in your particular use case.
LTSpice does not have a way to display numerical results, so the outcome of the formulas that emulate the DSP can be seen in the Error Log file.
re: INTEG(v(imeas)*v(sin))=9.81017e-005 FROM 0 TO 0.0002 yreal: (2*re*freq)/(n*gain*rfb)=0.000981017 im: INTEG(v(imeas)*v(cos))=3.58174e-006 FROM 0 TO 0.0002 yim: (2*im*freq)/(n*gain*rfb)=3.58174e-005 zmod: 1/sqrt(yreal**2+yim**2)=1018.67 phi: atan(yim/yreal)=2.09097 zre: zmod*cos(phi)=1017.99 zim: zmod*sin(phi)=37.1674
Hi,
Circuit: * C:\Users\galya\Downloads\EVAL-AD5933 with ADC.asc WARNING: Node N008 is floating. WARNING: Less than two connections to node N008. This node is used by A1. WARNING: Less than two connections to node COS. This node is used by V5. WARNING: Less than two connections to node IMEAS. This node is used by E3. Instance "m:u2:6": Length shorter than recommended for a level 2 MOSFET. Instance "m:u2:5": Length shorter than recommended for a level 2 MOSFET. Instance "m:u2:4": Length shorter than recommended for a level 2 MOSFET. Instance "m:u2:3": Length shorter than recommended for a level 2 MOSFET. Instance "m:u2:2": Length shorter than recommended for a level 2 MOSFET. Instance "m:u2:1": Length shorter than recommended for a level 2 MOSFET. Instance "m:u1:6": Length shorter than recommended for a level 2 MOSFET. Instance "m:u1:5": Length shorter than recommended for a level 2 MOSFET. Instance "m:u1:4": Length shorter than recommended for a level 2 MOSFET. Instance "m:u1:3": Length shorter than recommended for a level 2 MOSFET. Instance "m:u1:2": Length shorter than recommended for a level 2 MOSFET. Instance "m:u1:1": Length shorter than recommended for a level 2 MOSFET. Direct Newton iteration failed to find .op point. (Use ".option noopiter" to skip.) Starting Gmin stepping Increasing initial diagonal Gmin to 100 Increasing initial diagonal Gmin to 1000 Gmin = 1000 Gmin = 107.374 Gmin = 11.5292 Gmin = 1.23794 Gmin = 0.132923 Gmin = 0.0142725 Gmin = 0.0015325 Gmin = 0.00016455 Gmin = 1.76685e-005 vernier = 0.5 vernier = 0.25 vernier = 0.125 vernier = 0.0625 vernier = 0.03125 vernier = 0.015625 vernier = 0.0078125 Gmin = 1.13078e-005 vernier = 0.00390625 vernier = 0.00195313 vernier = 0.000976563 vernier = 0.000488281 Gmin = 0 Gmin stepping failed Starting source stepping with srcstepmethod=0 Could not converge to DC with sources off! Starting source stepping with srcstepmethod=1 Source Step = 3.0303% vernier = 0.25 Source Step = 7.57576% Source Step = 15.1515% Source Step = 22.7273% Source Step = 30.303% Source Step = 37.8788% vernier = 0.0625 Source Step = 37.1419% Source stepping failed Pseudo transient terminated at 102.851 s, solution accepted ready or not! re: INTEG(v(imeas)*v(sin))=-4.85799e-011 FROM 0 TO 0.0002 yreal: (2*re*freq)/(n*gain*rfb)=-4.85799e-010 im: INTEG(v(imeas)*v(cos))=7.34091e-011 FROM 0 TO 0.0002 yim: (2*im*freq)/(n*gain*rfb)=7.34091e-010 zmod: 1/sqrt(yreal**2+yim**2)=1.136e+009 phi: atan(yim/yreal)=-56.5046 zre: zmod*cos(phi)=6.26926e+008 zim: zmod*sin(phi)=-9.47348e+008 Date: Thu Jul 11 09:03:33 2024 Total elapsed time: 8.321 seconds. tnom = 27 temp = 27 method = modified trap totiter = 61738 traniter = 46503 tranpoints = 23251 accept = 18907 rejected = 4560 matrix size = 315 fillins = 552 solver = Normal Thread vector: 105.9/45.5[8] 19.8/10.7[8] 15.4/8.2[7] 3.1/20.3[1] 2592/500 Matrix Compiler1: 37.73 KB object code size 33.2/15.9/[4.6] Matrix Compiler2: 37.53 KB object code size 14.6/14.7/[6.0]
Thank you for answering.
Is there delay in the measurment time comparing to AD5933?,because of the change in the ADC sampling rate?
I could not run the stimulation in the LTspice software, error log is attached.
Best regards,
Gal
Yes, the AD9834 takes longer to perform the measurement than the AD9833.
Yes, the AD9834 takes longer to perform the measurement than the AD9833.