ADAS1000 RLD Simulation
 I am trying to simulate RLD Circuit as given in Page 35.
I am simulating. with all SW1 to SW5 swithes ON. and trying to understand its frequency response.
Here's the Circuit i am using  for Simulation.
AC Analysis / AC sweep from 0.1Hz to 10Mhz.
question:
1. As it can be observed the Frequency response maximum is -60db. why is this Frequency response starting negative magnitude. since the Loop gain is high and set with the help of 4Mohms it should be Positive right?
2.  What is the expected Frequency response of the RLD Amplifier?. is there any equation?.
3. Is it possible to explain this circuit?.
As per my understanding Vrld = (Vref- Vin)* Rfb/ Rin + Vref     am i right?.
Your support would be of great Help
 adas1000
Thanks & regards
Lokesh Kumar K.
  • Hi Lokesh,
    Correct, the loop gain will be high.
    Typically it's probably going to be set to be x100 or x200.
    In your simulation, when you close all those switches - the parallel 5x10k resistors all result in RIN=2k - so really you need to look at the RFB value - this should be much lower than 4M. This is captured in the snippit you pulled out of the datasheet and also in the datasheet text on page 36.
    There's two ways to configure the RLD.
    1. With external RIN component. (RIN =40k and Rfb - 4M) gain of 100. Sw1-5 = OPEN
    2. Using internal SW1-5 (maybe not all of them...might just be WCT = RA/LA/LL combination). Now Rfb value should change.
    Why is your bode plot showing negative - I would expect it to be positive - can you share the simulation file with me and I'll take a look. Here's what we would have seen in our own simulations.
    Catherine.
  • Dear Ms Catherine,
    Kindly find the Multisim simulation file in this link RLD.zip - Google Drive .
    *  I will correct RFB and check madam. In the 2nd Method as it is descirbed that i need to maintain the Loop gain of desired value. What is the Recommended RLD Loop gain. Is it 100 or 200?.
    *   Have you considered patient resistor in simulation?.
    *   is Rz resistor value fixed?. What is the effect of Rz resistor in the Circuit?.
    Regards
    Lokesh Kumar
  • Hi Catherine,
         I trouble shooted Myself and found out that the Offset value entered in the simulation was wrong. I maintained the Offset value correctly @1.3v  and i got the Response as shown below for Rin = 40k , Rfb = 400k & Cfb = 400pF.
    With this values theoretically  Gain = Rfb/ Rin = 4M/ 40K = 100. & Fc = 100hz.
     And hence the Frequency response has to start from 40db gain. Why is the below information given by your simulations starting from 70db gain?.
    Thanks & regards
    Lokesh Kumar K
  • Hi catherine.
    Can you reply for this question. I need to understand this question madam. I trouble shooted Myself and found out that the Offset value entered in the simulation was wrong.   I maintained the Offset value correctly @1.3v  and i got the Response  for Rin = 40k , Rfb = 400k & Cfb = 400pF.
    * And hence the Frequency response has to start from 40db gain. Why is the below information given by your simulations starting from 70db gain?.
    * Why should we maintain Very high gains of typically 100 or 200 in RLD Loop?.
  • HI Lokesh,
    Thanks for update, glad to hear you resolved your simulation issue.
    In response to your question about the gain, the purpose of the RLD loop is to reject common mode interference (noise, interference from external sources like mains or other patient connected instruments), the higher gain helps achieve this. The effectiveness of the loop would be reduced with gains of <50. The actual gain that you use is up to you - it's set using the external components, so you have flexibility in terms of what suits your application. In our evaluations, we used the component values as listed in the datasheet. They worked well.
    I'll have to look into the simulation plot in a bit more detail and get an answer for you on the 70dB. Appreciate your patience on this.
    Catherine.
  • Hi catherine,
           Thanks for the answer.I Appreciate your help. Looking forward for the answer.
    Regards
    Lokesh