I am using MAX32010_EVKIT_B. I use keysight 2231A-30-3 as the power source, set vcc=12V, Vee=-12v, Vdd=5V. And use keysight digital multimeter 34465A to measure voltage between MAX32010 Vmeas pin and gnd. I connect demo kit DUT_NODE and GND_NODE together with a short wire. And connect Sense and DUTGSNS together and then connect to gnd. I hope this configuration can test Vmeas op-amplifer offset voltage.
The evaluation kit setting is as the following picture.
The problem is, when I change the force current value from 0 to 1.2A, the Vmeas value is changing consequently for -2.2mV to 4.3mV. The offset is so big. From datasheet, voltage monitor op-amplifer temp coefficient is only +/-10uv. I can't think out what is the root cause.
By the way, I have do another test. I keep the force currect not change, and change max32010 temperature from 33℃ to 100℃, the Vmeas change no more than 1mV.