There is no mention of this terminal in the data sheet table 7, but from FIg68 is it ``V- - 0.7 V to V+ + 0.7 V''?
Edit Notes
add Fig[edited by: UBOTI at 8:54 AM (GMT -4) on 22 Mar 2023]
Strong EMC design
ADAQ7980
Recommended for New Designs
The ADAQ7980/ADAQ7988 are 16-bit analog-to-digital converter
(ADC) μModule® data acquisition systems that integrate four common signal processing and...
Datasheet
ADAQ7980 on Analog.com
Hi UBOTI ,
We will look into this. I will contact the product owner and get back to you.
JC
Hi UBOTI
You are correct. Just make sure to follow the conditions stated for PD_REF in the Theory of Operation section.
If there are other concerns, please feel free to reach out.
Best Regards,
Regine
Thank you for answering.
Please tell me one more thing.
I would like to connect the LT6654 externally to REFOUT without using the REFBUFF of the ADAQ7980, referring to the datasheet FIG69.
With PD_REF = V-, REF = GND and connecting the LT6654 to REFOUT,
The LT6654 output voltage will drop to approximately 2.5V.
(When the connection is cut off, the LT6654 normally outputs 5V)
When I connected the LT6654 to REFOUT with PD_REF = V- and REF = REFOUT, I was able to apply 5V normally.
Q When connecting the reference to REFOUT externally without using the REFBUFF of the ADAQ7980, what is the appropriate way to connect the REF pin?
Hi UBOTI ,
Apologies it took a while to get back to you.
We strongly recommend using the reference buffer circuit, even if you are using an external reference (in this case, LT6654). This is to ensure that we have sufficient current to drive the ADC. To do this, you may just simply apply REF = 5V.
The same concept goes if you are using multiple ADAQ7980 devices. Apply 5V to the REF pin of the first device, then proceed with the connections as illustrated below (with the REF_OUT pins connected together in the succeeding stages).
You may then check through the REF_OUT pins if we are seeing the same voltage that we applied at the REF pin.
Hope this helps!
Thanks and regards,
Chelsea
If REF=GND, 5V cannot be applied to REFOUT from the outside (cannot be applied according to FIG.69).
Isn't there a protection Di between REF and REFOUT?
(It was possible to supply 5V from the outside by setting REF=REFOUT)
Hi UBOTI ,
Apologies for the delay in response. Sending the reply that my colleague sent:
In the case, where the REF pin (non-inverting input) of the disabled internal reference buffer is tied to GND, the current is nearly 38mA that exceeds what the LT6654 can source or sink (+/- 10mA). So it would appear that a viable solution is to connect the REF pin to VCC (5V) or to the same potential as the externally applied reference like the LT6654.
Chelsea