Calibrating current sources on AD7124

Thread Summary

The user inquires about using AD7124-8 or AD7124-4 to measure RTD and thermistor temperatures, with a focus on calibrating current sources using internal reference and precision resistors. The final answer confirms that the current sources have good short-term stability, are minimally affected by the internal MUX, and should be individually calibrated for best performance. The reference design CN-0381 is suggested for a ratiometric configuration to avoid current source inaccuracies.
AI Generated Content

Hi,

I want to use an AD7124-8 or AD7124-4 to measure an RTD and a Thermistor.  In the reference design for the 3 wire RTD, CN-0383, the inaccuracy for the current sources is removed by generating the reference voltage by applying the current source to a precision resistor.  I would like to also be able to read other temperature devices, like a 22K thermistor. To that end I would like to use the internal reference and the two current sources as described in CN-0383, but instead of having the reference resistor in line with the RTD, have a set of reference resistors all hooked up to their own A/D channels such that I can measure the current sources with a full scale reading. I.e. a precision 2.2k resistor to calibrate the 1ma source. 5.11k for the 500uA, etc. each hooked up to their own A/D mux input and current source output.

A couple of issues come to mind.

1) The short term stability of the current sources.  Will they change values in a short period of time?  I.e. the time it takes to switch them from one mux output to another?  From the temperature device to the reference resistor?

2) The effect of the sink impedance on the current sources.  Will the current sources change value if I switch from a 10K thermistor in a 100 degree bath, approximately 800 ohms, and then switch it to my reference 10,000 ohm +- 0.1% low drift resistor?

3) Any effect that the internal MUX has on the accuracy of the current sources.  Is it safe to assume that this is zero? Or is there a practical way to calibrate this out?

4) Must I assume that all of the six current sources are independent?  I.e. If I calibrate the 1mA source will its cal scale to the 50uA source?

Thanks!

Chuck

  • Hi Chuck,

    Would it be alright if you share your schematic diagram so we could have a better understanding of your application?

    This reference circuit might be helpful to you, it's about thermocouple measurement system using AD7124 but in the Common Variations section on page 8, a reference circuit using a thermistor for cold junction temperature measurement can be found.

    Regards,

    RC

  • RC,

    Does this make more sense?  I am trying to measure my temperature device (pt100 or pt1000 or 10K Therm or 22K Therm) using AIN2 and AIN3, supplying excitation current from IOUT0 at 500uA via AIN1.  Afterh that measurement, I then remove IOUT0 from AIN1 and then apply it to AIN5 and measure the current source against the 2.5v reference via AIN6 and AIN7.

    more below....

    The issues that come to mind, again are:

    1) The short term stability of the current sources.  Will they change values in a short period of time?  I.e. the time it takes to switch them from one mux output to another?  From the temperature device to the reference resistor?

    2) The effect of the sink impedance on the current sources.  Will the current sources change value if I switch from a 10K thermistor in a 100 degree bath, approximately 800 ohms, and then switch it to my reference 10,000 ohm +- 0.1% low drift resistor?

    3) Any effect that the internal MUX has on the accuracy of the current sources.  Is it safe to assume that this is zero? Or is there a practical way to calibrate this out?

    4) Must I assume that all of the six current sources are independent?  I.e. If I calibrate the 1mA source will its cal scale to the 50uA source?

     

    Chuck

  • Hi Chuck,

    I have already contacted the product owner regarding your queries, I will get back to you as soon as I have the information.

    I noticed that in your schematic you are using a configuration for a 4 wire RTD, can you confirm if this is correct? If so, I would recommend that you use a ratio metric configuration as discussed in this reference design. This way you would need not to worry about the inaccuracies of the current source and you would also avoid the use of an external reference. The reference design could also be easily modified for thermistor measurement by just replacing the reference resistor and selecting an appropriate excitation current.

    Regards,

    RC

  • Hi Chuck,

    For your question,

    1) The short term stability of the current sources.  Will they change values in a short period of time?  I.e. the time it takes to switch them from one mux output to another?  From the temperature device to the reference resistor?

         -Best performance is achieved when a ratiometric configuration is used. However, you could Ping-Pong between the 2 measurements. The currents have good short term stability. So, after measuring the RTD/thermistor, the next measurement should be to measure the excitation current.

    2) The effect of the sink impedance on the current sources.  Will the current sources change value if I switch from a 10K thermistor in a 100 degree bath, approximately 800 ohms, and then switch it to my reference 10,000 ohm +- 0.1% low drift resistor?

         -Each current sources has an output compliance stated on the datasheet, once it is met, each circuit will see the same current.

    3) Any effect that the internal MUX has on the accuracy of the current sources.  Is it safe to assume that this is zero? Or is there a practical way to calibrate this out?

        -The internal mux has low on-resistance so it has little effect on the accuracy.

    4) Must I assume that all of the six current sources are independent?  I.e. If I calibrate the 1mA source will its cal scale to the 50uA source?

         -To get the best performance, each current should be calibrated. Any mismatch is then removed.

    Regards,

    RC