Hello, i am using LTC1864 ADC for Converting PT100 data into Digital count ,
Issue: when i am running the device on Battery (11.1 A , 1Ah) supply the counts given by ADC are stable like linear count but when i am running the deivce on Adapter (12v 1A) counts are not stable they are varrying with large values difference. After that i gave supply from SMPS 12v 1A , the reading s were still varrying with lots of large values so i connected 10uf tantalum capacitor at ADC Input and vref and VCC pins shorted to vcc and there i used bypass capacitor of 1uf tantalum and readings were stable , problem is only for Adapter when adapter supply is provided reading s are fluctuating. I am attathcing the circuit i used there i replaced c7 with tentalum and R12 removed , c8- tantalum 1uf and VCC and Vref is shorted and conncted to vcc +5v .