AD7124-8 : Internal and System calibration with external reference
The user inquires about the AD7124-8's calibration procedures in a 4-wire RTD setup with an external reference. The final answer confirms that internal calibration results are automatically applied to the data register, internal calibration can be performed with an external reference, and system calibration requires applying zero-scale and full-scale voltages to AIN2/AIN3, considering the reference voltage through RREF. Key parameters: full-power mode, FS = 200, SINC4 filter, bipolar mode, external reference, gain = 8, excitation current = 250 μA, Rref = 10 kΩ, Rheadroom ≈ 400 Ω.
Product Number: AD7124-8
I am working with the AD7124-8 in a 4-wire RTD measurement setup and need clarification on how internal and system calibration behave when using an external reference.
My Setup:
  • AD7124-8 in full-power mode
  • FS value = 200, SINC4 filter
  • Bipolar mode, all input and reference buffers enabled
  • External reference selected
  • Gain = 8
  • Excitation current = 250 µA
  • 4-wire RTD:
    • Excitation on AIN0
    • RTD connected between AIN2 (P) and AIN3 (N)
  • Reference resistor (Rref) = 10 kΩ
    • Lead resistor (Rheadroom) ≈ 400 Ω
  • Using ratiometric calculations to derive resistance from ADC code.
Calibration Procedure I Used:
After an ADC reset, I configure the device as above and then perform:
  1. Internal zero-scale (offset) calibration
  2. Internal full-scale (gain) calibration
Both were executed according to the datasheet procedure.
My Questions:
  1. After performing internal gain and offset calibration, are the ADC conversion results available in the data register adjusted with the offset and gain calibration registers? Or do I need to apply any correction in my firmware?
  2. Is it valid to perform internal calibration while external reference is selected in the configuration register ?, or Do we need to perform internal calibration with internal reference configured ?
  3. Regarding system calibration:
  4. For system zero-scale and full-scale calibration, what should be applied to the AIN2/AIN3 analog inputs in this configuration?
    • It is mentioned that full-scale input should be provided while performing system full-scale calibration. As we are using IOUT as external reference , do we need to consider this while feeding voltage for system calibration ?
I want to ensure I am calibrating the AD7124-8 correctly for accurate RTD measurement. Any guidance or recommended best practices would be very helpful.
  • Hi
      - You do not need to apply any correction on the firmware, it is already updated on the data register.
      - Internal full scale calibration should work just fine on using whether internal reference or external reference.
    3. For system zero-scale and full-scale calibration, what should be applied to the AIN2/AIN3 analog inputs in this configuration?
      - System calibrations expect the system zero-scale (offset) and system full-scale (gain) voltages to be applied to the ADC pins before initiating the calibration modes.
    4. It is mentioned that full-scale input should be provided while performing system full-scale calibration. As we are using IOUT as external reference, do we need to consider this while feeding voltage for system calibration?
      - Yes, the reference voltage through RREF should be considered for system calibrations
    Thanks and regards,
    Rod
  • Hi  ,
    Thanks for the quick response.
    I have a few additional questions regarding the system zero-scale and full-scale calibration steps.
    From your explanation, system calibration requires applying the system’s actual zero-scale (offset) and full-scale (gain) voltages to the ADC inputs before triggering the calibration modes.
    1. System Zero-Scale Calibration:
      Am I correct in assuming that for zero-scale calibration, I should short AIN2 and AIN3 to GND so that the ADC sees 0 V at the differential input?
    2. System Full-Scale Calibration:
      For full-scale calibration, since the PT1000 is normally driven by the excitation current, I assume I need to disconnect the RTD and apply a known resistor value instead.
      However, because my reference is generated ratiometrically using the excitation current and Rref, how should I correctly apply the full-scale voltage?
      • The REFOUT from the eval board measures around 2.495 V, not exactly 2.5 V.
      • If I apply an external voltage source for the full-scale calibration, will this introduce errors or conflict with the ratiometric setup?
  • Could someone please provide an update when convenient? I would appreciate any guidance or clarification.
  • Hi
    • Yes, you are correct
    • Ideally, if you want the most accurate system calibrations, actual system measurement values should be used. Applying external voltage on your system for calibration purposes may introduce additional errors on your overall system.
    Thanks and regards,
    Rod