Could not get expected result using AD5940 with demo application provided
Product Number: AD5940, Square Wave Voltammetry example, SensorPal
We purchased two AD5940 Eval boards and tried to use them in an electrochemical sensing application.
To get familiar with the AD5940 device and verify how well it performs in our application, we conducted an experiment using square wave voltammetry in three approaches. a) use a commercial potentiostat (CH); b). Use the Analog Device SensorPal APP; c). Use the Square Wave Voltammetry example code provided here: https://github.com/analogdevicesinc/ad5940-examples
The experiment settings include a glassy carbon electrode, a platinum electrode and a Ag/AgCl electrode, worked as working, counter, reference; 10 mM hexacyanoferrate(ii), PBS 1X.
The configurations applied in the three approaches are shown below:
The results generated from the three approaches are shown below:
From the results shown above, when plotting the current vs index with the data generated from the Square Wave voltammetry example code with the keil project, we do not manage to see a redox peak as we saw from the other approaches.
Can you please explain why it is like this? What might be wrong with what we did in the experiment?
Another question regarding the result generated from SensorPal:
 although a redox peak is observed during Square wave voltammetry in the SensorPal, it seems that the peak potential is shifted towards the negative values. Further testing revealed that applying a *-1 factor to both current and potentials gives a similar results as the one obtained from the commercial instrument (CH). Can you please explain why there is such a shift and how to correct it to make sure the visualized potential corresponds to the applied one?
Olive
    May I know if you are using the latest Square wave voltammetry code in Keil (attached below for your reference)?
    @file:    SWVTest.c
    @author:  $Author: mlambe $
    @brief:   Square Wave Voltammetry measurement sequences.
    @date:    Updated 15th May 2021
    Copyright (c) 2017-2019 Analog Devices, Inc. All Rights Reserved.
    This software is proprietary to Analog Devices, Inc. and its licensors.
    By using this software you agree to the terms of the associated
    Analog Devices Software License Agreement.
    /** @addtogroup AD5940_System_Examples
    *    @defgroup Ramp_Test_Example
    *    @brief Using sequencer to generate ramp signal and control ADC to sample data.
    *    @details
    * @note Need to update code when runs at S2 silicon.
    * @todo update LPDAC switch settings for S2 and LPDAC 1LSB bug.
    * @todo Calibrate ADC/PGA firstly to get accurate current. (Voltage/Rtia = Current)
    * @note The method to calculate LPDAC ouput voltage
    *        - #define LSB_DAC12BIT (2.2V/4095)
    *        - #define LSB_DAC6BIT  (2.2V/4095*64)
    *        - Volt_12bit = Code12Bit*LSB_DAC12BIT + 0.2V
    *        - Volt_6bit = Code6Bit*LSB_DAC6BIT + 0.2V
    * # Ramp Signal Parameters definition
    * @code
    * (Vbias - Vzero):
    *     RampPeakVolt   -->            /
    *    RampStartVolt   -->   /
    * Vzero: If there is no limitation on Vzero, Set VzeroStart to 2.2 and VzeroPeak to 0.4V
    * Voltage VzeroStart -->  ______
    * Voltage VzeroPeak  -->        |_____
    * Vbias: Vbias is calculated from RampPeakVolt, RampStartVolt, VzeroStart and VzeroPeak.
    * Voltage VbiasPeak  -->       /|    /
    * Voltage VbiasStart -->   /    |/
    * RampState define:    S0 | S1  | S2 |
    * @endcode
    *  # The sequencer method to do Ramp test.
    * The Ramp test need to update DAC data in real time to generate required waveform, and control ADC to start sample data. \n
    * We used two kinds of sequence to realize it. One is to control DAC where SEQ0 and SEQ1 are used, another sequence SEQ2 controls ADC.
    *  ## Sequence Allocation
    * SEQ3 is used to initialize AD5940.\n
    * SEQ0/1 is used to generate voltage step.\n
    * SEQ2 is used to startup ADC to sample one point.
    * |SRAM allocation|||
    * |SequenceID  | AddressRange   |  Usage  |
    * |SEQID_3     | 0x0000-0xzzzz  | Initialization sequence|
    * |SEQID_2     | 0xzzzz-0xyyyy  | ADC control sequence, run this sequence will get one ADC result|
    * |SEQID_0/1   | 0xyyyy-end     | DAC update sequence. If size if not enough for all steps, use it like a FIFO.|
    * Where 0xzzzz equals to SEQ3 length, 0xyyyy equals to sum of SEQ2 and SEQ3 length.
    * In one word, put SEQ2 commands right after SEQ3. Don't waste any SRAM resource.
    *  ##Sequencer Running Order
    * The sequencer running order is set to firstly update DAC then start ADC. Repeat this process until all waveform generated.
    * Below is explanation of sequencer running order.
    * @code
    * DAC voltage changes with sequencer, assume each step is 0.05V start from 0.2V
    * 400mV->                             _______
    * 350mV->                  _____     |
    * 300mV->       _____     |     |____|
    * 250mV->      |     |____|
    * 200mV->    __|
    * Update DAC:  ?     ?    ?     ?    ?       ?       -No update
    *            SEQ0  SEQ1  SEQ0  SEQ1 SEQ0    SEQ1    SEQ0
    *                 SEQ2  SEQ2 SEQ2  SEQ2    SEQ2    SEQ2   |The final sequence is set to disable sequencer
    * WuptTrigger  ?  ?    ?  ?    ?  ?    ?  ?    ?  ?    ?  ?    ?  ?    ?  ?
    * Time Spend   |t1| t2 |t1| t2 |t1| t2 |t1| t2 |t1| t2 |t1| t2 |t1| t2 |t1| t2
    *        |The following triggers are ignored because sequencer is disabled
    * Wupt: Wakeup Timer
    * @endcode
    * The final sequence will disable sequencer thus disable the whole measurement. It could be SEQ0 or SEQ1.  \n
    * SEQ2 will always follow SEQ0/SEQ1 to turn on ADC to sample data.                                         \n
    * SEQ0/1 and SEQ2 is managed by wakeup timer. The time delay between SEQ0/1 and SEQ
    * is set by user. Reason is that after updating DAC, signal needs some time to settle before sample it.    \n
    * In above figure, the time t1 is the delay set by user which controls where ADC starts sampling.
    * SEQ2 commands are fixed. Function is simply turn on ADC for a while and turn off it
    *      after required number of data ready.        \n
    * SEQ0/1 is always changing its start address to update DAC with different voltage.                        \n
    * Check above figure we can see SEQ0/SEQ1 is repeatedly trigged by Wakeuptimer, if we don't change the start
    * Address of SEQ0/SEQ1, they will always update DAC with same data, thus no waveform generated.
    * Considering below SEQ0 command which is similar for SEQ1 on modifying SEQxINFO register.:
    * **Sequencer Command Block 1**
    * @code
    * //Total sequence command length is **4**
    * SEQ_WR(REG_AFE_LPDACDAT0, 0x1234);           //update DAC with correct voltage
    * SEQ_WAIT(10);                                //wait 10clocks to allow DAC update
    * SEQ_WR(REG_AFE_SEQ1INFO, NextAddr|SeqLen);   //The next sequence is SEQ1, set it to correct address where stores commands.
    * SEQ_SLP();                                   //Put AFE to hibernate/sleep mode.
    * @endcode
    * It will update DAC with data 0x1234, then it wait 10 clocks to allow LPDAC update.
    * The final command is to send AFE to sleep state.
    * The third commands here is to allow modify sequence infomation by sequencer. Above piece of commands are running by SEQ0.
    * It modify the start address of **SEQ1**. SEQ1 has same ability to update DAC data but with **different** data.
    * By the time Wakeup Timer triggers SEQ1, it will update DAC with correct data.
    * The last block of sequencer command is to disable sequencer.
    * **Sequencer Command Block 2**
    * @code
    * SEQ_NOP();
    * SEQ_NOP();
    * SEQ_NOP();
    * SEQ_STOP();                                   //Put AFE to hibernate/sleep mode.
    * @endcode
    * Total SRAM is 6kB in AD594x. In normal other application, we use 2kB for sequencer and 4kB for FIFO.
    * Assume the ramp test require 128 steps, then the sequence length is 4*128 = 512, each command need 4Byte. So it costs 2kB SRAM.
    * When ramp test requires hundres of voltage steps(ADC samples), 2kB SRAM is far from enough. We recommend to use 4kB for sequencer
    * and 2kB for data FIFO.
    * If ramp test require more steps, then we need to update SRAM with commands dynamically, use it as a ping-pong buffer.
    * **Sequencer Command Block 3**
    * @code
    * SEQ_WAIT(10);
    * SEQ_INT0();                                  //Generate custom interrupt 0 to inform MCU to update ping-pong buffer.
    * @endcode
    #include "ad5940.h"
    #include <stdio.h>
    #include "string.h"
    #include "math.h"
    #include "SqrWaveVoltammetry.h"
    * @brief The ramp application paramters.
    * @details Do not modify following default parameters. Use the function in AD5940Main.c to change it.
    static AppSWVCfg_Type AppSWVCfg =
      .bParaChanged = bFALSE,
      .SeqStartAddr = 0,
      .MaxSeqLen = 0,
      .SeqStartAddrCal = 0,
      .MaxSeqLenCal = 0,
      .LFOSCClkFreq = 32000.0,
      .SysClkFreq = 16000000.0,
      .AdcClkFreq = 16000000.0,
      .RcalVal = 10000.0,
      .ADCRefVolt = 1820.0f,              /* 1.8V or 1.82V? */
      /* Describe Ramp signal */
      .RampStartVolt = -1000.0f,          /* -1V */
      .RampPeakVolt = +1000.0f,           /* +1V */
      .VzeroStart = 2200.0f,              /* 2.2V */
      .VzeroPeak = 400.0f,                /* 0.4V */
      .StepNumber = 866,
      /* Receive path configuration */
      .SampleDelay = 1.0f,                /* 1ms */
      .LPTIARtiaSel = LPTIARTIA_20K,      /* Maximum current decides RTIA value */
      .ExternalRtiaValue = 20000.0f,      /* Optional external RTIA resistore value in Ohm. */
      .AdcPgaGain = ADCPGA_1,
      .ADCSinc3Osr = ADCSINC3OSR_4,
      .FifoThresh = 4,
      /* Priviate parameters */
      .SWVInited = bFALSE,
      .StopRequired = bFALSE,
      .RampState = SWV_STATE0,
      .bFirstDACSeq = bTRUE,
      .bSqrWaveHiLevel = bFALSE,
      /* Configure Square wave */
      .Frequency = 25,              /* Frequency in Hz */
      .SqrWvAmplitude = 25,         /* Square wave amplitude in mV */
      .SqrWvRampIncrement = 5,      /* Ramp increment in mV*/
    * @todo add paramater check.
    * SampleDelay will limited by wakeup timer, check WUPT register value calculation equation below for reference.
    * SampleDelay > 1.0ms is acceptable.
    * @brief This function is provided for upper controllers that want to change
    *        application parameters specially for user defined parameters.
    * @param pCfg: The pointer used to store application configuration structure pointer.
    * @return none.
    AD5940Err AppSWVGetCfg(void *pCfg)
      if(pCfg)
        *(AppSWVCfg_Type**)pCfg = &AppSWVCfg;
        return AD5940ERR_OK;
      return AD5940ERR_PARA;
    * @brief Control application like start, stop.
    * @param Command: The command for this application, select from below paramters
    *        - APPCTRL_START: start the measurement. Note: the ramp test need firstly call function AppSWVInit() every time before start it.
    *        - APPCTRL_STOPNOW: Stop the measurement immediately.
    *        - APPCTRL_STOPSYNC: Stop the measuremnt when current measured data is read back.
    *        - APPCTRL_SHUTDOWN: Stop the measurement immediately and put AFE to shut down mode(turn off LP loop and enter hibernate).
    * @return none.
    AD5940Err AppSWVCtrl(uint32_t Command, void *pPara)
      switch (Command)
      case APPCTRL_START:
          WUPTCfg_Type wupt_cfg;
          if(AD5940_WakeUp(10) > 10)  /* Wakeup AFE by read register, read 10 times at most */
            return AD5940ERR_WAKEUP;  /* Wakeup Failed */
          if(AppSWVCfg.SWVInited == bFALSE)
            return AD5940ERR_APPERROR;
          *  SWV example is special, because the sequence is dynamically generated.
          *  Before 'START' ramp test, call AppSWVInit firstly.
          if(AppSWVCfg.RampState == SWV_STOP)
          /* Start it */
          wupt_cfg.WuptEn = bTRUE;
          wupt_cfg.WuptEndSeq = WUPTENDSEQ_D;
          wupt_cfg.WuptOrder[0] = SEQID_0;
          wupt_cfg.WuptOrder[1] = SEQID_2;
          wupt_cfg.WuptOrder[2] = SEQID_1;
          wupt_cfg.WuptOrder[3] = SEQID_2;
          wupt_cfg.SeqxSleepTime[SEQID_2] = 1;
          wupt_cfg.SeqxWakeupTime[SEQID_2] = (uint32_t)(AppSWVCfg.LFOSCClkFreq*AppSWVCfg.SampleDelay/1000.0f) - 1;
          wupt_cfg.SeqxSleepTime[SEQID_0] = 1;
          wupt_cfg.SeqxWakeupTime[SEQID_0] = (uint32_t)(AppSWVCfg.LFOSCClkFreq*((1/AppSWVCfg.Frequency*500) - AppSWVCfg.SampleDelay)/1000.0f) - 4;
          wupt_cfg.SeqxSleepTime[SEQID_1] = wupt_cfg.SeqxSleepTime[SEQID_0];
          wupt_cfg.SeqxWakeupTime[SEQID_1] = wupt_cfg.SeqxWakeupTime[SEQID_0];
          AD5940_WUPTCfg(&wupt_cfg);
          break;
      case APPCTRL_STOPNOW:
          /* Start Wupt right now */
          AD5940_WUPTCtrl(bFALSE);
          /* There is chance this operation will fail because sequencer could put AFE back
          to hibernate mode just after waking up. Use STOPSYNC is better. */
          break;
      case APPCTRL_STOPSYNC:
          AppSWVCfg.StopRequired = bTRUE;
          break;
      case APPCTRL_SHUTDOWN:
          AppSWVCtrl(APPCTRL_STOPNOW, 0);  /* Stop the measurement if it's running. */
          AD5940_ShutDownS();
        break;
      default:
        break;
    * @brief Generate initialization sequence and write the commands to SRAM.
    * @return return error code.
    static AD5940Err AppSWVSeqInitGen(void)
      AD5940Err error = AD5940ERR_OK;
      const uint32_t *pSeqCmd;
      uint32_t SeqLen;
      AFERefCfg_Type aferef_cfg;
      LPLoopCfg_Type lploop_cfg;
      DSPCfg_Type dsp_cfg;
      /* Start sequence generator here */
      AD5940_SEQGenCtrl(bTRUE);
      AD5940_AFECtrlS(AFECTRL_ALL, bFALSE);  /* Init all to disable state */
      aferef_cfg.HpBandgapEn = bTRUE;
      aferef_cfg.Hp1V1BuffEn = bTRUE;
      aferef_cfg.Hp1V8BuffEn = bTRUE;
      aferef_cfg.Disc1V1Cap = bFALSE;
      aferef_cfg.Disc1V8Cap = bFALSE;
      aferef_cfg.Hp1V8ThemBuff = bFALSE;
      aferef_cfg.Hp1V8Ilimit = bFALSE;
      aferef_cfg.Lp1V1BuffEn = bFALSE;
      aferef_cfg.Lp1V8BuffEn = bFALSE;
      /* LP reference control - turn off them to save power*/
      aferef_cfg.LpBandgapEn = bTRUE;
      aferef_cfg.LpRefBufEn = bTRUE;
      aferef_cfg.LpRefBoostEn = bFALSE;
      AD5940_REFCfgS(&aferef_cfg);
      lploop_cfg.LpAmpCfg.LpAmpSel = LPAMP0;
      lploop_cfg.LpAmpCfg.LpAmpPwrMod = LPAMPPWR_BOOST3;
      lploop_cfg.LpAmpCfg.LpPaPwrEn = bTRUE;
      lploop_cfg.LpAmpCfg.LpTiaPwrEn = bTRUE;
      lploop_cfg.LpAmpCfg.LpTiaRf = LPTIARF_20K;
      lploop_cfg.LpAmpCfg.LpTiaRload = LPTIARLOAD_SHORT;
      lploop_cfg.LpAmpCfg.LpTiaRtia = AppSWVCfg.LPTIARtiaSel;
      if(AppSWVCfg.LPTIARtiaSel == LPTIARTIA_OPEN) /* User want to use external RTIA */
        lploop_cfg.LpAmpCfg.LpTiaSW = LPTIASW(13)|LPTIASW(2)|LPTIASW(4)|LPTIASW(5)|LPTIASW(9)/*|LPTIASW(10)*/; /* SW5/9 is closed to support external RTIA resistor */
      else
        lploop_cfg.LpAmpCfg.LpTiaSW = /*LPTIASW(13)|*/LPTIASW(2)|LPTIASW(4);
      lploop_cfg.LpDacCfg.LpdacSel = LPDAC0;
    	lploop_cfg.LpDacCfg.DacData6Bit = (uint32_t)((AppSWVCfg.VzeroStart - 200.0f)/DAC6BITVOLT_1LSB);
      lploop_cfg.LpDacCfg.DacData12Bit = (int32_t)((AppSWVCfg.RampStartVolt)/DAC12BITVOLT_1LSB) + lploop_cfg.LpDacCfg.DacData6Bit*64  ;
      lploop_cfg.LpDacCfg.DataRst = bFALSE;
      lploop_cfg.LpDacCfg.LpDacSW = LPDACSW_VBIAS2LPPA/*|LPDACSW_VBIAS2PIN*/|LPDACSW_VZERO2LPTIA/*|LPDACSW_VZERO2PIN*/;
      lploop_cfg.LpDacCfg.LpDacRef = LPDACREF_2P5;
      lploop_cfg.LpDacCfg.LpDacSrc = LPDACSRC_MMR;
      lploop_cfg.LpDacCfg.LpDacVbiasMux = LPDACVBIAS_12BIT; /* Step Vbias. Use 12bit DAC ouput */
      lploop_cfg.LpDacCfg.LpDacVzeroMux = LPDACVZERO_6BIT;  /* Base is Vzero. Use 6 bit DAC ouput */
      lploop_cfg.LpDacCfg.PowerEn = bTRUE;
      AD5940_LPLoopCfgS(&lploop_cfg);
      AD5940_StructInit(&dsp_cfg, sizeof(dsp_cfg));
      dsp_cfg.ADCBaseCfg.ADCMuxN = ADCMUXN_LPTIA0_N;
      dsp_cfg.ADCBaseCfg.ADCMuxP = ADCMUXP_LPTIA0_P;
      dsp_cfg.ADCBaseCfg.ADCPga = AppSWVCfg.AdcPgaGain;
      dsp_cfg.ADCFilterCfg.ADCSinc3Osr = AppSWVCfg.ADCSinc3Osr;
      dsp_cfg.ADCFilterCfg.ADCRate = ADCRATE_800KHZ;  /* ADC runs at 16MHz clock in this example, sample rate is 800kHz */
      dsp_cfg.ADCFilterCfg.BpSinc3 = bFALSE;        /* We use data from SINC3 filter */
      dsp_cfg.ADCFilterCfg.Sinc2NotchEnable = bTRUE;
      dsp_cfg.ADCFilterCfg.BpNotch = bTRUE;
      dsp_cfg.ADCFilterCfg.ADCSinc2Osr = ADCSINC2OSR_1067;  /* Don't care */
      dsp_cfg.ADCFilterCfg.ADCAvgNum = ADCAVGNUM_2;   /* Don't care because it's disabled */
      AD5940_DSPCfgS(&dsp_cfg);
      /* Sequence end. */
      AD5940_SEQGenInsert(SEQ_STOP()); /* Add one extra command to disable sequencer for initialization sequence because we only want it to run one time. */
      /* Stop sequence generator here */
      AD5940_SEQGenCtrl(bFALSE); /* Stop sequencer generator */
      error = AD5940_SEQGenFetchSeq(&pSeqCmd, &SeqLen);
      if(error == AD5940ERR_OK)
        AD5940_StructInit(&AppSWVCfg.InitSeqInfo, sizeof(AppSWVCfg.InitSeqInfo));
        if(SeqLen >= AppSWVCfg.MaxSeqLen)
          return AD5940ERR_SEQLEN;
        AppSWVCfg.InitSeqInfo.SeqId = SEQID_3;
        AppSWVCfg.InitSeqInfo.SeqRamAddr = AppSWVCfg.SeqStartAddr;
        AppSWVCfg.InitSeqInfo.pSeqCmd = pSeqCmd;
        AppSWVCfg.InitSeqInfo.SeqLen = SeqLen;
        AppSWVCfg.InitSeqInfo.WriteSRAM = bTRUE;
        AD5940_SEQInfoCfg(&AppSWVCfg.InitSeqInfo);
      else
        return error; /* Error */
    * @brief Generate ADC control sequence and write the commands to SRAM.
    static AD5940Err AppSWVSeqADCCtrlGen(void)
      uint32_t SeqLen;
      uint32_t WaitClks;
      ClksCalInfo_Type clks_cal;
      clks_cal.DataCount = 1; /* Sample one point everytime */
      clks_cal.DataType = DATATYPE_SINC3;
      clks_cal.ADCSinc3Osr = AppSWVCfg.ADCSinc3Osr;
      clks_cal.ADCSinc2Osr = ADCSINC2OSR_1067;  /* Don't care */
      clks_cal.ADCAvgNum = ADCAVGNUM_2; /* Don't care */
      clks_cal.RatioSys2AdcClk = AppSWVCfg.SysClkFreq/AppSWVCfg.AdcClkFreq;
      AD5940_ClksCalculate(&clks_cal, &WaitClks);
      AD5940_SEQGpioCtrlS(AGPIO_Pin2);
      AD5940_AFECtrlS(AFECTRL_ADCPWR, bTRUE);
      AD5940_SEQGenInsert(SEQ_WAIT(16*100));  /* wait 250us for reference power up */
      AD5940_AFECtrlS(AFECTRL_ADCCNV, bTRUE);  /* Start ADC convert and DFT */
      AD5940_SEQGenInsert(SEQ_WAIT(WaitClks));  /* wait for first data ready */
      AD5940_AFECtrlS(AFECTRL_ADCPWR|AFECTRL_ADCCNV, bFALSE);  /* Stop ADC */
      AD5940_SEQGpioCtrlS(0);
      AD5940_EnterSleepS();/* Goto hibernate */
      /* Sequence end. */
        AD5940_StructInit(&AppSWVCfg.ADCSeqInfo, sizeof(AppSWVCfg.ADCSeqInfo));
        if((SeqLen + AppSWVCfg.InitSeqInfo.SeqLen) >= AppSWVCfg.MaxSeqLen)
        AppSWVCfg.ADCSeqInfo.SeqId = SEQID_2;
        AppSWVCfg.ADCSeqInfo.SeqRamAddr = AppSWVCfg.InitSeqInfo.SeqRamAddr + AppSWVCfg.InitSeqInfo.SeqLen ;
        AppSWVCfg.ADCSeqInfo.pSeqCmd = pSeqCmd;
        AppSWVCfg.ADCSeqInfo.SeqLen = SeqLen;
        AppSWVCfg.ADCSeqInfo.WriteSRAM = bTRUE;
        AD5940_SEQInfoCfg(&AppSWVCfg.ADCSeqInfo);
      else
    * @brief Calculate DAC code step by step.
    * @details The calculation is based on following variables.
    *          - RampStartVolt
    *          - RampPeakVolt
    *          - VzeroStart
    *          - VzeroPeak
    *          - StepNumber
    *          Below variables must be initialzed before call this function. It's done in function @ref AppSWVInit
    *          - RampState
    *          - CurrStepPos
    *          - bDACCodeInc
    *          - CurrRampCode
    static AD5940Err RampDacRegUpdate(uint32_t *pDACData)
      uint32_t VbiasCode, VzeroCode;
      if (AppSWVCfg.bRampOneDir)
    	if(AppSWVCfg.RampStartVolt > AppSWVCfg.RampPeakVolt)
    		AppSWVCfg.bDACCodeInc = bFALSE;
        switch(AppSWVCfg.RampState)
        case SWV_STATE0: /* Begin of Ramp  */
          AppSWVCfg.CurrVzeroCode = (uint32_t)((AppSWVCfg.VzeroStart - 200.0f)/DAC6BITVOLT_1LSB);
          AppSWVCfg.RampState = SWV_STATE1;
          break;
        case SWV_STATE1:
          if(AppSWVCfg.CurrStepPos >= AppSWVCfg.StepNumber/2)
            AppSWVCfg.RampState = SWV_STATE2;   /* Enter State2 */
            AppSWVCfg.CurrVzeroCode = (uint32_t)((AppSWVCfg.VzeroPeak - 200.0f)/DAC6BITVOLT_1LSB);
          break;
        case SWV_STATE2:
          if(AppSWVCfg.CurrStepPos >= AppSWVCfg.StepNumber)
            AppSWVCfg.RampState = SWV_STOP;     /* Enter Stop */
          break;
        case SWV_STOP:
          break;
      else
          break;
        case SWV_STATE1:
          if(AppSWVCfg.CurrStepPos >= AppSWVCfg.StepNumber/4)
          break;
        case SWV_STATE2:
          if(AppSWVCfg.CurrStepPos >= (AppSWVCfg.StepNumber*2)/4)
            AppSWVCfg.RampState = SWV_STATE3;   /* Enter State2 */
            AppSWVCfg.bDACCodeInc = AppSWVCfg.bDACCodeInc ? bFALSE : bTRUE;
          break;
        case SWV_STATE3:
          if(AppSWVCfg.CurrStepPos >= (AppSWVCfg.StepNumber*3)/4)
            AppSWVCfg.RampState = SWV_STATE4;   /* Enter State2 */
          break;
        case SWV_STATE4:
          if(AppSWVCfg.CurrStepPos >= (AppSWVCfg.StepNumber))
          break;
        case SWV_STOP:
          break;
      AppSWVCfg.CurrStepPos++;
      if(AppSWVCfg.bSqrWaveHiLevel)
        if(AppSWVCfg.bDACCodeInc)
          AppSWVCfg.CurrRampCode -= (AppSWVCfg.DACCodePerStep - AppSWVCfg.DACCodePerRamp);
        else
          AppSWVCfg.CurrRampCode -= AppSWVCfg.DACCodePerStep;
        AppSWVCfg.bSqrWaveHiLevel = bFALSE;
      }else
          AppSWVCfg.CurrRampCode += AppSWVCfg.DACCodePerStep;
        else
          AppSWVCfg.CurrRampCode += (AppSWVCfg.DACCodePerStep - AppSWVCfg.DACCodePerRamp);
        AppSWVCfg.bSqrWaveHiLevel = bTRUE;
      VzeroCode = AppSWVCfg.CurrVzeroCode;
      VbiasCode = (uint32_t)(VzeroCode*64 + AppSWVCfg.CurrRampCode);
      if(VbiasCode < (VzeroCode*64))
        VbiasCode --;
      /* Truncate */
      if(VbiasCode > 4095) VbiasCode = 4095;
      if(VzeroCode >   63) VzeroCode =   63;
      *pDACData = (VzeroCode<<12)|VbiasCode;
    /* Geneate sequence(s) to update DAC step by step */
    /* Note: this function doesn't need sequencer generator */
    * @brief Update DAC sequence in SRAM in real time.
    * @details This function generates sequences to update DAC code step by step. It's also called in interrupt
    *          function when half commands in SRAM has been completed. We don't use sequence generator to save memory.
    *          Check more details from documentation of this example. @ref Ramp_Test_Example
    static AD5940Err AppSWVSeqDACCtrlGen(void)
    #define SEQLEN_ONESTEP    4L  /* How many sequence commands are needed to update LPDAC. */
    #define CURRBLK_BLK0      0   /* Current block is BLOCK0 */
    #define CURRBLK_BLK1      1   /* Current block is BLOCK1 */
      uint32_t BlockStartSRAMAddr;
      uint32_t DACData, SRAMAddr;
      uint32_t i;
      uint32_t StepsThisBlock;
      BoolFlag bIsFinalBlk;
      uint32_t SeqCmdBuff[SEQLEN_ONESTEP];
      /* All below static variables are inited in below 'if' block. They are only used in this function */
      static BoolFlag bCmdForSeq0 = bTRUE;
      static uint32_t DACSeqBlk0Addr, DACSeqBlk1Addr;
      static uint32_t StepsRemainning, StepsPerBlock, DACSeqCurrBlk;
      AppSWVCfg.StepNumber = (uint32_t)(2*(AppSWVCfg.RampPeakVolt - AppSWVCfg.RampStartVolt)/AppSWVCfg.SqrWvRampIncrement);
      if(AppSWVCfg.bRampOneDir == bFALSE)
        AppSWVCfg.StepNumber*=2;
    		AppSWVCfg.StepNumber-=2;
      //AppSWVCfg.FifoThresh = AppSWVCfg.StepNumber;
      if(AppSWVCfg.StepNumber >1020)
        printf("Error: Selected Increment, StartVolt and PeakVolt exceed accepted limits \n");
        while(1){}
      /* Do some math calculations */
      if(AppSWVCfg.bFirstDACSeq == bTRUE)
        /* Reset bIsFirstRun at end of function. */
        int32_t DACSeqLenMax;
        StepsRemainning = AppSWVCfg.StepNumber;
        DACSeqLenMax = (int32_t)AppSWVCfg.MaxSeqLen - (int32_t)AppSWVCfg.InitSeqInfo.SeqLen - (int32_t)AppSWVCfg.ADCSeqInfo.SeqLen;
        if(DACSeqLenMax < SEQLEN_ONESTEP*4)
          return AD5940ERR_SEQLEN;  /* No enough sequencer SRAM available */
        DACSeqLenMax -= SEQLEN_ONESTEP*2;  /* Reserve commands each block */
        StepsPerBlock = DACSeqLenMax/SEQLEN_ONESTEP/2;
        DACSeqBlk0Addr = AppSWVCfg.ADCSeqInfo.SeqRamAddr + AppSWVCfg.ADCSeqInfo.SeqLen;
        DACSeqBlk1Addr = DACSeqBlk0Addr + StepsPerBlock*SEQLEN_ONESTEP;
        DACSeqCurrBlk = CURRBLK_BLK0;
        /* Analog part */
        AppSWVCfg.DACCodePerStep = AppSWVCfg.SqrWvAmplitude/DAC12BITVOLT_1LSB;
        AppSWVCfg.DACCodePerRamp = AppSWVCfg.SqrWvRampIncrement/DAC12BITVOLT_1LSB;
    #if ALIGIN_VOLT2LSB
        AppSWVCfg.DACCodePerStep = (int32_t)AppSWVCfg.DACCodePerStep;
        AppSWVCfg.DACCodePerRamp = (int32_t)AppSWVCfg.DACCodePerRamp;
    #endif
        if(AppSWVCfg.DACCodePerStep > 0)
          AppSWVCfg.bDACCodeInc = bTRUE;
        else
        AppSWVCfg.CurrRampCode = AppSWVCfg.RampStartVolt/DAC12BITVOLT_1LSB;
        AppSWVCfg.RampState = SWV_STATE0;   /* Init state to STATE0 */
        AppSWVCfg.CurrStepPos = 0;
        bCmdForSeq0 = bTRUE;      /* Start with SEQ0 */
      if(StepsRemainning == 0) return AD5940ERR_OK; /* Done. */
      bIsFinalBlk = StepsRemainning <= StepsPerBlock?bTRUE:bFALSE;
      if(bIsFinalBlk)
        StepsThisBlock = StepsRemainning;
      else
        StepsThisBlock = StepsPerBlock;
      StepsRemainning -= StepsThisBlock;
      BlockStartSRAMAddr = (DACSeqCurrBlk == CURRBLK_BLK0)?\
    DACSeqBlk0Addr:DACSeqBlk1Addr;
    SRAMAddr = BlockStartSRAMAddr;
    for(i=0; i<StepsThisBlock - 1; i++)
      uint32_t CurrAddr = SRAMAddr;
      SRAMAddr += SEQLEN_ONESTEP;  /* Jump to next sequence */
      RampDacRegUpdate(&DACData);
      SeqCmdBuff[0] = SEQ_WR(REG_AFE_LPDACDAT0, DACData);
      SeqCmdBuff[1] = SEQ_WAIT(10); /* !!!NOTE LPDAC need 10 clocks to update data. Before send AFE to sleep state, wait 10 extra clocks */
      SeqCmdBuff[2] = SEQ_WR(bCmdForSeq0?REG_AFE_SEQ1INFO:REG_AFE_SEQ0INFO,\
        (SRAMAddr<<BITP_AFE_SEQ1INFO_ADDR)|(SEQLEN_ONESTEP<<BITP_AFE_SEQ1INFO_LEN));
      SeqCmdBuff[3] = SEQ_SLP();
      AD5940_SEQCmdWrite(CurrAddr, SeqCmdBuff, SEQLEN_ONESTEP);
      bCmdForSeq0 = bCmdForSeq0?bFALSE:bTRUE;
    /* Add final DAC update */
    if(bIsFinalBlk)/* This is the final block */
      /* After update LPDAC with final data, we let sequencer to run 'final final' command, to disable sequencer.  */
      CurrAddr += SEQLEN_ONESTEP;
      /* The final final command is to disable sequencer. */
      SeqCmdBuff[0] = SEQ_NOP();    /* Do nothing */
      SeqCmdBuff[1] = SEQ_NOP();
      SeqCmdBuff[2] = SEQ_NOP();
      SeqCmdBuff[3] = SEQ_STOP();   /* Stop sequencer. */
      /* Disable sequencer, END of sequencer interrupt is generated. */
    else /* This is not the final block */
      /* Jump to next block. */
    DACSeqBlk1Addr:DACSeqBlk0Addr;
    SeqCmdBuff[3] = SEQ_INT0(); /* Generate Custom interrupt 0. */
    DACSeqCurrBlk = (DACSeqCurrBlk == CURRBLK_BLK0)?\
    CURRBLK_BLK1:CURRBLK_BLK0; /* Switch between Block0 and block1 */
    if(AppSWVCfg.bFirstDACSeq)
      AppSWVCfg.bFirstDACSeq = bFALSE;
      if(bIsFinalBlk == bFALSE)
      { /* Otherwise there is no need to init block1 sequence */
        error = AppSWVSeqDACCtrlGen();
        if(error != AD5940ERR_OK)
          return error;
      /* This is the first DAC sequence. */
      AppSWVCfg.DACSeqInfo.SeqId = SEQID_0;
      AppSWVCfg.DACSeqInfo.SeqLen = SEQLEN_ONESTEP;
      AppSWVCfg.DACSeqInfo.SeqRamAddr = BlockStartSRAMAddr;
      AppSWVCfg.DACSeqInfo.WriteSRAM = bFALSE; /* No need to write to SRAM. We already write them above. */
      AD5940_SEQInfoCfg(&AppSWVCfg.DACSeqInfo);
    * @brief Calibrate LPTIA internal RTIA resistor(s).
    * @details This function will do calibration using parameters stored in @ref AppEDACfg structure.
    static AD5940Err AppSWVRtiaCal(void)
      fImpPol_Type RtiaCalValue;  /* Calibration result */
      LPRTIACal_Type lprtia_cal;
      AD5940_StructInit(&lprtia_cal, sizeof(lprtia_cal));
      lprtia_cal.LpAmpSel = LPAMP0;
      lprtia_cal.bPolarResult = bTRUE;                /* Magnitude + Phase */
      lprtia_cal.AdcClkFreq = AppSWVCfg.AdcClkFreq;
      lprtia_cal.SysClkFreq = AppSWVCfg.SysClkFreq;
      lprtia_cal.ADCSinc3Osr = ADCSINC3OSR_4;
      lprtia_cal.ADCSinc2Osr = ADCSINC2OSR_22;        /* Use SINC2 data as DFT data source */
      lprtia_cal.DftCfg.DftNum = DFTNUM_2048;         /* Maximum DFT number */
      lprtia_cal.DftCfg.DftSrc = DFTSRC_SINC2NOTCH;        /* @todo For frequency under 12Hz, need to optimize DFT source. Use SINC3 data as DFT source */
      lprtia_cal.DftCfg.HanWinEn = bTRUE;
      lprtia_cal.fFreq = AppSWVCfg.AdcClkFreq/4/22/2048*3;  /* Sample 3 period of signal, 13.317Hz here. Do not use DC method, because it needs ADC/PGA calibrated firstly(but it's faster) */
      lprtia_cal.fRcal = AppSWVCfg.RcalVal;
      lprtia_cal.LpTiaRtia = AppSWVCfg.LPTIARtiaSel;
      lprtia_cal.LpAmpPwrMod = LPAMPPWR_NORM;
      lprtia_cal.bWithCtia = bFALSE;
      AD5940_LPRtiaCal(&lprtia_cal, &RtiaCalValue);
      AppSWVCfg.RtiaValue = RtiaCalValue;
      printf("Rtia,%f,%f\n", RtiaCalValue.Magnitude, RtiaCalValue.Phase);
    * @brief Initialize the ramp test. Call this functions every time before start ramp test.
    * @param pBuffer: the buffer for sequencer generator. Only need to provide it for the first time.
    * @param BufferSize: The buffer size start from pBuffer.
    AD5940Err AppSWVInit(uint32_t *pBuffer, uint32_t BufferSize)
      FIFOCfg_Type fifo_cfg;
      SEQCfg_Type seq_cfg;
      /* Configure sequencer and stop it */
      seq_cfg.SeqMemSize = SEQMEMSIZE_2KB;
      seq_cfg.SeqBreakEn = bFALSE;
      seq_cfg.SeqIgnoreEn = bFALSE;
      seq_cfg.SeqCntCRCClr = bTRUE;
      seq_cfg.SeqEnable = bFALSE;
      seq_cfg.SeqWrTimer = 0;
      AD5940_SEQCfg(&seq_cfg);
      /* Start sequence generator */
      /* Initialize sequencer generator */
      if((AppSWVCfg.SWVInited == bFALSE)||\
        (AppSWVCfg.bParaChanged == bTRUE))
        if(pBuffer == 0)  return AD5940ERR_PARA;
        if(BufferSize == 0) return AD5940ERR_PARA;
        if(AppSWVCfg.LPTIARtiaSel == LPTIARTIA_OPEN) /* Internal RTIA is opened. User wants to use external RTIA resistor */
          AppSWVCfg.RtiaValue.Magnitude = AppSWVCfg.ExternalRtiaValue;
          AppSWVCfg.RtiaValue.Phase = 0;
        else
          AppSWVRtiaCal();
        AppSWVCfg.SWVInited = bFALSE;
        AD5940_SEQGenInit(pBuffer, BufferSize);
        /* Generate sequence and write them to SRAM start from address AppSWVCfg.SeqStartAddr */
        error = AppSWVSeqInitGen();      /* Application initialization sequence */
        if(error != AD5940ERR_OK) return error;
        error = AppSWVSeqADCCtrlGen();   /* ADC control sequence */
        AppSWVCfg.bParaChanged = bFALSE; /* Clear this flag as we already implemented the new configuration */
      /* Reconfigure FIFO, The Rtia calibration function may generate data that stored to FIFO */
      AD5940_FIFOCtrlS(FIFOSRC_SINC3, bFALSE);        /* Disable FIFO firstly */
      fifo_cfg.FIFOEn = bTRUE;
      fifo_cfg.FIFOSrc = FIFOSRC_SINC3;
      fifo_cfg.FIFOThresh = AppSWVCfg.FifoThresh;    /* Change FIFO paramters */
      fifo_cfg.FIFOMode = FIFOMODE_FIFO;
      fifo_cfg.FIFOSize = FIFOSIZE_4KB;
      AD5940_FIFOCfg(&fifo_cfg);
      /* Clear all interrupts */
      AD5940_INTCClrFlag(AFEINTSRC_ALLINT);
      /* Generate DAC sequence */
      AppSWVCfg.bFirstDACSeq = bTRUE;
      /* Configure sequence info. */
      AppSWVCfg.InitSeqInfo.WriteSRAM = bFALSE;
      AD5940_SEQCtrlS(bTRUE); /* Enable sequencer */
      AD5940_SEQMmrTrig(AppSWVCfg.InitSeqInfo.SeqId);
      while(AD5940_INTCTestFlag(AFEINTC_1, AFEINTSRC_ENDSEQ) == bFALSE);
      AD5940_INTCClrFlag(AFEINTSRC_ENDSEQ);
      AppSWVCfg.ADCSeqInfo.WriteSRAM = bFALSE;
      AD5940_SEQCtrlS(bFALSE);
      AD5940_WriteReg(REG_AFE_SEQCNT, 0);
      AD5940_SEQCtrlS(bTRUE);   /* Enable sequencer, and wait for trigger */
      AD5940_ClrMCUIntFlag();   /* Clear interrupt flag generated before */
      AD5940_AFEPwrBW(AFEPWR_LP, AFEBW_250KHZ); /* Set to low power mode */
      AppSWVCfg.SWVInited = bTRUE;  /* SWV application has been initialized. */
    * @brief This function is called in ISR when AFE has been wakeup and we can access registers.
    * @param pData: the buffer points to data read back from FIFO. Not needed for this application-SWV
    * @param pDataCount: The data count in pData buffer.
    static int32_t AppSWVRegModify(int32_t * const pData, uint32_t *pDataCount)
      if(AppSWVCfg.StopRequired == bTRUE)
    * @brief Depending on the data type, do appropriate data pre-process before return back to controller
    static int32_t AppSWVDataProcess(int32_t * const pData, uint32_t *pDataCount)
      uint32_t i, datacount;
      datacount = *pDataCount;
      float *pOut = (float *)pData;
      for(i=0;i<datacount;i++)
        pData[i] &= 0xffff;
        pOut[i] = AD5940_ADCCode2Volt(pData[i], AppSWVCfg.AdcPgaGain, AppSWVCfg.ADCRefVolt)/AppSWVCfg.RtiaValue.Magnitude * 1e3f;  /* Result unit is uA. */
      return 0;
    * @brief The interrupt service routine for SWV test.
    * @param pBuff: The buffer provides by host, used to store data read back from FIFO.
    * @param pCount: The available buffer size starts from pBuff.
    AD5940Err AppSWVISR(void *pBuff, uint32_t *pCount)
      uint32_t BuffCount;
      uint32_t FifoCnt;
      BuffCount = *pCount;
      uint32_t IntFlag;
      AD5940_SleepKeyCtrlS(SLPKEY_LOCK);
      *pCount = 0;
      IntFlag = AD5940_INTCGetFlag(AFEINTC_0);
      if(IntFlag & AFEINTSRC_CUSTOMINT0)          /* High priority. */
        AD5940Err error;
        AD5940_INTCClrFlag(AFEINTSRC_CUSTOMINT0);
    		//AD5940_McuSetHigh();
    		//AD5940_McuSetLow();
       // AD5940_SleepKeyCtrlS(SLPKEY_UNLOCK);
        //AD5940_EnterSleepS(); /* If there is need to do AFE re-configure, do it here when AFE is in active state */
      if(IntFlag&AFEINTSRC_DATAFIFOTHRESH)
        FifoCnt = AD5940_FIFOGetCnt();
        if(FifoCnt > BuffCount)
          ///@todo buffer is limited.
        AD5940_FIFORd((uint32_t *)pBuff, FifoCnt);
        AD5940_INTCClrFlag(AFEINTSRC_DATAFIFOTHRESH);
        AppSWVRegModify(pBuff, &FifoCnt);
        /* Process data */
        AppSWVDataProcess((int32_t*)pBuff,&FifoCnt);
        *pCount = FifoCnt;
        return 0;
      if(IntFlag & AFEINTSRC_ENDSEQ)
        /* Process data */
        *pCount = FifoCnt;
        AppSWVCtrl(APPCTRL_STOPNOW, 0);    /* Stop the Wakeup Timer. */
      return 0;
     @file:    AD5940Main.c
     @author:  $Author: nxu2 $
     @brief:   Used to control specific application and process data.
     @version: $Revision: 766 $
     @date:    $Date: 2017-08-21 14:09:35 +0100 (Mon, 21 Aug 2017) $
       User could configure following parameters
    #define APPBUFF_SIZE 1024
    uint32_t AppBuff[APPBUFF_SIZE];
    float LFOSCFreq;    /* Measured LFOSC frequency */
     * @brief An example to deal with data read back from AD5940. Here we just print data to UART
     * @note UART has limited speed, it has impact when sample rate is fast. Try to print some of the data not all of them.
     * @param pData: the buffer stored data for this application. The data from FIFO has been pre-processed.
     * @param DataCount: The available data count in buffer pData.
     * @return return 0.
    static int32_t RampShowResult(float *pData, uint32_t DataCount)
      static uint32_t index;
      /* Print data*/
      for(int i=0;i<DataCount;i++)
        printf("index:%d, %.3f\n", index++, pData[i]);
        //printf("%.3f\n",pData[i]);
        //index++;
        //i += 10;  /* Print though UART consumes too much time. */
      return 0;
     * @brief The general configuration to AD5940 like FIFO/Sequencer/Clock.
     * @note This function will firstly reset AD5940 using reset pin.
     * @return return 0.
    static int32_t AD5940PlatformCfg(void)
      CLKCfg_Type clk_cfg;
      AGPIOCfg_Type gpio_cfg;
      LFOSCMeasure_Type LfoscMeasure;
      /* Use hardware reset */
      AD5940_HWReset();
      AD5940_Initialize();    /* Call this right after AFE reset */
      /* Platform configuration */
      /* Step1. Configure clock */
      clk_cfg.HFOSCEn = bTRUE;
      clk_cfg.HFXTALEn = bFALSE;
      clk_cfg.LFOSCEn = bTRUE;
      clk_cfg.HfOSC32MHzMode = bFALSE;
      clk_cfg.SysClkSrc = SYSCLKSRC_HFOSC;
      clk_cfg.SysClkDiv = SYSCLKDIV_1;
      clk_cfg.ADCCLkSrc = ADCCLKSRC_HFOSC;
      clk_cfg.ADCClkDiv = ADCCLKDIV_1;
      AD5940_CLKCfg(&clk_cfg);
      /* Step2. Configure FIFO and Sequencer*/
      fifo_cfg.FIFOEn = bTRUE;           /* We will enable FIFO after all parameters configured */
      fifo_cfg.FIFOSize = FIFOSIZE_4KB;   /* 2kB for FIFO, The reset 4kB for sequencer */
      fifo_cfg.FIFOSrc = FIFOSRC_SINC3;   /* */
      fifo_cfg.FIFOThresh = 4;            /*  Don't care, set it by application paramter */
      seq_cfg.SeqMemSize = SEQMEMSIZE_2KB;  /* 4kB SRAM is used for sequencer, others for data FIFO */
      seq_cfg.SeqIgnoreEn = bTRUE;
      /* Step3. Interrupt controller */
      AD5940_INTCCfg(AFEINTC_1, AFEINTSRC_ALLINT, bTRUE);   /* Enable all interrupt in INTC1, so we can check INTC flags */
      AD5940_INTCCfg(AFEINTC_0, AFEINTSRC_DATAFIFOTHRESH|AFEINTSRC_ENDSEQ|AFEINTSRC_CUSTOMINT0, bTRUE);
      /* Step4: Configure GPIOs */
      gpio_cfg.FuncSet = GP0_INT|GP1_SLEEP|GP2_SYNC;  /* GPIO1 indicates AFE is in sleep state. GPIO2 indicates ADC is sampling. */
      gpio_cfg.InputEnSet = 0;
      gpio_cfg.OutputEnSet = AGPIO_Pin0|AGPIO_Pin1|AGPIO_Pin2;
      gpio_cfg.OutVal = 0;
      gpio_cfg.PullEnSet = 0;
      AD5940_AGPIOCfg(&gpio_cfg);
      /* Measure LFOSC frequency */
      /**@note Calibrate LFOSC using system clock. The system clock accuracy decides measurement accuracy. Use XTAL to get better result. */
      LfoscMeasure.CalDuration = 1000.0;  /* 1000ms used for calibration. */
      LfoscMeasure.CalSeqAddr = 0;        /* Put sequence commands from start address of SRAM */
      LfoscMeasure.SystemClkFreq = 16000000.0f; /* 16MHz in this firmware. */
      AD5940_LFOSCMeasure(&LfoscMeasure, &LFOSCFreq);
      printf("LFOSC Freq:%f\n", LFOSCFreq);
     // AD5940_SleepKeyCtrlS(SLPKEY_UNLOCK);         /*  */
      return 0;
     * @brief The interface for user to change application paramters.
     * @return return 0.
    void AD5940RampStructInit(void)
      AppSWVCfg_Type *pRampCfg;
      AppSWVGetCfg(&pRampCfg);
      /* Step1: configure general parmaters */
      pRampCfg->SeqStartAddr = 0x10;                /* leave 16 commands for LFOSC calibration.  */
      pRampCfg->MaxSeqLen = 512-0x10;              /* 4kB/4 = 1024  */
      pRampCfg->RcalVal = 10000.0;                  /* 10kOhm RCAL */
      pRampCfg->ADCRefVolt = 1.820f;               /* The real ADC reference voltage. Measure it from capacitor C12 with DMM. */
      pRampCfg->FifoThresh = 1023;                   /* Maximum value is 2kB/4-1 = 512-1. Set it to higher value to save power. */
      pRampCfg->SysClkFreq = 16000000.0f;           /* System clock is 16MHz by default */
      pRampCfg->LFOSCClkFreq = LFOSCFreq;           /* LFOSC frequency */
    	pRampCfg->AdcPgaGain = ADCPGA_1P5;
    	pRampCfg->ADCSinc3Osr = ADCSINC3OSR_4;
    	/* Step 2:Configure square wave signal parameters */
      pRampCfg->RampStartVolt = -400.0f;     /* Measurement starts at 0V*/
      pRampCfg->RampPeakVolt = 0.0f;     		 /* Measurement finishes at -0.4V */
      pRampCfg->VzeroStart = 1300.0f;           /* Vzero is voltage on SE0 pin: 1.3V */
      pRampCfg->VzeroPeak = 1300.0f;          /* Vzero is voltage on SE0 pin: 1.3V */
      pRampCfg->Frequency = 750;                 /* Frequency of square wave in Hz */
      pRampCfg->SqrWvAmplitude = 150;       /* Amplitude of square wave in mV */
      pRampCfg->SqrWvRampIncrement = 5; /* Increment in mV*/
      pRampCfg->SampleDelay = 0.2f;             /* Time between update DAC and ADC sample. Unit is ms and must be < (1/Frequency)/2 - 0.2*/
      pRampCfg->LPTIARtiaSel = LPTIARTIA_1K;      /* Maximum current decides RTIA value */
    	pRampCfg->bRampOneDir = bFALSE;//bTRUE;			/* Only measure ramp in one direction */
    void AD5940_Main(void)
      uint32_t temp;
      AD5940PlatformCfg();
      AD5940RampStructInit();
      AppSWVInit(AppBuff, APPBUFF_SIZE);    /* Initialize RAMP application. Provide a buffer, which is used to store sequencer commands */
    	AD5940_Delay10us(100000);		/* Add a delay to allow sensor reach equilibrium befor starting the measurement */
      AppSWVCtrl(APPCTRL_START, 0);          /* Control IMP measurement to start. Second parameter has no meaning with this command. */
      while(1)
        if(AD5940_GetMCUIntFlag())
          temp = APPBUFF_SIZE;
          AppSWVISR(AppBuff, &temp);
          RampShowResult((float*)AppBuff, temp);
     @file:    RampTest.H
     @brief:   Ramp Test header file.
    #ifndef _SWVTEST_H_
    #define _SWVTEST_H_
    #include "ad5940.h"
    #include <stdio.h>
    #include "string.h"
    #include "math.h"
    /* Do not modify following parameters */
    #define ALIGIN_VOLT2LSB     0   /* Set it to 1 to align each voltage step to 1LSB of DAC. 0: step code is fractional. */
    #define DAC12BITVOLT_1LSB   (2200.0f/4095)  //mV
    #define DAC6BITVOLT_1LSB    (DAC12BITVOLT_1LSB*64)  //mV
     * The Ramp application related paramter structure
    typedef struct
    /* Common configurations for all kinds of Application. */
      BoolFlag  bParaChanged;         /**< Indicate to generate sequence again. It's auto cleared by AppBIAInit */
      uint32_t  SeqStartAddr;         /**< Initialaztion sequence start address in SRAM of AD5940  */
      uint32_t  MaxSeqLen;            /**< Limit the maximum sequence.   */
      uint32_t  SeqStartAddrCal;      /**< Not used for Ramp.Calibration sequence start address in SRAM of AD5940 */
      uint32_t  MaxSeqLenCal;         /**< Not used for Ramp. */
    /* Application related parameters */
      float     LFOSCClkFreq;         /**< The clock frequency of Wakeup Timer in Hz. Typically it's 32kHz. Leave it here in case we calibrate clock in software method */
      float     SysClkFreq;           /**< The real frequency of system clock */
      float     AdcClkFreq;           /**< The real frequency of ADC clock */
      float     RcalVal;              /**< Rcal value in Ohm */
      float     ADCRefVolt;           /**< The real ADC voltage in mV. */
      float     RampStartVolt;        /**< The start voltage of ramp signal in mV */
      float     RampPeakVolt;         /**< The maximum or minimum voltage of ramp in mV */
      float     VzeroStart;           /**< The start voltage of Vzero in mV. Set it to 2400mV by default */
      float     VzeroPeak;            /**< The peak voltage of Vzero in mV. Set it to 200mV by default */
      uint32_t  StepNumber;           /**< Total number of steps. Limited to 4095. */
      float     SampleDelay;          /**< The time delay between update DAC and start ADC */
      uint32_t  LPTIARtiaSel;         /**< Select RTIA */
      float     ExternalRtiaValue;    /**< The optional external RTIA value in Ohm. Disconnect internal RTIA to use external RTIA. When using internal RTIA, this value is ignored. */
      uint32_t  AdcPgaGain;           /**< PGA Gain select from GNPGA_1, GNPGA_1_5, GNPGA_2, GNPGA_4, GNPGA_9 !!! We must ensure signal is in range of +-1.5V which is limited by ADC input stage */
      uint8_t   ADCSinc3Osr;          /**< We use data from SINC3 filter. */
      /* Digital related */
      uint32_t  FifoThresh;           /**< FIFO Threshold value */
    /* Private variables for internal usage */
      BoolFlag  SWVInited;           /**< If the program run firstly, generated initialization sequence commands */
      fImpPol_Type  RtiaValue;        /**< Calibrated Rtia value */
      SEQInfo_Type  InitSeqInfo;
      SEQInfo_Type  ADCSeqInfo;
      BoolFlag      bFirstDACSeq;     /**< Init DAC sequence */
      SEQInfo_Type  DACSeqInfo;       /**< The first DAC update sequence info */
      uint32_t  CurrStepPos;          /**< Current position */
      float     DACCodePerStep;       /**< DAC codes in square waveform */
      float     DACCodePerRamp;       /**< DAC codes needed to ramp increment */
      float     CurrRampCode;         /**<  */
      float     Frequency;						/**< Frequency of square wave */
      float     SqrWvAmplitude;       /**< Set amplitude of square wave */
      float     SqrWvRampIncrement;   /**< Ramp increase in mV */
      uint32_t  CurrVzeroCode;
      BoolFlag  bDACCodeInc;          /**< Increase DAC code.  */
      BoolFlag  bSqrWaveHiLevel;			/**< Flag to indicate square wave high level */
    	BoolFlag bRampOneDir;						/**< Ramp in one direction only */
      BoolFlag  StopRequired;         /**< After FIFO is ready, stop the measurement sequence */
      enum _RampState{SWV_STATE0 = 0, SWV_STATE1, SWV_STATE2, SWV_STATE3, SWV_STATE4, SWV_STOP} RampState;
    }AppSWVCfg_Type;
    #define APPCTRL_START          0
    #define APPCTRL_STOPNOW        1
    #define APPCTRL_STOPSYNC       2
    #define APPCTRL_SHUTDOWN       3   /**< Note: shutdown here means turn off everything and put AFE to hibernate mode. The word 'SHUT DOWN' is only used here. */
    AD5940Err AppSWVInit(uint32_t *pBuffer, uint32_t BufferSize);
    AD5940Err AppSWVGetCfg(void *pCfg);
    AD5940Err AppSWVISR(void *pBuff, uint32_t *pCount);
    AD5940Err AppSWVCtrl(uint32_t Command, void *pPara);
    void AD5940_McuSetLow(void);
    void AD5940_McuSetHigh(void);
    #endif
  • Hi Akila,
    Thanks for your response!
    No, I am using a version created in 2017 and got from Analog Device github.
    Can you point me to where I can get the latest examples?
    Olive
  • The codes are in the process of update in Github, For now, kindly use the above posted codes.
  • Hello Akila,
    When do you expect the new commit to be on Github?
    Thanks!
  • Hi Akila,
    I am also interested in knowing when the new code could be available.
    Please, kindly inform if you guys have any updates.
    Thanks!
     The new Square wave voltammetry code is posted above in this thread. Github will be updated soon.
