Impedance measurement range for the ADuCM355 respectively AD5940

Thread Summary

The user inquired about the feasibility of performing impedance measurements up to 1 megaohm at 20 mHz using the ADuCM355 or AD5940. The support engineer confirmed that both devices support measurements down to 0.016 Hz and up to 10 megaohms, despite the datasheet stating a range of 0.4 ohms to 10,000 ohms. The M355_Biased_EIS project demonstrates this capability, and the datasheet will be corrected.
AI Generated Content

Hello,

we would like to use the ADUCM355 respectively the ADC5940 for impedance measurements.

On the datasheet of the ADuCM355 (https://www.analog.com/media/en/technical-documentation/data-sheets/ADuCM355.pdf), on page 1 it is stated that the range for impedance measurement is <1 ohm to 10 megaohms. On page 7, the impedance measurement range is specified to be 0,4 ohms to 10,000 ohms. We would like to perform measurements up to about 1 megaohms at a frequency of 0.02 Hz (20 mHz). Is this measurement feasible either for the ADuCM355 or for the AD5940?