i use ad7792 with pt100。
the voltage of pt100 is 0.107V,
7792 setting is REF=3.3v 1/16, ( 0.20625v ) IOUT=1mA,
But it“s readback data is always half of 65536,
It's wrong.I wonder know how.
Best regards.
Aaron
i use ad7792 with pt100。
the voltage of pt100 is 0.107V,
7792 setting is REF=3.3v 1/16, ( 0.20625v ) IOUT=1mA,
But it“s readback data is always half of 65536,
It's wrong.I wonder know how.
Best regards.
Aaron
Hi,
Can you share your schematic diagram please and register configuration?
Are you also using a ratio metric configuration? If you want to know more about RTD circuits, I would recommend to check CN-0383 (Rev. C) (analog.com) and also download https://www.analog.com/media/en/designtools/calculators/rtd_configurator_and_error_budget_calculator-rel1.0.0.exe
This tool uses a much newer product AD7124 which I also recommend for new designs as it has better performance and more advanced features suitable for RTD sensing application.
Thanks,
Jellenie
Hi,JellenieR
Thanks for your help,it's OK now,
I reset chip after power on.
Best regards,
Aaron.
Hi Aaron.
Good to hear that you have solved the issue. Though I still recommend to configure your circuit in a ratiometric configuration for better accuracy.
Thanks,
Jellenie