Hi,
I am considering developing a temperature measurement environment using LTM2985.
I connected Linduino board (DC2026C) with a cable, LTM2985 demo board (DC2872A), and thermocouple demo board (DC2212A), and am measuring the ambient temperature (approximately 23°C) using 4-terminal method with RTD (PT-100) installed on DC2212A.
But the measured values vary by approximately 0.1°C.
Looking at the RMS noise performance of the RTD PT-100 shown on page.9 of LTM2985 datasheet, it is nearly 0°C.
Even if we multiply it by 6.6 to predict P-P, we estimate that the product performance in terms of variation is far less than 0.1°C. What considerations should be made to obtain measurement quality equivalent to the product performance shown in the datasheet? Mounting the RTD near the IC rather than on the demo board ?
averaging the measurement values, etc.
Thanks and regards,
kanatko