ADAS1000: Why the amplitude of ECG is changing when I am trying to use electrode configuration?

Hi,

I am trying to configure ADAS1000 in electrode mode as per fig60 in the datasheet. When I record the data in the single ended mode, with gain 4.2, the amplitude of the ECG signal is increased to 2.0mv (originally fed signal amplitude =1.0mv).  Can anyone let me know why this is happening? Please find the attached document with screenshots of the register configuration.

https://docs.google.com/document/d/1gYhmpxkSDmIOxi4LLl5fzq-C25UkHgWl7PStEJBmZQ8/edit?usp=sharing 

Thanks in advance

Parents Reply Children
  • Hi Mark

    We have tried with other gain settings also (1.4, 2.1 and 2.8) and we are facing the same issue. When we feed ECG with 2.0mv we are not getting any waveform. Can you please help regrading this. Can you please share the common electrode A mode (ref fig 60, pg.33 from datasheet) working configuration file? I have also attached the register configuration we have used. Please verify it and let us know the changes to be made.

    Thanks in advance


    00000001	00D803BE	00F003DE
    00000002	00000000	00000000
    00000003	00000000	00000000
    00000004	00000F88	00000F88
    00000005	0020010A	00000004
    00000006	00000000	00000000
    00000007	00242424	00242424
    00000008	00000000	00000000
    00000009	00002000	00002000
    0000000A	00079018	00079018
    0000000B	00000004	00000004
    0000000C	0000FFFF	0000FFFF
    0000000D	00000000	00000000
    0000000E	00000000	00000000
    0000000F	00000000	00000000
    00000020	00000000	00000000
    00000021	00000051	00000FAD
    00000022	00000FF8	00000FCA
    00000023	00000036	00000F4D
    00000024	0000002F	00000FEF
    00000025	00000FFB	00000FC8
    0000002A	00000000	00000000
    0000002B	00001200	00001200
    0000002C	00FFF6FB	00FFF6FB
    0000002D	00000000	00000000
    0000002E	00000000	00000000
    

  • Hi Mark,

    We are waiting for the register configuration and solution. Since we have to proceed with our activity, can you please send it the earliest?

    Thanks in advance

  • Hi,
    Are you using the applications board from ADI? If yes, the ADAS1000 GUI will be very helpful on experimenting with the ADAS1000 settings. 

    In your register settings, your ECGCTL disables the RA read and RLD setting in CMREFCTL is also off, you'll want to enable LARLD, LLRLD, and RARLD in CMREFCTL. 

    For more details on ADAS1000 setup, please check the EXAMPLES OF INTERFACING TO THE ADAS1000 section of ADAS1000/ADAS1000-1/ADAS1000-2 (Rev. C) (analog.com) Example 1 should help you get up to speed on ADAS1000

    regards,
    Mark

  • Hi Mark, 

    We are using ADAS1000SDZ EVM module. We will verify the register settings suggested by you. On the board we had routed RA to CE_IN as per fig60 in datasheet. Enabling RA in ECGCTL  was inducing noise so we had to disable it. 

    Can you share the working configuration for common electrode mode?

  • Hi Mark,

    We configured ADAS1000 as per fig62 in the data sheet and we are getting ECG data with proper amplitude. When we capture the data from write to file option, the columns displayed are LA,LL,RA, and V1 to V6. Can you clarify if the displayed columns are actually LA -CE, LL-CE and so on?

    Thanks in advance.

  • Hi,
    Since CE=RA, it its better to tag it as LA-RA, LL-RA, V1-RA, V2-RA,...Please refer to the table in Figure 62 for data word arrangement 


    regards,
    Mark

  • Hi Mark, 

    We tried the configuration you mentioned, we are still facing the same problem of amplitude change  and not able to configure the ADAS1000 according to fig 60.

    Can you please share the register configuration to configure ADAS1000 according to fig60?

    Thanks in advance.

  • Hi Mark,

    We are waiting for the register configuration and solution. Since we have to proceed with our activity, can you please send it the earliest?

    Thanks in advance

  • Hi,
    The critical register bits for this configuration are stated in figure 62 as well. DATAFMT = 1, CHCONFIG = 0, and CEREFEN = 1. You should also enable all channels in ECGCTL. 


    Regards,
    Mark

  • Hi Mark, 

    We are getting data properly according to fig 62, we want to capture data in vector/lead mode as per fig 60.  For fig 60 DATAFMT=0, can you help us with this configuration?

    Thanks in advance