I have a few questions regarding the AD5940.
1) LPDAC calibration: Can you give me more information how the calculations of the parameters (bC2V_DACXB, kC2V_DACxB) are done in the function "AD5940_LPDACCal"? For example, it is unclear to me where the factor 1e3f/32768*1.835f/1.82f comes from.
2) LPDAC calibration: I have a deviation from the expected output for both the 6 and 12 bit output of the LPDAC. For example, with a register value (Vzero0/6bit output) of 0x00, I expect a voltage of 200mV according to the datasheet, but I only measure 190mV. I was hoping that the LPDAC calibration would minimize the deviation. Is there a configuration or setup to minimize these deviations/tolerances?
3) LPDAC calibration: I used the formula from the example code (ad5940.h) (Code(LSB) = kV2C_DACxB * Voltage(mV) + bV2C_DACxB) to calculate the register value for a given voltage. Again, I was able to find a discrepancy. For example, if I want to achieve a voltage of 800mV, I get a calculated register value of 0x11. However, the measured voltage at the LPDAC output is only 730mV. Probably this is related to the calculated factors from the function "AD5940_LPDACCal". Am I missing something or why do I still have deviations even after calibration?
4) Excitation signal: Is it possible to set a negative voltage between the excitation signal and the voltage at the Vzero0 pin (voltage at SE0)?
thanks in advance
    1) AD5940 is factory calibrated for 1.835V internal reference voltage.
    Since the reference voltage to be given is 1.82V, correction factor is multiplied in LPDAC calibration function AD5940_LPDACCa().
    2) Are you sure there is no additional resistor in the measurement path?
    3) You may check out the code2 voltage conversion function below:
      @brief Convert ADC Code to voltage.
      @param ADCPga: The ADC PGA used for this result.
      @param code: ADC code.
      @param VRef1p82: the actual 1.82V reference voltage.
      @return Voltage in volt.
    float AD5940_ADCCode2Volt(uint32_t code, uint32_t ADCPga, float VRef1p82)
      float kFactor = 1.835/1.82;
      float fVolt = 0.0;
      float tmp = 0;
      tmp = (int32_t)code - 32768;
      switch(ADCPga)
      case ADCPGA_1:
        break;
      case ADCPGA_1P5:
        tmp /= 1.5f;
        break;
      case ADCPGA_2:
        tmp /= 2.0f;
        break;
      case ADCPGA_4:
        tmp /= 4.0f;
        break;
      case ADCPGA_9:
        tmp /= 9.0f;
        break;
      default:break;
      fVolt = tmp*VRef1p82/32768*kFactor;
      return fVolt;
    4) With VZero chosen in the middle, excitation voltage can swing in both positive and negative directions w.r.t. VZero.
  • Thank you for your quick response!
    2) What exactly do you mean with additional resistance? I have internally disconnected everything except the HSTIA. The positive input of the HSTIA is connected to Vzero0 and the negative input is connected to SE0. Then I directly measured the Vzero0 pin output and at the SE0 pin with an oscilloscope.
    3) As I understood, you can use the function "AD5940_ADCCode2Volt" to convert an ADC code to the corresponding voltage. However, my question was about the LPDAC. I want to set a specific voltage on the LPDAC's 6/12bit output and need a formula/function that will return the corresponding register value and include the calibration parameters (kV2C_DACxB, bV2C_DACxB).
    4) Figure 26 (AD5940 datasheet) shows a sinusoidal voltage with a positive offset (VBias0-Vzero0). Is it also possible to set a negative DC voltage (negative offset) in respect to Vzero0 for the generated sinusoidal voltage?
    1) There is a loading effect between the 12 bit DAC and the 6 Bit DAC.
    Is the part below added in your code?
    12BITCODE = LPDACDAT0 [11:0];
    6BITCODE = LPDACDAT0 [17:12];
    if (12BITCODE < (6BITCODE *64))
    LPDACDAT [11:0] = (12BITCODE – 1);
    2) The LPDAC transfer function is in page 31 of AD5940 datasheet:
    When the 12-bit output is less than the 6-bit output,
    12-Bit DAC Output Voltage = 0.2 V + (LPDACDAT0, Bits[11:0] × 12-BIT_LSB_DAC)
    6-Bit DAC Output Voltage = 0.2 V + (LPDACDAT0, Bits[17:12] × 6-BIT_LSB_DAC) – 12-BIT_LSB_DAC)
    When the 12-bit output is ≥ the 6-bit output,
    3) (VBIAS - VZERO) can be from  - 600mV upto +600mV.