I'm currently working on a project that involves using the AD7771 delta-sigma ADC. While the SPI interface seems to be functioning properly, I'm encountering an issue with the received data. Upon measuring the voltage output on the REF_OUT pin, I noticed it is at -1.65V, which matches the AVSS voltage level. I have thoroughly checked all the power supplies and couldn't identify any apparent problems.
I have an external reference (2.5V), but it's referenced to the GND, so i can't use it.
I would greatly appreciate any assistance.
Thank you in advance for your support.
HP
Top Replies
    Jun 14, 2023 +1 verified
    Hi  ,
    While it is true that the potential at REF_OUT with respect to GND at power-up is -1.65V, this is not the correct connection for the reference. REF_OUT matches AVSS since the internal reference…
Parents
  • Hi  ,
    While it is true that the potential at REF_OUT with respect to GND at power-up is -1.65V, this is not the correct connection for the reference. REF_OUT matches AVSS since the internal reference output is disabled at start-up. To enable the output, you can configure the PDB_REFOUT_BUF register to 1.
    Furthermore, if you want to use this as your reference, you can configure ADC_MUX_CONFIG bits to 01 as per the table below:
    Remember to use 100 nF bypass caps.
    And if you want to further improve its noise performance, you can connect a low-pass filter with a cut-off frequency of 10 Hz, and buffer it up again by:
    1. Connecting REF_OUT to REFx+
    2. Connecting AVSSx to REFx-
    3. Configuring REF_MUX_CTRL to use external reference i.e. ADC_MUX_CONFIG bits to 00.
    For more details, you can check the Internal Reference and Reference Buffers section in the datasheet.
    AD7771 (Rev. A) (analog.com)
    Yugel
  • Hey Yugel,
    I wanted to express my gratitude for your assistance. I followed your advice, and it worked! Thank you.
    HP
  • Hello  ,
    Glad to be of help. Don't hesitate to post new threads whenever you need help in our products. This thread will be closed now.
    Yugel