Three electrode measurement for AD5933?

Thread Summary

The user is using the AD5933 for bioelectrical electrode impedance measurements in a three-electrode configuration. The issue was discontinuities in the Nyquist plot and large errors at low frequencies. The solution involved recalculating the MCLK for each frequency, ensuring the excitation signal was within the ADC range, and adjusting the high-pass filter cutoff frequency to 0.14 Hz. This resulted in continuous and more accurate impedance measurements.
AI Generated Content

Hi everybody,

 

we're using the AD5933 for bioelectrical electrode impedance measurements. Usually, the electrodes have to be immersed in a saline solution for measuring, and a 2 point measurement (Electrode to Reference electrode) adds unwanted potential offsets to the measurement. For a more accurate measurement, a technique using 3 electrodes is employed:

 

(Fig. 4 from this link)

Ralph Group : projects : Fabrication and characterization of graphene electrodes for electrochemistry experiments

Figure 4

Is there any way to connect the AD5933 in such an arrangement to measure impedance? Do you have any additional suggestions for performing such a measurement?

 

Thanks and Kindest Regards,

Oscar Cota

Parents
  • The RFB is 20k ohms. I'm using the AFE below and in practice the Reference and Counter are not shorted together, but connected to electrodes in the biomarker solution like the examples Oscar gave above. 

    Edit:

    The red boxes which I tried to add in Paint are multiplexers and the red thing between them in the calibration resistor.

    For measuring the impedance, I start at 100 kHz and measure the calibration resistor and then the biomarker solution, then I manually step the frequency down to 63 kHz and measure the calibration resistor and then the biomarker solution, and I keep decreasing the frequency until 1 Hz. Then at the end of the sweep, the code calculates the gain factor and the actual real and imaginary values for the solution.

Reply
  • The RFB is 20k ohms. I'm using the AFE below and in practice the Reference and Counter are not shorted together, but connected to electrodes in the biomarker solution like the examples Oscar gave above. 

    Edit:

    The red boxes which I tried to add in Paint are multiplexers and the red thing between them in the calibration resistor.

    For measuring the impedance, I start at 100 kHz and measure the calibration resistor and then the biomarker solution, then I manually step the frequency down to 63 kHz and measure the calibration resistor and then the biomarker solution, and I keep decreasing the frequency until 1 Hz. Then at the end of the sweep, the code calculates the gain factor and the actual real and imaginary values for the solution.