Hi,
I have a AD5940 Electrochemical Eval Board, and I am testing it with the Square Wave Voltammetry example provided analogdevicesinc/ad5940-examples: AD594x related application examples and block level examples. (github.com). I configured the LPDAC switches(LPDACSW0[1], LPDACSW0[2],LPDACSW0[3],LPDACSW0[4]) and LPTIA switches (SW2,SW4,SW5,SW6,SW7,SW8,SW9,SW12) to enable the path for taking a measurement on Vbias0 and Vzero0 as shown in Figure 22 of the AD5940 datasheet. According to my understanding, If I configure RampStartVolt = RampPeakVolt = 0 mV, VzeroStart = VzeroPeak = 1300 mV, then measurement at the point of Vzero0 (C12 on the board ) and Vbias0 (C11 on the board) should be 1300 mV, however, what I measured at both points is 800 mV. Just wondering if my understanding was wrong or is there other explanation why the measured voltage is different from the desired voltage at the two points?
Thank,
Olive