The user inquired about the meaning and triggering conditions of GRDALM (Guard Alarm) in the AD5560. The final answer clarified that GRDALM is triggered if a 100mV difference is detected between the FORCE and GUARD outputs for at least 200μS, indicating a potential short to GND or another potential. The AD5560 prevents SW17 and SW18 from closing simultaneously to avoid connecting the guard amplifier output to DUTGND.
AI Generated Content
Category: Datasheet/Specs
Product Number: AD5560

Hola

What is GRDALM standing for by Alarm setup register, 0x06?  No information in the datasheet?  How is it triggered by event?

Thread Notes

Parents Reply
  • Hi, .

    After discussing this with the team, these are what we found:

    The difference between FORCE and Guard amplifier output that would trigger the \KELALM is 100mV. Just realized it was stated in the datasheet as well:

    AD5560 has its own scheme that prevents SW17 and SW18 to be closed at the same time, subsequently avoiding a scenario that the guard amplifier output gets connected to the DUTGND.

    The guard amplifier output being shorted to GND or any potential 100mV away from the FORCE may happen external to the part (i.e. if the shield gets shorted to another potential).

    \KELALM is triggered if a difference of 100mV is detected for a minimum duration of 200uS.

    Best regards,
    Mac