Abnormal Analysis Value on EV-AD7609, AD7609EDZ
I tested DC 10Vinput for AD7609 with same condition on EV-Board
Normal case CH 1,2,3,4,5,6,7 on EV-AD7609
But CH8 has abnormal Analysis Vaule and has more offset than normal case CH.
And I attached screen captured of GUI.
 AD7609EDZ-edit.zip
Please check why the result value is different only in 8 channels.
    •  Super User
    on Feb 24, 2021 1:52 AM
    We will look into this, I'll contact the product owner and get back to you.
    Andrei
    on Feb 25, 2021 2:05 PM
    Hello Ryan,
    From the screen shots it looks like you are indeed saturating the ADC? So, is it a matter of offset or gain?
    How would it look if applied say 0V or 9V.
    Lluis.
    • I attached test result of AD7609, Customer is requesting the FAR for this issue.
    • Unit 1. (CH4 is failure channel)
    Information (Voltage)
    1. Accuracy ± ( 0.1 ) % of span
    Channel
    Standard
    0V
    2V
    4V
    6V
    8V
    10V
    1
    Measuring
    0.000
    V
    1.999
    V
    4.000
    V
    6.001
    V
    8.002
    V
    10.003
    V
    2
    Measuring
    0.000
    V
    1.999
    V
    4.000
    V
    6.000
    V
    8.002
    V
    10.003
    V
    3
    Measuring
    0.000
    V
    1.999
    V
    3.999
    V
    6.000
    V
    8.001
    V
    10.002
    V
    4
    Measuring
    0.000
    V
    2.000
    V
    4.000
    V
    6.001
    V
    8.003
    V
    10.005
    V
    5
    Measuring
    0.000
    V
    1.999
    V
    4.000
    V
    6.000
    V
    8.001
    V
    10.003
    V
    6
    Measuring
    0.000
    V
    1.999
    V
    3.999
    V
    6.000
    V
    8.001
    V
    10.003
    V
    7
    Measuring
    0.000
    V
    1.999
    V
    3.999
    V
    6.000
    V
    8.001
    V
    10.002
    V
    8
    Measuring
    0.000
    V
    1.999
    V
    4.000
    V
    6.000
    V
    8.001
    V
    10.002
    V
    • Unit 2.(CH1, CH2 is failure channel)
    Channel
    Standard
    0V
    2V
    4V
    6V
    8V
    10V
    1
    Measuring
    0.000
    V
    2.000
    V
    4.000
    V
    6.001
    V
    8.004
    V
    10.005
    V
    2
    Measuring
    0.000
    V
    1.999
    V
    4.000
    V
    6.001
    V
    8.003
    V
    10.005
    V
    3
    Measuring
    0.000
    V
    1.999
    V
    4.001
    V
    6.002
    V
    8.003
    V
    10.003
    V
    4
    Measuring
    0.000
    V
    2.000
    V
    4.000
    V
    6.001
    V
    8.003
    V
    10.003
    V
    5
    Measuring
    0.000
    V
    1.999
    V
    4.000
    V
    6.001
    V
    8.002
    V
    10.003
    V
    6
    Measuring
    0.000
    V
    2.000
    V
    4.000
    V
    6.000
    V
    8.001
    V
    10.002
    V
    7
    Measuring
    0.000
    V
    2.000
    V
    4.000
    V
    6.001
    V
    8.002
    V
    10.003
    V
    8
    Measuring
    0.000
    V
    2.000
    V
    4.000
    V
    6.001
    V
    8.002
    V
    10.003
    V
    Thanks
    on Feb 26, 2021 9:30 AM in reply to Ryan King
    Hi Ryan,
    From your table, first column it seems no offset error on any channel.
    From your table, last coulmn, it seems like 3mV Positive Full Scale mismatch, in LSBs that is 19 LSBs? That looks to be within the 80LSBs maximum specification, isn´t it?
    Lluis.
  • Hi Lluis,
    Why are these channels saturated in EV-Board's GUI?
    Thanks