We are experiencing issues with results in cyclic voltammetry with AD5940 evaluation kit. When we test it with extermal impedance, we get correct voltages and currents. But when we test it with an electrode with a solution, we can´t get good results. VRE0 - VSE0 varies from 200mV to 1000mV, but reads almost 0 A every time. In this kit, VZERO (VSE or VWE) is being used fixed to 1.3V and VBIAS (VCE) varies from 1500 to 2300 to get VRE-VSE 200 to 1000
We have made a cyclic voltametry to same electrode and solution with a commercial potentiostat. Corresponding current was 1 to 9uA. We measured VCE and it was a reference potential of 0 V and voltage varied from 1500 to 2300 in VWE. Just the opposite to the Analog kit.
Would it be possible to use the analog kit with fixed VCE and variable VWE, just like in the commercial potentiostat?