SYNC_IN to first DRDY time

Category: Hardware
Product Number: AD7779

Hello!

I use the AD7779 ADC for a new hardware design.

I need to know how the SYNC_IN to first DRDY time is calculated.

I use an FPGA to read the samples from the DOUT interface.

The SYNC_IN pulse timing for the ADC is generated from an FPGA internal 250 us synchronization pulse. The ADC sampling shall be synchronized to that pulse.

 So, the DRDY pulse shall be synchronous to our 250 us sync pulse. We want to monitor the synchronous ADC sampling permanently by comparing DRDY with the 250 us sync pulse.

Q1: Can the SYNC_IN to DRDY time be calculated by a formula?

Q2: Is the SYNC_IN to first DRDY time stable after every Power on / ADC Reset / Reconfiguration?

The DS shows a minimum time of 145 us and a settling time of the sinc3 filter, but no more details.

My ADC settings: High resolution mode, MCLK=8.192 MHz, Decimation Rate 128, ODR = 16 kSps

Furthermore we use the CHx_SYNC_OFFSET registers to delay the ADC channels individually.

All 8 channels get different values > 0.

We would like to delay all channels with fixed values. No channel has a zero delay.

The problem is, the SYNC_IN to first DRDY time changes as soon a different value is written to one of the CHx_SYNC_OFFSET register.

The greater the value in the CHx_SYNC_OFFSET registers, the bigger is the SYNC_IN to first DRDY time. But this delays all samples additionally.

Q3: How do the CHx_SYNC_OFFSET registers influence the SYNC_IN to first DRDY time?

Q4: How to delay the first 4 channels with a fixed value for example 20 us and the other 4 channels with 50 us?

best regards,

Sebastian

Parents
  • Hi  ,

    Please refer to the answers below and let me know if you have other questions.

    Q1.The SYNC_IN to DRDY time is computed as 3/ODR for sinc3 and 5/ODR for sinc5

    I am in the process of correcting the datasheet to 188 uS for this value. Please refer to the calculation below

    for ODR = 16kSPS

    tinit = 3/(16k) = 188 uS

    Q2. Yes. This time should be relatively stable.

    Q3. The sync signal is delayed by the value written in the CHx_SYNC_OFFSET registers 

    is added to the 3 ODR settling time

    Q4. Typically, these registers are used to match the phase of multiple channels and compensate for delays. The channel with the biggest delay is used as the reference and the difference from the other channels is written to the registers as discussed in these links

    AD777x: What is the best phase offset calibration process? - Documents - Precision ADCs - EngineerZone

    AD7771 Phase Adjustment Question - Q&A - Precision ADCs - EngineerZone

    ad777x phase adjustment - Q&A - Precision ADCs - EngineerZone

    To add a phase delay per channel, please refer to the steps below.

    ODR = 16kSPS; dec_rate = 128 MOD Clk = 2.048MHz

    Compute for the delay in Mod cycles

    delay = 20uS/(1/2.048MHz) = 20uS * 2.048 MHz = 40.96 cycles ~= 41 cycles

    Register value = (dec_rate - mod cycles)/Phase Adjustment compensation = (128-41)/1 = 87 = 0x57

    for delay of 50uS

    delay = 50uS * 2.048 MHz = 102.4 ~= 102 cycles

    Register Value = (128-102)/1 = 26 = 0X1A

    Thanks,
    Janine

  • Hello Janine!

    Thank you for your fast reply.

    To Q1:

    I have measured the time between rising edge of SYNC_IN and the rising edge of the first DRDY pulse. I always measure 142.2 us with my ADC settings (MCLK=8.192 MHz, Decimation Rate 128, ODR=16 kSps, High Resolution Mode, all CHx_SYNC_OFFSET unwritten).

    If I write the value 127 to all CHx_SYNC_OFFSET registers, the time changes from 142 us to 204 us. This is exactly the 62 us delay I expect.

    But how to get 3/ODR = 188 us? 

    to Q4:

    As I understand your answer, there is no chance to delay ALL channels by given delay values in the CHx_SYNC_OFFSET registers. The channels can only be delayed relative to each other.The one channel (for example CH0) with the biggest delay remains undelayed (CH0_SYNC_OFFSET=0), the other channels will be delayed until they match CH0?

    thanks a lot for your support

    Sebastian

  • Hi  ,

    Please refer to the answers below.

    Q1. 142 uS is correct. I apologize for the confusion. 188 uS is the estimated settling time of the sinc3 filter.

    If you are able to get the 62 uS, the offset register is indeed working as expected.

    Q4. There is a chance to delay all channels by using the offset registers, but this is not the usual case for the phase offset adjustment.

    The usual use case for this is to use one channel with the highest delay and set it as the reference and delay the other channels to maintain phase coherence between the channels.

    I hope I have clarified your confusion. Let me know if there is anything else.

    Thanks,
    Janine

Reply
  • Hi  ,

    Please refer to the answers below.

    Q1. 142 uS is correct. I apologize for the confusion. 188 uS is the estimated settling time of the sinc3 filter.

    If you are able to get the 62 uS, the offset register is indeed working as expected.

    Q4. There is a chance to delay all channels by using the offset registers, but this is not the usual case for the phase offset adjustment.

    The usual use case for this is to use one channel with the highest delay and set it as the reference and delay the other channels to maintain phase coherence between the channels.

    I hope I have clarified your confusion. Let me know if there is anything else.

    Thanks,
    Janine