Hi Teams:
I used AD7124-8 TEST PT100,But the test resut is only 50ohm?
AD7124-8
Recommended for New Designs
The AD7124-8 is a low power, low noise, completely integrated
analog front end for high precision measurement applications. The
AD7124-8 W grade is AEC...
Datasheet
AD7124-8 on Analog.com
Hi Teams:
Hi ZFJC ,
Can you share your schematic for this?
Have you also checked the reference design for RTD measurement system using AD7124? If not you may refer to the link below:
CN0383 Circuit Note | Analog Devices
The reference designs are engineered and tested for quick and easy system integration to help solve today’s analog, mixed-signal, and RF design challenges. This circuit note provided above uses a Class B Pt100 RTD sensor with an accuracy of ±0.3°C at 0°C but it can support other classes such as Class A, Class AA, 1/3 DIN, or 1/10 DIN that are higher accuracy RTDs. You may refer to the circuit note for better understanding on how to setup your PT100 using AD7124.