AD9517 clk-gen of the AD9467 Eval Board

Hello,

the AD9467 evaluation board clocked with the onboard oscillator runs fine with the HSC-ADC-EVALC FIFO-Board and Matlab as visualization and analysis software.


Now I tried to change the sample frequency by enabling the clock generator AD9517 and configuring it with the SPIController software.

I did the following steps to enable the AD9517:

  • I mounted the capacitors C302 an C303 to route the 250 MHz clock from oscillator to the clock generator.
  • I double checked that the resistor R310 is mounted to enable the chip-select line.
  • I didn't mounted C304 and C305 yet (they foreward the generated clock to the ADC),
    because at first I want to get a working SPI connection to the AD9517.

I loaded the AD9517 config file to the SPIController software, but an error pops up that the SPI connection to the DUT cannot be established.
I also corrected the chip-select line to CSB_1 at the SPIController config, but it still doesn't work!

Unfortunately the user guide of the evaluation board doesn't give any (detailed) instructions on how to use the clock generator!

Any idea what steps i missed?

Thanks!!

  • This may be a bit of a challenge to run down as the Eval board is supported by the ADC team, but the AD9517 is supported by the Clock team.  Both teams have been made aware of your question, though.

    This probably won't be too much help, but the data sheets for the AD9517 family can be quickly accessed here:

    http://search.analog.com/search/ProductSearch.aspx?query=AD9517&locale=en

  • Hello MtheK,

    On behalf of ADI, I apologize with the trouble that you're having with the board.

    Looking at the AD9547 schematic, I see that R309 is a very strong pull-up on CSB that should either be removed or replaced with a 10k resistor.  This p/u resistor may be preventing CSB from going low. It should be easy to probe this.

    Also, the AD9547 schematic appears to use the AD9517 in two ways: simple clock distribution, and as a PLL. Because I don't have the AD9517 register setup file, it's not obvious to me which input on the AD9517 is being used. If you're using the PLL, then install C300/C301. If you're not using the PLL, install C302/C303.

    I also see that C304/305/305/306 are DNI's, so you'll need to address that as well. Make sure that you don't have two sources driving the same input into the AD9467.

    On a minor note, If you notice that the LVPECL performance needs to be a bit better, you can try 180 ohm pull-downs on the AD9517 LVPECL output to increase the drive strength. The 250 ohm pull-downs are ok, and may be fine here, but the 180's should inprove the rise/fall times of the clock.

    -Paul Kern

    Applications Engineer

    I'm checking with my fellow apps engineer on the AD9467 to see

  • Hello Guys!

    After hours of probing, debugging, reading schematics and drinking coffee I didn't very much progress, at least I still cannot connect to the clk-gen.

    • Working throuth the schematic I also found out the the SPI interface of the AD9517 (Clk-Gen) is connected directly to the FIFO-Board (HSC-ADC-EVALC) connector, and the SPI interface of the AD9467 (ADC) is connected via the jumpers at the header P600 -- so I removed all the four jumpers to avoid any SPI interference of the ADC (shouldn't happen because of the differen CS-lines, but just to be on the safe side!).

    • Rob Reeder sent me the evaluation software especially for the AD9517 clk-gen, but unfortunately this software can't connect to the clk-gen either.
      What makes sense to me because the AD9517-SW uses the AD9517 Eval-Board as USB-SPI bridge, and I use the HSC-ADC-EVALC board (the AD9517-SW doesn't even detect my eval board). Is there any way to make the Eval-SW work with the FIFO-Board, maybe a modified USB-Controller firmware or something like that?

    • I probed all the SPI lines (CS, CLK, SDI) and they all work as expected, and the ADC SPI Controller software (from http://www.analog.com/en/analog-to-digital-converters/ad-converters/ad9251/products/evaluation-boardstools/CU_High-Speed_ADC_FIFO_evaluation_tools/resources/fca.html#FPGA, it comes with cfg files for the AD9517 clk-gen !!) sends the commands correctly! I double checked every bit of the transmission with the AD9517 datasheet, the commands are ok!
      The strange thing is that at read commands, the SDO line keeps quite at all, it's stuck to the 3.3V Vdd-level. No response from the chip

      FYI: As mentioned in the first post, I needed to change the cfg-file so that the correct CS-line is used (CSB_2).
    • Is there any easy way to check if the AD9517 chip is still alive and working correctly? ... maybe it's damaged?

    • BTW: The AD9517 uses the SDI line as bidirectional data-line per default; but the SPI-Master (Cypress USB-Controller) uses unidirectional data-lines. I thought through this and think that this shouldn't cause problems -- any other opinions?

    • BTW: In the first post I forgot to mention that i removed the AD9517 power-down jumper at J300, of course

    @Paul: Thanks for all your hints! At the moment I just installed C302/C303 to use the PLL -- this clock should be ok for the chip to work correctly, shouldn't it?

    C304/305/305/306 are also in my mind, but at first I want to be able to configure the chip.

    I'll think about changing the pullup tomorrow at work (it's 23:30 PM in Austria now ), but I think the 1k pullup is ok too because the signal goes down correctly.

    Thanks for the great support!

    Martin Klepatsch

    Infineon Austria GmbH

  • Hello Guys,

    I finally made the clk-gen work today !!

    Don't ask me me what the mistake was, I don't know exactly ...

    At first I wrote a litte C++ tool that allows me to send single read and write commands to the clk-gen via the FIFO-board. I used the FIFO-Interface Library for that. While probing an analysing the SPI bus, i noticed several crappy voltage levels on the the data lines: sometimes 3,3V (ok), sometimes ~1.8V (worked sometimes), and during some transmissions the high-level of the SDI line was only at 0.3V !!!!!

    Just for fun I disconnected the power jack of the ADC-Eval-Board .......... and suddenly it worked => at least the commands and the response from the clk-gen looked perfect at my oscilloscope; with a high-level of 1.4V and a low-level of 0V

    The AVDD_3P3V level was 1.4V (without the power jack connected!).    I concluded that the low 1k pullup resistor of the SPI bus coupled some voltage of 1.4V  to the supply-lane that was enough to supply the control-logic of the A9517 ... I could successfully read out the chip ID an so on *happy*

    So I dediced to disable possible inferers by ONLY supplying the AVDD_3P3V power lane (via a separate DC power supply; needed to remove E706 for that) and let the main power jack disconnected. Actually the chip worked fine, just als well as the SPIController software !

    I wanted to counterproof the problems with the power supply and reassembled E706 and supplying the complete board with the main power jack again,  but now everything worked fine again .... no crappy voltage-levels at the SPI bus .... no connection problems with the SPIController ....  no problems at all!

    To cut a long story short, removing and reassembling E706 magically removed all my problems!

    Tanks for your support!

    Regards,

    Martin K.

  • Hi Martin,

    Thank you for the update and I am glad you have everything up and running.

    We have tried this out in our lab and certainly didn't have the same issues you were encountering.

    If you would like to exchange the evaluation board for a new please let me know.

    Please don't hesitate to contact us if you have any other questions.

    Regards,

    Rob

  • Hi Martin,

    Glad to hear. Thanks for the feedback.

    -Paul Kern

  • Thanks for the offer to replace the board, Rob! I may come back to that later.

    I have one last issue with the oscillator / clk-gen:

    The input clock (source: the stock 250 MHz oscillator) seems to be trippled somewhere in the AD9517. (The VCO divider is bypassed, and I just use the Divider1 to set the ouput frequency).

    If I want to get a output frequency of  50 MHz, I have to set a divider of 15  [M=7, N=6].

    If I want to get a output frequency of  125 MHz, I have to set a divider of 5 [M=2, N=1].

    ... and so on.

    If I enable the internal 1.6GHz VCO, everything seems to work fine.

    Any ideas?

    Maybe you can check this in your laboratory!

    Kind regards,

    Martin K.

    BTW: I did the following modifications to route the clk to the adc:

    unmounted C209/C210

    mounted C304/C305 (clk-gen -> adc)

    mounted C302/C303 (osc -> clk-gen)

    and replaced R3077/R308 with 180's; tanks for the hint, Paul!

  • Hi Martin,

    > The input clock (source: the stock 250 MHz oscillator) seems to be  trippled somewhere in the AD9517.

    > (The VCO divider is bypassed, and I  just use the Divider1 to set the ouput frequency).

    > If I want to get a output frequency of  50 MHz, I have to set a divider of 15  [M=7, N=6].

    > If I want to get a output frequency of  125 MHz, I have to set a divider of 5 [M=2, N=1].

    >... and so on.

    >If I enable the internal 1.6GHz VCO, everything seems to work fine.

    > Any ideas?

    It appears that you're trying to input 250 MHz into the CLK input. If that's true, make sure the PLL is indeed off. It's possible that the internal VCO is still running at 1500 MHz.

    There are no doublers or triplers inside the part. Something isn't configured in the maner that you suspect it is, or the PLL isn't locked (if you are using the PLL).

    -Paul

  • Yes that's what I'm trying to do!

    Take the 250 MHz clock from the on-board oscillator and derive the clock for the ADC (in a range of 50 to 250 MHz) => so no PLL or VCO is used.

    • PLL is powered down (async pwr down)
    • VCO is disconnected (power dn VCO clk interface) => I found no option to power it down...
    • VCO divider is bypassed
    • "Input to VCO divider" is set to "External clk"
    • all ouputs are disabled except output 3 (LVPECL output to the clk-input of the ADC)

    The output frequency should be fout = fin / Dx, but it is fout = (fin*3) / Dx, with Dx = N+M+2.

    I attached the cal-file with my register settings, you can view it with the SPIController software.

    Regards,

    Martin K.

  • This question has been assumed as answered either offline via email or with a multi-part answer. This question has now been closed out. If you have an inquiry related to this topic please post a new question in the applicable product forum.

    Thank you,
    EZ Admin